Embedded Power for
Business-Critical Continuity
NLP110 Medical
Single and
Triple Output
Total Power:
80 - 110 W
Input Voltage:
90 - 264 Vac
# of Outputs:
Single, triple
Rev. 10.06.11_155
NLP110 Medical Series
1 of 4
Special Features
Provides low voltage outputs
Compliance to EN61000-3-2
(Power Factor = 0.98)
Package 6.5 x 3.0 x 1.26 inches
90 Vac to 264 Vac universal input
range
UL, cUL, and VDE safety approvals
Overvoltage and short circuit
protection
VDE0871-A, EN55022-A,
EN55011-A conducted noise
EN55011-A radiated noise
Meets all applicable and relevant
immunity standards EN61000-4-
2,4,5 and 6
RoHS compliant
2 year warranty
Electrical Specifications
Input
Input ra nge:
Frequency:
Input surge current:
(cold start)
Safety ground
leakage current:
Input current:
Universal input
(See Note 6)
120 Vac
230 Vac
230 Vac, 50 Hz
120 Vac @ 80 W
120 Vac @ 110 W
230 Vac @ 80 W
230 Vac @ 110 W
UL/IEC127
Main output
Auxiliary outputs
At turn-on
Main output
75% to 100%
step at 0.1 A µs
Main outputs
Cyclic operation
Single and multiple
90-264 Vac
47-63 Hz
18 A max.
35 A max.
100 µA
0.9 A rms
1.3 A rms
0.48 A rms
0.7 A rms
F3.15A H, 250 Vac
±2.0%
±5.0%
1.0 s, max.
5.0% or 250 mV
max. dev., 1 ms max.
recovery to 1%
±0.02%/°C
125%, ±10%
Continuous
See table
Input fuse:
Output
Total regulation:
(line and load)
Rise time:
Transient response:
Safety
• TUV EN60601-1/IEC601
• UL2601-1 2
nd
edition
• CSA C22.2 No. 601.1-M90
File No. E182560
Temperature coefficient:
Overvoltage protection:
Short circuit protection:
Minimum output current:
Embedded Power for
All specifications are typical at nominal input, full load at 25 °C unless otherwise stated.
EMC Characteristics
Conducted emissions:
Radiated emissions:
ESD air:
ESD contact:
Surge:
Fast transients:
Radiated immunity:
Conducted immunity:
EN55022, FCC part 15
EN61000-4-2
EN61000-4-2
EN61000-4-5
EN61000-4-4
EN61000-4-3
EN61000-4-6
120 Vac, 60 Hz
120 Vac. 80 W (-9905J)
Input/output
Input/chassis
Level A
Level A
Compliant
Level 3
Level 3
Level 3
Level 3
Level 3
Level 3
16.7 ms @ 110 W
75% min.
4000 Vac
1500 Vac
Rev. 10.06.11_155
2 of 4
Harmonic current emission correction: EN61000-3-2
General Specifications
Hold-up time:
Efficiency:
Isolation voltage:
Approvals and standards pending:
Weight:
MTBF demonstrated:
MIL-HDBK-217F
EN60601-1, IEC60601-1, VDE0750, UL2601-1, C22.2 No. 601.1-M90
283 g (10 oz)
220,000 hours min.
Environmental Specifications
Thermal performance:
Operating, ambient
(see derating curve)
Non-operating
50 °C to 70 °C ambient convection cooled
0 °C to +50 °C, ambient convection cooled
Peak (0 °C to +50 °C, 60 s)
Relative humidity:
Altitude:
Vibration (See Note 5):
Shock
Non-condensing
Operating
Non-operating
5 - 500 Hz
per MIL-STD-810E
0° C to +50 °C
-40 °C to +85 °C
Derate to 50% load
80 W
(See Note 3)
5% to 95% RH
10,000 feet max.
30,000 feet max.
