Regulated
Converter
2MOPP, 250VAC working voltage isolation
Clearance and creepage distance >8mm
Up to 10kVDC reinforced insulation
IEC/EN/UL 60601 certified with CB Report
(3rd Ed. Safety, 4th Ed. EMC)
-40°C to +75°C operation, no derating
2:1 wide input range
DC/DC Converter
REM6E
6 Watt
2:1 Input
DIP24 or SMD
Single & Dual
Output
w
YEA R
The REM6E series of medical grade regulated DC/DC converters feature reinforced 250VAC
continuous working isolation with >8mm creepage/clearance. The compact DIP24/SMD package offers
industry standard pinouts with tightly regulated single/dual outputs and UVLO, SCP and OCP.
The operating ambient temperature range is from -40°C to +75°C without derating. The converters
are UL marked and certified to CB, IEC, EN and ANSI/AAMI 60601 3rd. Ed. Safety and 4th Ed. EMC
medical standards. The low 1μA leakage current complies with medical applied part B, BF and CF limits
as defined by IEC60601-1.
arrant
Selection Guide
Part
Number
REM6E-xx09S
(3,4,5)
REM6E-xx12S
(3,4,5)
REM6E-xx15S
(3,4,5)
REM6E-xx24S
(3,4,5)
REM6E-xx09D
(3,4,5)
REM6E-xx12D
(3,4,5)
REM6E-xx15D
(3,4,5)
Input
Voltage Range
[VDC]
9-18 / 18-36 / 36-75
Output
[VDC]
±9
±12
±15
Output
Current
[mA]
667
±335
±250
±200
Efficiency
typ.
81 / 82 / 83
82 / 83 / 84
83 / 84 / 84
83 / 84 / 85
81 / 82 / 83
82 / 83 / 84
83 / 84 / 84
Max. Capacitive
Load
(2)
[µF]
2200
2200
2200
±2200
±2200
±2200
Notes:
Note1: Efficiency is tested at nominal input and full load at +25°C ambient
Note2: Max Cap Load is tested at nominal input and full resisitive load
RE
RE
CONVER
CONVER
PL
U
IA
PRO
PL
U
IA
PRO
Model Numbering
compliant
TS
REACH
compliant
CO
TS
10 from 10
REM6E-
Output Power
nom. Input Voltage
Output Voltage
Single
or
Dual
/R /A /
UVLO Option
(5)
Case Style or CTRL Option
(4)
Pinning
Isolation
(3)
2MOPP
250VAC
E314885
CAN/CSA-C22.2 No. 60601-1:14 certified
ANSI/AAMI ES60601-1 certified
EN60601-1 certified
IEC60601-1 certified
IEC60601-1-2 compliant
EN55032 compliant
Notes:
Note3: add suffix „/R8“ for 8kVDC or „/R10“ for 10kVDC isolation (DIP24 only)
if SMD package is used, always add suffix „/R6“ for 6kVDC isolation
Note4: add suffix „/CTRL“ for fitted CTRL pin (DIP24 only)
if SMD package is used do not add suffix „/CTRL“, CTRL pin is always mounted
Note5: add suffix „/X1“ for Under Voltage Lockout Option
Ordering Examples
REM6E-1209S/R8/A
REM6E-1212D/R10/A/CTRL
REM6E-1209S/R6/A/SMD
REM6E-2412D/R10/A/CTRL/X1
= 12Vin, 9Vout, Single, 8kVDC Isolation and „A“ pinning, DIP24
= 12Vin, 12Vout, Dual, 10kVDC Isolation, „A“ pinning, with CTRL pin
= 12Vin, 9Vout, Single, 6kVDC Isolation, „A“ pinning, SMD with CTRL pin
= 24Vin, 12Vout, Dual, 10kVDC Isolation, „A“ pinning, DIP24, CTRL pin and UVLO Option
www.recom-power.com
REV.: 3/2020
MED-1
y
CO
REM6E
DC/DC Converter
(measured @ Ta= 25°C, nom. Vin, full load and after warm-up unless otherwise stated)
BASIC CHARACTERISTICS
Internal Input Filter
Input Voltage Range
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
DC-DC ON
nom. Vin= 12VDC
DC-DC OFF
nom. Vin= 24VDC
nom. Vin= 48VDC
Input Current
DC-DC ON
DC-DC OFF
9VDC
18VDC
36VDC
12VDC
24VDC
48VDC
7.9VDC
18VDC
16.7VDC
36VDC
34.3VDC
650mA
320mA
150mA
35mA
25mA
7mA
10%
0.6ms
0.45ms
0.6ms
Open or 0VDC<V
CTRL
<1.2VDC
Short or 4.8VDC<V
CTRL
<12VDC
25mA
350µA
120kHz
20MHz BW
Notes:
Note6: Measurements are made with a 0.1µF MLCC across output. (low ESR)
150mVp-p
Condition
Min.
