The MBC300 Series of open-frame medical power supplies, with its
wide universal 90-264 VAC input range, is available at 300 W of output
power and a variety of single output voltages.
The MBC series was designed to 4th edition medical approvals and
provides 2x MOPP (Means of Patient Protection) isolation for Class I
and Class II installations.
These medical power supplies are ideal for monitoring, home health
equipment as well as surgical devices.
200 W convection cooled
-20 to 50°C full load operation
2x MOPP
5.0 x 3.0 x 1.5 inch (127.0 x 76.2 x 38.1 mm)
12 V fan & 5 V standby outputs
Inhibit and Power Good signals
No minimum load required
IEC Protection Class Options:
o
Class I: Earth pin J4 (no suffix)
o
Class II: No Earth pin (-2 suffix)
Conducted EMI EN 55022-B, FCC Part 15 Level B
Medical Safety Agency Approvals
RoHS Compliant
CE marked
Diagnostic
Drug Pump
Dialysis
Home Health Care
Monitoring
Imaging
MBC300 Series
MODEL
CONNECTOR
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
Screw Terminal
OUTPUT
VOLTAGE
5 VDC
5 VDC
12 VDC
12 VDC
15 VDC
15 VDC
24 VDC
24 VDC
30 VDC
30 VDC
48 VDC
48 VDC
MAX LOAD
CONVECTION
1,2,3
28.0 A
28.0 A
16.67 A
16.67 A
13.33 A
13.33 A
7.5 A
7.5 A
6.0 A
6.0 A
3.75 A
3.75 A
MAX LOAD
300 LFM
1,2,3
40.0 A
40.0 A
25.0 A
25.0 A
20.0 A
20.0 A
13.54 A
13.54 A
10.83 A
10.83 A
6.77 A
6.77 A
MINIMUM
LOAD
0A
0A
0A
0A
0A
0A
0A
0A
0A
0A
0A
0A
RIPPLE &
NOISE
2%
2%
2%
2%
2%
2%
2%
2%
2%
2%
2%
2%
TOTAL
REGULATION
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
± 2.5%
MBC300-1T05G
MBC300-1T05G-2
MBC300-1T12G
MBC300-1T12G-2
MBC300-1T15G
MBC300-1T15G-2
MBC300-1T24G
MBC300-1T24G-2
MBC300-1T30G
MBC300-1T30G-2
MBC300-1T48G
MBC300-1T48G-2
Cover-300-XCB
Metal cover kit accessory
NOTES:
Peak current rating on main output is 120% of max., lasting < 30 s with a maximum 10% duty cycle.
Combined output power of main output, fan supply and standby supply shall not exceed max. power rating.
Derate output power linearly to 80% from 90 VAC to 80 VAC input.
Ripple is peak to peak with 20 MHz bandwidth and 10
μF
(Tantalum capacitor) in parallel with a 0.1
μF
capacitor at rated line
voltage and load ranges.
Specifications are for nominal input voltage, 25°C unless otherwise stated.
PARAMETER
Input Voltage
Input Frequency
Input Current
Inrush Current
Leakage Current
Switching Frequency
120 VAC
230 VAC
120 VAC
230 VAC
120 VAC
230 VAC
PFC converter (fixed)
Resonant converter (variable)
DESCRIPTION / CONDITION
Universal
SPECIFICATION
90-264 VAC / 120-390 VDC
47 to 63 Hz
3.2 A max
1.65 A max
35 A max
65 A max
< 125
μA
<250
μA
80 kHz typical
35 to 250 kHz, 90 kHz typical
PARAMETER
Output Power
Efficiency
Hold Up Time
Power Factor
Line Regulation
Load Regulation
Transient Response
Rise Time
120 VAC
230 VAC
120 / 230 VAC
120 VAC
230 VAC
SPECIFICATION
200 to 325 W
88% typical
92% typical
10 ms
0.98
0.95
+/-0.5%
+/-2%
< 10%, recovery time < 5 ms
< 100 ms
50% to 100% load change, 50 Hz, 50% duty cycle, 0.1 A/μs,
tech.support@psbel.com
MBC300 Series
Set Point Tolerance
Voltage Output Adjustment
Over Voltage Protection
Over Current Protection
Short Circuit Protection
Over Temperature Protection
Short term, automatic recovery
Automatic Recovery
110° C primary heat sink
Automatic recovery
± 1%
±3%
110 to 150 %
110 to 150 %
NOTES:
Standby output voltage tolerance including set point accuracy, line and load regulation is +/-10%. Ripple and noise is less than 5%.
Fan supply output voltage tolerance including set point accuracy, line and load regulation is +/-30% and needs min. 1% load on main
output to be within regulation band. Ripple and noise is less than 10%.