2.4 G rms peak
516.4 Part IV
0 °C to +50 °C, ambient, 150 LFM forced air 110 W
DERATING CURVE
Output Power (Watts)
110W
80W
150 LFM FORCED
AIR COOLING
NATURAL
CONVECTION
COOLING
55W
40W
0C
10C 20C 30C 40C 50C 60C 70C
Embedded Power for
Ordering Information
Output
+5 V
+12 V
+48 V
+24 V
+5 V (A)
+3.3 V (B)
+12 V
+3.3 V (A)
+2.5 V (B)
+12 V
+12 V (A)
+3.3 V (B)
-12 V
+12 V (A)
+5 V (B)
-12 V
Min
(6)
0.5 A
0.3 A
0 A
0.2 A
0.5 A
0.2 A
0 A
0.5 A
0.1 A
0 A
0.2 A
0.5 A
0 A
0.2 A
0.2 A
0 A
15 A
6.4 A
1.6 A
3.2 A
13 A
13 A
0.65 A
13 A
13 A
0.65 A
6.5 A
13 A
0.65 A
6.5 A
13 A
0.65 A
150 LFM
(2)
22 A
9.2 A
2.3 A
4.6 A
16 A
16 A
1 A
16 A
16 A
1 A
8.5 A
16 A
1 A
8.5 A
16 A
1 A
300 LFM
(3)
22 A
9.2 A
2.3 A
4.6 A
18 A
20 A
1 A
20 A
20 A
1 A
8.5 A
20 A
1 A
8.5 A
18 A
1 A
Peak
(4)
22 A
11.5 A
2.5 A
6.0 A
18 A
20 A
1 A
22 A
22 A
1 A
10 A
22 A
1 A
10 A
22 A
1 A
Ripple
(5)
50 mV
120 mV
240 mV
240 mV
50 mV
50 mV
120 mV
50 mV
50 mV
120 mV
1200 mV
50 mV
120 mV
120 mV
50 mV
120 mV
Total
Regulation
± 2.0%
± 2.0%
± 2.0%
± 2.0%
± 2.0%
± 2.0%
± 5.0%
± 2.0%
± 2.0%
± 5.0%
± 2.0%
± 2.0%
± 5.0%
± 2.0%
± 2.0%
± 5.0%
NLP110-9908J
NLP110-9995J
NLP110-9994J
Model Numbers
(7, 14, 15)
NLP110-9905J
NLP110-9912J
NLP110-9917J
NLP110-9924J
NLP110-9993J
Rev. 10.06.11_155
3 of 4
Free air convection.
Multiple output units: maximum continuous output power not to exceed
80 W.
For -9993J; I
3.3 V
= 13 A max.; I
5.0 V
= 13 A max.; I
3.3 V
+ I
5.0 V
² 16 A.
For -9994J; I
3.3 V
= 13 A max.; I
2.5 V
= 13 A max.; I
3.3 V
+ I
2.5 V
² 16 A
For -9995J; I
3.3 V
= 13 A max.; I
12 V
= 6.5 A max.; I
3.3 V
+ I
12 V
² 16 A.
For -9908J; I
5.0 V
= 13 A max.; I
12 V
= 6.5 A max.; I
5.0 V
+ I
12 V
² 16 A.
Single output units: maximum continuous output power not to exceed; 75 W
on -9905J; 76.8 W on -9912J,-9924J, and -9917J.
150 LFM forced air cooling from L4 side.
Multiple output units: maximum continuous output power not to exceed 105
W.
For -9993J; I
3.3 V
= 16 A max.; I
5.0 V
= 16 A max.; I
3.3 V
+ I
5.0 V
² 20 A.
For -9994J; I
3.3 V
= 16 A max.; I
2.5 V
= 16 A max.; I
3.3 V
+ I
2.5 V
² 20 A
For -9995J; I
3.3 V
= 16 A max.; I
12 V
= 8.5 A max.; I
3.3 V
+ I
12 V
² 20 A.
For -9908J; I
5.0 V
= 16 A max.; I
12 V
= 8.5 A max.; I
5.0 V
+ I
12 V
² 20 A.
Single output units: maximum continuous output power not to exceed 110 W
for all models.
300 LFM forced air cooling from L4 side.
Multiple output units: maximum continuous output power not to exceed 110
W.
For -9993J; I
3.3 V
= 20 A max.; I
5.0 V
= 18 A max.; I
3.3 V
+ I
5.0 V
² 22 A.
For -9994J; I
3.3 V
= 20 A max.; I
2.5 V
= 20 A max.; I
3.3 V
+ I
2.5 V
² 22 A
For -9995J; I
3.3 V
= 20 A max.; I
12 V
= 8.5 A max.; I
3.3 V
+ I
12 V
² 22 A.
For -9908J; I
5.0 V
= 20 A max.; I
12 V
= 8.5 A max.; I
5.0 V
+ I
12 V
² 22 A.
for all models.
Peak output current lasting less than 30 seconds with duty cycle less than
5%. During peak loading, output voltage may exceed total regulation limits.
Figure is peak-to-peak for convection power rating. Output noise
measurements are made across a 20 MHz bandwidth using a 6’ twisted pair,
terminated with a 10 µF electrolytic capacitor and a 0.1 µF ceramic capacitor.
Minimum load required for correct start-up and operation on single outputs
and on main output of multiple versions. Failure to observe minimum
load on main output will not allow the supply to start-up correctly. Some
electronic test loads have a large delay time before they start drawing
current even though the voltage from the supply is present. During this time
delay, there is no load on the output and as a result, the supply cannot start-
up properly and maintain its correct output voltage. In these instances, a
dummy resistive load across the output may be necessary to load the output
of the supply until the test load can function correctly and draw the intended
minimum load. Minimum load required on auxiliary outputs to maintain
regulation.
For models NLP110-9908J and NLP110-9995J, the 12 V output is floating.
