The MBC350 Series of open frame medical power supplies feature a
wide universal AC input range of 90 – 264 VAC, offering 350 W of
output power in a compact 3 x 5 x 1 inch footprint, with a variety of
isolated single output voltages.
The MBC series is designed and approved to the latest Medical
standards (EN/IEC 60601-1), providing 2 x MOPP isolation for Class I
& Class II applications.
These power supplies are ideal for medical, telecom, datacom,
industrial equipment and other applications.
3 x 5 x 1 Inch Form Factor
350 W with Forced Air Cooling & 200 W with Convection Cooling
Approved to EN/IEC 60601-1
Efficiencies up to 94%
-40 to 70°C Operating Temperature
12 V / 0.5 A Fan Output, Thermal Shut-Down Feature
2.56 Million Hours, Telcordia -SR332-Issue 3 MTBF
Standby Power < 0.5 W
Medical (BF) Safety Approvals
RoHS Compliant
Diagnostic
Drug Pump
Dialysis
Home Health Care
Monitoring
Portable Equipment
MBC350 Series
VOLTAGE
MAX. LOAD
(CONVECTION)
MAX. LOAD
(375 LFM)
MIN. LOAD
RIPPLE & NOISE
MODEL NUMBER
Specifications are for nominal input voltage, 25°C unless otherwise stated.
PARAMETER
Input Voltage
Input Frequency
Input Current
No Load Power
Inrush Current
Leakage Current
Power Factor
Switching Frequency
115 VAC:
230 VAC:
Typical
115 VAC:
230 VAC:
264 VAC:
Typical
Touch Current
Full Load
PFC:
PWM:
DESCRIPTION / CONDITION
Universal
(Derate from 100% at 100 VAC to 90% at 90 VAC)
SPECIFICATION
90-264 VAC / 390 VDC
47 - 63 Hz
3.6 A max.
1.8 A max.
>0.5 W
25 A
45 A
75 A
300 uA
<100 uA
>0.95
70 - 130 KHz
50 - 80 KHz
tech.support@psbel.com
MBC350 Series
PARAMETER
Output Power
Efficiency
Hold-up Time
Line Regulation
Load Regulation
Transient Response
Voltage Adjustment
Rise Time
Set Point Tolerance
Over Current Protection
Over Voltage Protection
Short Circuit Protection
Hiccup mode / Auto Recovery
Typical
50-100% step load change, at 0.1A/uS slew rate,
50% duty cycle, 50 Hz = 5% ,
With 375 LFM:
Convection:
48 V, 58 V:
24 V, 30 V:
12 V, 15 V:
Full Load:
Convection Load:
SPECIFICATION
350 W
200 W
94%
93%
92%
8 ms typical
14 ms typical
+/-0.5%
+/-1%
recovery time < 5 ms
+/-3%
55 ms
+/-1%
>110%
110 to 140%
PARAMETER
Storage Temperature
Cooling
Altitude
Humidity
Reliability
Startup guaranteed, with spec deviation, see Fig. 1
With 375 LFM forced air cooling at 100 to 264 VAC:
With natural convection cooling at 100 to 264 VAC:
Operating:
Non-operating:
Non Condensing
MTBF according to Telcordia - SR332-Issue 3
SPECIFICATION
-40 to +70°C
-40 to 0°C
-40 to 85° C
350 W
200 W
16,000 ft.
40,000 ft.
5% to 95%
2.56 million hours
Conducted Emissions
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Radiated Emissions
Surge Susceptibility
Harmonic Current
AC Flicker
EN55022-B, CISPR22-B, FCC PART15 – B
EN61000-4-2:
EN61000-4-3:
EN61000-4-4:
Radiated:
Radiated with external core:
(King core K5B RC 25x12x15-M in input cable with 5 Turns)
EN61000-4-5:
EN61000-3-2:
EN61000-3-3:
Level-3
Level-3
Level-3
Level A
Level B
Level-3
Class D
Pass
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00561_AE
© 2017 Bel Power Solutions & Protection
MBC350 Series
Isolation Voltage
Protection Level
Safety Standard(s)
Agency Approvals
Input to Output: (For medical applications)
Input to GND: (Not Applicable For Class II Option)
Output to GND: for type BF
for type B (N/A For Class II Option)
Primary to Secondary:
Primary to Earth:
Secondary to Earth:
Approved to the latest edition of the following standards:
CSA/UL60601-1, EN60601-1 and IEC60601-1.
Nemko, UL, C-UL
4000 VAC
1500 VAC
1500 VAC
500 VAC
2 MOPP
1 MOPP
1 MOPP
CE mark
Complies with LVD Directive
Figure 1. Derating Curves
MBC350 Series
CONNECTOR
AC Input Connector
PIN
J1
Screw Terminal
(Option 1)
Molex Connector
(Option 2)
Aux (Fan) Output
Earth
J3
J4
Pin 1
Pin 2
Pin 3
Pin 1
Pin 2
Pin 1,2,3,4
Pin 5,6,7,8
Pin 1
Pin 2
AC Line
Not Fitted
AC Neutral
V1 +VE
V1 - VE
V1 +VE
V1 - VE
FAN +VE
FAN -VE
MANUFACTURER / PN
Molex: 26-60-4030
Mating: 09-50-3031; Pins: 08-50-0106
6-32 inches Screw Pan HD
Mating: 16 AWG wire crimped to Ring
Tongue Terminal AMP: 8-31886-1
Molex: 26-60-4080
Mating: 09-50-3081; Pins: 08-50-0106
AMP :640456-2
Mating: 640440-2
Molex: 19705-4301
Mating: 19003-0001
DC Output Connector
J2
Weight
Dimensions
300 g
76.2 x 127.0 x 25.4 mm (3 x 5 x 1 inch)
Figure 2. Mechanical Drawing - Screw Terminal (Option 1)
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00561_AE