PARAMETER
Storage Temperature
Refer to de-rating curves (Fig. 1)
Start-up is guaranteed
5 V model
12 V, 15 V, 24 V, 30 V & 48 V models
5 V model
12 V and 15 V models
24 V, 30V and 48 V models
SPECIFICATION
-20 to 70°C
-20 to 0°C
-40 to 85° C
140 W max
200 W max
200 W max
300 W max
325 W max
95% Rh
10,000 ft.
40,000 ft.
1.77 million hours
Convection:
Cooling*
With 300LFM:
Relative Humidity
Altitude
Reliability
Non Condensing
Operating:
Non-Operating:
MTBF according to Telcordia –SR332-issue 3
* Refer de-rating curves (Fig. 1) to determine output power over the entire operating temperature range
Figure 1. Output Power Vs. Temperature Derating Curves
NOTE:
The de-rating curves are valid for input voltages of 115 VAC to 264 VAC. Below 115 VAC to 90 VAC the convection rating is 180 W max.
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00290_AD
© 2017 Bel Power Solutions & Protection
MBC300 Series
EN55022-B, CISPR22-B, FCC PART15-B
EN61000-3-2
EN61000-4-2
EN61000-4-3
EN61000-4-4
EN61000-4-5
Non Condensing
Operating:
Non-Operating:
PARAMETER
Conducted Emissions
Radiated Emissions
Harmonic Current
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Surge Susceptibility
Humidity
Altitude
SPECIFICATION
Class II product meets Class A limit
line for conducted emission.
To be controlled in end system
Class D
Level-3
Level-3
Level-3
Level 3
95%
10,000 ft.
40,000 ft.
PARAMETER
Isolation Voltage
Safety Standards
Agency Approvals
CE mark
Input to Output:
Approved to the latest edition of the following standards:
CSA/UL60601-1, EN60601-1 and IEC60601-1.
Nemko, Nemko-CCL
Complies with LVD Directive
SPECIFICATION
4242 VDC min
PARAMETER
Power Good *
Remote On/ Off
Remote Sense
TTL signal goes high after main output is within regulation band, delay is 0.1 to 0.3 s
To turn on PSU short remote pin to ground
Compensates for 200 mV cable drop
NOTE:
Power good signal cannot be used as a current source. Internal pull up resistor from PG signal to 5 V is 10K.
It is recommended to use external transistor if intended to source current.
CONNECTOR
AC Input Connector
PIN
J1
Pin 1
Pin 2
AC LINE
AC NEUTRAL
MANUFACTURER / PN
Molex: 26-60-4030
Mating: 09-50-3031; Pins: 08-50-0106
6-32 inches Screw Pan HD
Mating: Designed to accept Ring Tongue Terminal AMP : 8-31886-1,
wherein one 16 AWG (max) wire can be crimped.
Note: One Ring Tongue Terminal with 16 AWG is recommended for
current up to 11 A only.
Use multiple tongue terminals with wire for more current.
DC Output Connector
J2
Pin 1
Pin 2
RTN
V1
Signals & Aux Power
J3
Pin 1
Pin 2
Pin 3
Pin 4
Pin 5
Pin 6
Pin 7
Pin 8
Earth
J4
REMOTE ON/OFF
RTN
VFAN (+12 V/0.5 A)
-VE REMOTE SENSE
VSTBY (+5 V/2 A, +/-5%)
+VE REMOTE SENSE
RTN
POWER GOOD
Spade Connector
(Class I product only)
Molex: 22-23-2081
Mating: 22-01-2087; Pins: 08-50-0113
Molex: 19705-4301
Mating: 190030001
NOTE:
PSU is supplied with J3 housing, pin-1 and pin-2 shorted to enable main output without remote on/off feature.
MBC300 Series
PARAMETER
Weight
Dimensions
450 g (0.99 lbs)
127.0 x 76.2 x 38.1 mm (5.0 x 3.0 x 1.5 inch)
Figure 2. Mechanical Drawing
NOTES:
In case the PCB is mounted in a metal enclosure, using metal hardware ensure the following:
Stand off, used to mount PCB has OD of 5.4 mm max.
Screws, used to fix PCB on stand off, have head dia of 6.0 mm max.
Washer, if used, to have dia of 6.5 mm max.
NUCLEAR AND MEDICAL APPLICATIONS
- Products are not designed or intended for use as critical components in life support systems,
equipment used in hazardous environments, or nuclear control systems.
TECHNICAL REVISIONS
- The appearance of products, including safety agency certifications pictured on labels, may change depending on the
date manufactured. Specifications are subject to change without notice.
Asia-Pacific
+86 755 298 85888
Europe, Middle East
+353 61 225 977
North America
+1 408 785 5200
BCD.00290_AD