For -12 V output, pin 11 on J2 has to be connected to Return making pin 12
the -12 V output
Three orthogonal axes, random vibration 10 minutes for each axes, 2.4 G rms
5 Hz to 500 Hz.
For optimum reliability, no part of the heatsink should exceed 110 ˚C, and no
semiconductor case temperature should exceed 120 ˚C.
CAUTION: Allow a minimum of 1 second after disconnecting line power
when making thermal measurements.
This product is only for inclusion by professional installers within other
equipment and must not be operated as a stand alone product.
The EMI specifications reference measurements made with the power supply
mounted on a grounded metal sheet extending 1 inch beyond each edge,
using an unshielded cable. No external filtering required during conducted
emissions testing but some applications may require additional filtering
to achieve system compliance. A line choke, (ac input cords looped twice
through an EMI suppression toroid) was used during radiated emissions
testing. Considerable radiated testing in 1U six-sided boxes has shown that
units can meet level B in typical systems. Application support is available
from the factory to assist with EMI compliance.
All models require a minimum mounting stand-off of 6.35 mm (0.25 inches)
in the end use product.
The ‘J’ suffix indicates that these parts are Pb-free (RoHS 6/6) compliant.
NOTICE: Some models do not support all options. Please contact your local
Emerson Network Power representative or use the on-line model number
search tool at http://www.Emerson.com/EmbeddedPower
Embedded Power for
6.50
(165.10)
6.050
(153.670)
0.23
(5.842)
(64.770)
3.00
(76.20)
Americas
Rev. 10.06.11_155
4 of 4
J4
J2
F1
J1
ALL DIMENSIONS IN INCHES (mm)
MAXIMUM COMPONENT HEIGHT
IS 1.26 (32.00)
Recommended direction
for forced air relative to
power supply orientation
shown below.
Optimum.
Very good.
Good.
Not recommended.
5810 Van Allen Way
Carlsbad, CA 92008
USA
Telephone: +1 760 930 4600
Facsimile: +1 760 930 0698
Europe (UK)
T1 (REF.)
J3
J5
Waterfront Business Park
Merry Hill, Dudley
West Midlands, DY5 1LX
United Kingdom
Telephone: +44 (0) 1384 842 211
Facsimile: +44 (0) 1384 843 355
Asia (HK)
0.23
(5.842)
14/F, Lu Plaza
2 Wing Yip Street
Kwun Tong, Kowloon
Hong Kong
Telephone: +852 2176 3333
Facsimile: +852 2176 3888
For global contact, visit:
techsupport.embeddedpower
@emerson.com
Input Pin Connections
J1
Pin 1
Pin 2
Pin 3
AC Neutral
No Connection
AC Line
V (A) Sense +
V (A) Sense -
Safety Earth
Output Pin Connections
J2
Pin 1
Pin 2
Pin 3
Pin 4
Pin 5
Pin 6
Pin 7
Pin 8
Pin 9
Pin 10
Pin 11
Pin 12
Single
No Connection
No Connection
No Connection
Return
Return
Return
Return
V (A)
V (A)
V (A)
No Connection
No Connection
Triple
V (B)
V (B)
V (B)
Return
Return
Return
Return
V (A)
V (A)
V (A)
V (C)
V (C) Return
(7
While every precaution has been taken to ensure
accuracy and completeness in this literature, Emerson
Network Power assumes no responsibility, and disclaims
all liability for damages resulting from use of this
information or for any errors or omissions.
J3
Pin 1
Pin 2
Emerson Network Power.
The global leader in enabling
business-critical continuity.
AC Power
Connectivity
DC Power
Embedded Computing
Embedded Power
Monitoring
Outside Plant
Power Switching & Controls
Precision Cooling
Racks & Integrated Cabinets
Services
Surge Protection
EmersonNetworkPower.com
Emerson Network Power and the Emerson
Network Power logo are trademarks and
service marks of Emerson Electric Co.
©2011 Emerson Electric Co.
J4
Pin 1
J5
Pin 1
Pin 2
V (B) Sense +
V (B) Sense -
Input and output connectors
AC (J1) connector type
Molex 26-60-4030 or equivalent.
DC (J2) connector type
12 position Molex Spox type 26-48
1125 or equivalent.
Sense (J3) connector type
Molex 22-23-2021 or equivalent.
Earth (J4) connector type
Male 0.250 quick disconnect.
Sense (J5) connector type
Leoco 2421P02H000.
Mating connectors
AC (J1) mating connector type
Molex 09-50-3031 or equivalent with Molex 08-50-0105 or
equivalent crimp terminals.
DC (J2) mating connector type
Molex Spox type 26-03-3121 and contact 08-52-0113.
Sense (J3) mating connector type
Molex 22-01-3027 and contact 08-50-0113.
Earth (J4) mating connector type
Molex 90028.
Sense (J5) mating connector type