Micro Displacement Sensor
Compact, High-performance
Meets various needs with high performance at the
Z4D-F
resolution of 5
µm.
Numerous applications are possible when the
analog output model is integrated with OMRON’s
K3TX Intelligent Signal Processors.
ON/OFF output model (NPN open collector) is
available.
Ordering Information
Z4D-F
Output
Model
Z4D-F04A
Z4D-F04D
Analog
ON/OFF
Accessories (Attached)
Item
Z4D-F04A
1 pc.
2 pcs.
1 pc.
1 copy
1 pc.
2 pcs.
1 pc.
1 pc.
1 copy
Z4D-F04D
Mounting bracket
Mounting screws (M3 x 2)
Connector with 1-m cord
Screwdriver
Operation Manual
Z4D-F
Z4D-F
Ratings/Characteristics
Item
Z4D-F04A
12 to 24 VDC±10%, ripple voltage (p-p): 10 mV max.
35 mA max.
4±1.25 mm
Red LED with a wavelength of 700 nm
1.5 x 1.5 mm max. (at the sensing range of 4 mm)
200 kΩ min.
Z4D-F04D
Supply voltage
Current consumption
Sensing range
(see note 1)
Light source
Spot diameter
(see note 2)
Connection imped-
ance
(see note 3)
Resolution
(see note 4)
Hysteresis
(see note 1)
Linearity (see note 5)
Analog output
(see note 1)
Control output
Residual output volt-
age
Response time
(see note 6)
Temperature influ-
ence (see note 7)
Indicator
µm
max.
±1%
FS
1 to 5 V (1.6 V/mm±10%)
5 ms max.
0.15% FS/°C
µm
NPN open collector, 12 to 24 VDC, 50 mA max.
1 V max.
1.5 ms max.
PWR indicator (green): Lit when power is supplied
STB indicator (green): Lit when illumination is suffi-
ALM indicator (red): Lit when illumination is insufficient. cient.
OPE indicator (orange): Lit when output is ON.
Operating: –10°C to 55°C (with no icing)
Storage: –15°C to 60°C (with no icing)
Operating: 35% to 85% (with no icing)
Operating: 3000 lx max. (incandescent lamp), 10,000 lx max. (sunlight)
IP50 (IEC standard)
Destruction: 10 to 500 Hz (1-mm max. single amplitude) for 11 min for 3 times each in X, Y, and Z directions
Destruction: 500 m/s
for 3 times each in X, Y, and Z directions (total of 18 times)
1,000 VAC for 1 min between current carry parts and case
0.6 N·m max.
ABS (Polycarbonate for the sensing part)
Approx. 10 g (without connector cable)
20 MΩ min. (at 500 VDC)
1m
Ambient temperature
Ambient humidity
Ambient illumination
Enclosure rating
Vibration resistance
Shock resistance
Dielectric strength
Tightening torque
Material
Weight
Insulation resistance
Cable length
1.
The values are for N8.5 Munsell paper.
2.
The spot diameter is defined by 1/e
of the Sensor’s laser beam center. Although a sub-beam may be observed around the main
beam, this does not mean that the product is defective.
3.
Refers to input impedance of a device to be connected.
4. Resolution (Z4D-F04A)
The resolution is the peak-to-peak displacement conversion value of the analog displacement output (Conditions: N8.5 Munsell
mat paper at the center of measurement point with a power ripple noise of 10 mV (p-p) max.)
Output fluctuation
5.
The value is the peak-to-peak error rate of a displacement output voltage line with N8.5 Munsell mat paper. The value varies
with the sensing object.
Z4D-F
Z4D-F
6. Response Time (Z4D-F04A)
The response time of the Sensor (Z4D-F04A) is the time required for the analog displacement output to increase from 10% to
90% of the full value (at the rise time) or decrease from 90% to 10% of the full value (at the fall time).
Displacement
value
tr: rise time
tf: fall time
Analog dis-
placement
output
The response time of the Sensor (Z4D-F04D): ON/OFF switching time
7.
The value is at the sensing range of 4 mm.
Engineering Data (Reference Value)
Characteristics Data
Angle Characteristics with Difference in Horizontal
Inclination
Standard paper
PBT
SPCC
Analog Output Model (Z4D-F04A)
Sensing Distance vs. Analog Output Characteristics
(Typical)
Linear output (V)
Error (%FS)
Positive
inclination
Negative
inclination
Sensing distance (mm)
Inclination angle (°)
Angle Characteristics with Difference in Vertical Inclination
Standard paper
PBT
SPCC
Analog Output Characteristics with Difference in Sensing
Object
Standard paper
PBT
SPCC
Error (%FS)
Positive
inclination
Negative
inclination
Distance (mm)
Error (%FS)
Z4D-F
Temperature Characteristics
Vcc = 24 V
Vcc = 10.8 V
Vcc = 26.4 V
Z4D-F
Error (%FS)
Temperature (°C)
Resolution Characteristics with Difference in Reflection Rate
Resolution (µm)
Reflection factor (%)
ON/OFF Output Model (Z4D-F04D)
Sensing Distance vs. Adjustor Angle Characteristics
Standard paper
PBT
SPCC
Adjustor angle (°)
The PBT plastic referred in this data was detected under
its natural state.
Z4D-F
Z4D-F
Nomenclature
Analog Output (Z4D-F04A)
(2)
(1) Power (PWR) Indicator
The green PWR indicator is lit when power is properly supplied to
the Sensor.
(2) Alarm (ALM) Indicator
The red ALM indicator is lit if there is no sensing object or the
reflectivity of the sensing object is low and it is impossible for the
Sensor to process the light reflected to the Sensor. The indicator
is lit when the amount of light is less than what is required for
computation by the Sensor. When the ALM indicator is lit, make
necessary adjustments so that the indicator turns OFF.
Appropriate amount of light:
Insufficient amount of light:
Not lit
Lit in red
ON/OFF Output (Z4D-F04D)
(3)
(4)
(5)
(3) Stability (STB) Indicator
The green STB indicator is lit when the Sensor is receiving suffi-
cient light to process. If the indicator is not lit, adjust the amount of
light to turn on the indicator.
(4) Operation (OPE) Indicator
The Sensor is synchronized with the ON/OFF output and the
orange OPE indicator is lit when the Sensor has ON output.
(5) Distance Adjustor
The ON/OFF position can be set with the distance adjustor if a
sensing object is at the sensing range (i.e., 4±1.25 mm). The ON/
OFF position will shift farther from the Sensor if the distance
adjustor is turned clockwise and the ON/OFF position will shift
closer to the Sensor if the distance adjustor is turned counter-
clockwise.