7013-C
7013-s
20-channel Isolated Switch Cards
This isolated switch card contains 20 indepen-
dent channels that can be connected in a wide
variety of configurations . Each channel is 2-pole .
The isolated switch configuration provides the
greatest flexibility because the switches can be
connected as needed . Both sides of each 2-pole
relay are available for connection .
ISOLATION: Channel to Channel:
>10
W,
<25pF .
Differential:
>10
W,
<50pF .
Common Mode:
>10
W,
<100pF .
CROSSTALK (1MHz, 50W Load):
<–50dB .
INSERTION LOSS (50W Source, 50W Load):
<0 .1dB below
1MHz, <3dB below 10MHz .
RELAy DRIVE CURRENT (per relay):
16mA .
EMC:
Conforms to European Union Directive 89/336/EEC .
SAFETy:
Conforms to European Union Directive 73/23/
EEC (meets EN61010-1/IEC 1010) .
ENVIRONMENT: Operating:
0° to 50°C, up to 35°C <80% RH .
Storage:
–25°C to 65°C .
Use with 7001 and 7002 switch mainframes
20 independent 2-pole switches
500nV, 100pa offsets
aCCEssorIEs aVaIlablE
for 7013-C
7011-KIT-R 96-pin Female Connector Kit
7011-MTC-1 96-pin Mass Terminated Cable, Female to Female, 1m
7011-MTC-2 96-pin Mass Terminated Cable, Female to Female, 2m
for 7013-s
7013-ST
Extra Screw Terminal Connection Board
ordering Information
7013-C
20-channel, 2-pole
Independent switch with
Connector board
20-channel, 2-pole Inde-
pendent switch with screw
Terminal Connector board
sErVICEs aVaIlablE
7013-C-3Y-EW
7013-S-3Y-EW
1-year factory warranty extended to 3 years
from date of shipment
7013-s
7015-C
7015-s
40-channel Solid State Multiplexer Cards
Quad 1×10 Configuration
The Model 7015 40-channel solid state multi-
plexer is designed for multipoint measurement
applications that require high reliability and
increased scanning speeds . With an MTBF of
more than 30,000 hours, the 7015 can handle
applications that require continuous use over
longer periods of time . The solid state switch
technology also provides fast switching times for
scanning rates of over 300 channels/measure-
ments per second when used with the 7002/2001
or 7001/2001 combination .
MULTIPLEX CONFIGURATION:
4 independent 1×10 2-pole
multiplex banks or 2 independent 1×10 4-pole multiplex
banks . Adjacent banks can be connected together . Jumpers
can be removed to isolate any bank from the backplane .
CONTACT CONFIGURATION:
2-pole Form A (Hi, Lo) .
CONNECTOR TyPE:
7015-C:
96-pin male DIN connector .
7015-S:
Screw terminal, #16AWG maximum wire size, with
0 .092 inch O .D . 28 conductors per card maximum . #22AWG
typical wire size with 0 .062 inch O .D . 88 conductors per card
maximum .
MAXIMUM SIGNAL LEVEL:
175V peak between any two pins,
34mA resistive load, 0 .3VA max ., 1×10
V·Hz max .
COMMON MODE VOLTAGE:
175V peak, any pin to chassis .
CONTACT TyPE:
Solid state switch .
CHANNEL RESISTANCE (per conductor):
<210W .
CONTACT POTENTIAL: 7015-C:
<5µV per channel contact pair .
7015-S:
<4µV per channel contact pair .
OFFSET CURRENT:
<1nA .
ACTUATION TIME:
<500µs .
ISOLATION: Bank:
>10
W,
<25 pF .
Channel to Channel:
>10
W,
<50 pF .
Differential:
Configured as 1×10: >10
W,
<100pF .
Configured as 1×40: >10
W,
<200pF .
Common Mode:
W,
<375pF .
W,
<1100pF .
INSERTION LOSS (50W Source, 1MW Load):
<0 .1dB below
250kHz, <3dB below 500kHz .
Quad 1×10 (40-channel)
solid-state multiplexer
30,000 hours MTbf
scan/measure over 300 ch/s
for 7015-C
SWITCHING AND CONTROL
7015-C
40-channel, 2-pole
Connector board
40-channel, 2-pole
Independent switch
with screw Terminal
Connector board
sErVICEs aVaIlablE
7015-C-3Y-EW
7015-S-3Y-EW
7015-s
1.888.KEITHLEY
(U.S. only)
www.keithley.com
A Greater Measure of Confidence
RELAy SWITCH CONFIGURATION:
20 independent channels
of 2-pole switching .
CONNECTOR TyPE: 7013-C:
7013-S:
maximum .
MAXIMUM SIGNAL LEVEL: DC Signals:
110V DC between any
two pins, 1A switched . 30VA (resistive load) .
AC Signals:
125V rms and 175V AC peak, between any two
pins, 1A switched, 60VA (resistive load) .
CONTACT LIFE: Cold Switching:
closures .
At Maximum Signal Levels:
closures .
<1W .
OFFSET CURRENT:
<100pA .
CONTACT POTENTIAL: 7013-C:
<2µV per channel contact pair
(HI, LO); <5µV per single contact .
7013-S:
<2µV per contact
pair (HI, LO); <5µV per single contact .
ACTUATION TIME:
3ms .