MINT1180
180W Single Output
Medical Series
3” x 5” x 1.3” Package
180W Convection Cooled Power
Ideal for 1U Applications
Class I and Class II versions
Universal Input 90-264 VAC
Approved to CSA/EN/IEC/UL60601-1, 3
rd
Edition, 2 x MOPP Isolation
Approved to CSA/EN/IEC/UL60950-1 2
nd
Edition
Efficiency 90% typical
3 Year Warranty
RoHS Compliant
The
MINT1180 product family is the most cost-effective medically approved supply in the market with 180 watts
of convection cooled output power in 3”X5” footprint. The series comes in eight different models ranging from
12 to 48V and has been approved to IEC60601-1 3
rd
Edition, rated for 2 MOPP (Means of Patient Protection)
isolation. This new single output power supply family has both Class I and II input versions and accepts wide
input range of 90 to 264Vac. The typical 90% efficiency and convection cooled make this supply an excellent
solution for HomeCare applications where the use of fans is not desired.
Model Selection
Model Number
MINT1180A1275K01
MINT1180A1575K01
MINT1180A1875K01
MINT1180A2475K01
MINT1180A2875K01
MINT1180A3275K01
MINT1180A4875K01
Notes:
Volts
12V
15V
18V
24V
28V
32V
48V
Output
Current*
15.0A
12.0A
10.0A
7.50A
6.40A
5.62A
3.75A
Minimum Load
0A
0A
0A
0A
0A
0A
0A
Ripple &
Noise**
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
280mV pk-pk
320mV pk-pk
480mV pk-pk
Total
Regulation
±3%
±3%
±3%
±3%
±3%
±3%
±3%
OVP Threshold
14.0
1.1V
18.5 ± 1.2V
21.5 ± 2.0V
29.0 ± 2.5V
33.5 ± 2.5V
36.0 ± 3.0V
56.0 ± 3.0V
* Total convection power is 180 Watts.
** Measured with noise probe directly across output terminals, and load terminated with 0.1µF ceramic and 10µF low ESR capacitors.
General Specifications
AC Input
100-240Vac, ±10%, 47-63Hz, 1∅
120–370Vdc
115Vac: 1.8A, 230Vac: 0.9A
Turn On Time
Less than 3 sec. @115Vac & Full Load
Input Current
Hold-up Time
>16mS at 180W, 120Vac
www.slpower.com
MINT1180 REV 1.8, 15-OCT-13
MINT1180
180W Single Output
Medical Series
(continued)
Inrush Current
264Vac, cold start: will not exceed 55A
Signals
Overload
Protection
Short Circuit
Protection
Overvoltage
Protection
Switching
Isolation
Operating
Temperature
Temperature
Derating
Storage
Temperature
Altitude
N/A
Input Fuses
Earth Leakage
Current
Efficiency
F1, F2: 4.00A, 250VAC fuses provided on all
models
120%-150% of rating, Hiccup Mode
<275µA@264Vac, 60Hz, NC; <400µA SFC
Provided - no damage will occur if the output is
shorted, auto recovery
OVP latch at 110%-130% of rated output
voltage.
PFC: Fixed, 65kHz
Main Converter: Variable 35-200kHz, 65-70kHz
at full load.
Input-Output: 4000Vac, 2 x MOPP
Input-Ground: 1800Vac, 1 x MOPP
Output-Ground: 1500Vac
-10° to +70°
Start Up at -40° full load
C,
Derate output power linearly above 50° to 50%
at 70°
88% typical
Output Power
Transient
Response
Ripple and Noise
180W convection cooled
500µS typ. for return to within 0.5% of
nominal, 50% load step.
∆i/∆t
<0.2A/µS.
Max Volt Deviation = 3%
See chart
Output Voltage
Adjustability
Minimum Load
See chart
Fixed output
-40° to +85°
Not required
Operating: -500 to 10,000 ft.
Non-operating: -500 to 40,000 ft.
Total Regulation
+/- 3% combined line, load and initial setting.
Operating: 0.003g²/Hz, 1.5grms overall, 3
axes, 10 min/axis
Non-Operating: 0.026g²/Hz, 5.0grms overall,
3 axes, 1 hr/axis
W: 3.0” x L: 5.0” x H: 1.3”
Relative Humidity
5% to 95%, non-condensing
Operating: Half-sine, 20gpk, 10ms, 3 axes, 6
shocks total
Non-Operating: Half-sine, 40 gpk, 10 ms, 3
axes, 6 shocks total
EN/CSA/UL/IEC 60601-1, 3
rd
Edition
EN/CSA/UL/IEC 60950-1, 2
nd
Edition
Vibration
Shock
Dimensions
Safety Standards
Weight
325g
MTBF
214,194 hours, 25°C, 110Vac
Output vs. Temperature Derating Curve
180W convection cooled, derate output power to 50% at 70°C.
Output Power (W)
Convection
Temperature ( C)
www.slpower.com
MINT1180
180W Single Output
Medical Series
EMI/EMC Compliance
Conducted Emissions
Radiated Emissions
Static Discharge Immunity
Radiated RF Immunity
EFT/Burst Immunity
Line Surge Immunity
Conducted RF Immunity
Power Frequency Magnetic Field Immunity
Voltage Dip Immunity
Line Harmonic Emissions
Flicker Test
EN55011/22 Class B, FCC Part 15, Subpart B, Class B
EN55011/22; FCC Part 15, Class A with 6dB Margin
EN61000-4-2, Criteria A, 6kV Contact Discharge, 8kV air discharge
EN61000-4-3, 3V/m. Criteria A.
EN61000-4-4, 2kV/5kHz, Criteria A
EN61000-4-5, 1kV differential, 2kV common-mode, Criteria A
EN61000-4-6, 3Vrms, Criteria A
EN61000-4-8, 3A/m, Criteria A
EN61000-4-11, 5% Vnom: 0.5cycle; 40% Vnom: 5 cycles, 70% Vnom:
25 cycles, Criteria A
EN61000-3-2, Class A, B, C, & D
EN61000-3-3, Complies (dmax<6%)
Mechanical Drawing
Notes:
1. All dimensions in inches (mm), tolerance
is +/-0.000”.
2. Mounting holes should be grounded for
EMI purposes.
3. FG is safety ground connection.
4. The power supply requires mounting on
metal standoffs 0.20” (5mm) in height, min.
[4] x 0.156
4 PLCS
MTG HOLES
Connector Information
Input Connector
J100
PIN 1) AC LINE
PIN 2) EMPTY
PIN 3) AC NEUTRAL
Mating Connector:
AMP 640250-3
Pins: 640252-2
Ground
(FG)
0.25” FASTON TAB
DC Output Connector
J300
Term. 1,2,3: RTN
Term. 4,5,6: +Vout
Mating Connector:
AMP 640250-6
Pins: 640252-2
Mating Connector:
Molex 190020001
Data Sheet © 2013 SL Power Electronics Corp. The information and specifications contained herein are believed to be correct at the time of publication. However, SL Power accepts no responsibility for
consequences arising from reproduction errors or inaccuracies. Specifications are subject to change without notice.
www.slpower.com