Typ.
Pi-type
18VDC
36VDC
75VDC
9VDC
Under Voltage Lockout (UVLO)
(„/X1“ version)
Quiescent Current
Minimum Load
(7)
Start-up time
Rise time
Hold-up time
ON/OFF CTRL
Input Current of CTRL Pin
Standby Current
Internal Operating Frequency
Output Ripple and Noise
(6)
DC-DC ON
DC-DC OFF
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
nom. Vin = 12VDC
nom. Vin = 24VDC
nom. Vin = 48VDC
DC-DC ON
DC-DC OFF
CTRL
=5VDC
DC-DC OFF
ON/OFF CTRL Option
+V
in
R1
CTRL
+V
out
R1
470Ω
-V
out
-V
in
continued on next page
www.recom-power.com
REV.: 3/2020
MED-2
REM6E
DC/DC Converter
Efficiency vs. Output Load
REM6E-1212S
REM6E-2412S
Efficiency [%]
9VDC
12VDC
15VDC
18VDC
Efficiency [%]
18VDC
24VDC
30VDC
36VDC
Output Load [%]
Output Load [%]
Efficiency vs. Input Voltage
(@ full Load)
REM6E-1212S
REM6E-2412S
Efficiency [%]
Efficiency [%]
24
26
28
34
36
Input Voltage [VDC]
Input Voltage [VDC]
REGULATIONS
Output Accuracy
Line Regulation
Load Regulation
(7)
Cross Regulation
Transient Response
low line to high line, full load
10% to 100% load
dual output only
25% load step change
Notes:
Note7: Operation below 10% load will not harm the converter, but specifications may not be met
www.recom-power.com
REV.: 3/2020
MED-3
Condition
Value
±1.5% typ.
±0.3% max.
0.5% typ.
±5.0% max.
5ms
REM6E
DC/DC Converter
Deviation vs. Load
REM6E-2412S
REM6E-1212S
Deviation [%]
Deviation [%]
-1.0
-1.5
-1.0
-1.5
Output Load [%]
Output Load [%]
PROTECTIONS
Short Circuit Protection (SCP)
below 100mΩ
“/R8” suffix
DIP24
“/R10” suffix
SMD
“/R6” suffix
tested for 1 second
rated for 1 minute
tested for 1 second
rated for 1 minute
rated for 1 minute
Value
continuous, hiccup mode, automatic recovery
8kVDC
4kVAC/60Hz
10kVDC
5kVAC/60Hz
6kVDC
Isolation Voltage
(8)
I/P to O/P
Isolation Resistance
Isolation Capacitance
Insulation Grade
Leakage Current
Means of Protection
Medical Device Classification
Internal
External
clearance/creepage
clearance/creepage
Notes:
Note8: For repeat Hi-Pot testing, reduce the time and/or the test voltage
Note9: Refer to local safety regulations if input over-current protection is also required. Recommended fuse: slow blow type
10GΩ min.
20pF typ.
reinforced
0.8µA typ. / 1µA max.
2MOPP
built-in power supply
>8mm
>8mm
ENVIRONMENTAL
Maximum Case Temperature
Temperature Coefficient
Thermal Impedance
Operating Altitude
Operating Humidity
Pollution Degree
MTBF
according to MIL-HDBK-217F, G.B.
www.recom-power.com
REV.: 3/2020
MED-4
Condition
full load @ natural convection 0.1m/s (see graph)
Value
-40°C to +75°C
+105°C
±0.02%/K typ. / ±0.05%/K max.
0.1m/s, horizontal
non-condensing
+25°C
+75°C
20K/W
3000m
5% - 95% RH max.
PD2
2100 x 10
hours
620 x 10
hours
REM6E
DC/DC Converter
Derating Graph
(@ Chamber and natural convection 0.1m/s)
Output Load [%]
-40 -30 -20 -10 0 10 20 30 40 50 60 70 80 90 100 110
75 85 95
Ambient Temperature [°C]
SAFETY AND CERTIFICATIONS
Certificate Type (Safety)
Medical Electric Equipment, General Requirements for Safety and
Essential Performance
Essential Performance (CB Scheme)
RoHS 2
Report / File Number
E314885
E314885
WD-SE-R-180524-A0
Standard
CAN/CSA-C22.2 No. 60601-1:14, 3rd Edition: 2014
ANSI/AAMI ES60601-1:2012
IEC60601-1:2005, 3rd Edition + AM1:2012
EN60601-1:2006 + A12:2014
RoHS 2011/65/EU + AM2015/863
EMC Compliance
Information technology equipment - Radio disturbance
characteristics - Limits and methods of measurement
Information technology equipment - Immunity characteristics - Limits
ESD Electrostatic discharge immunity test
Radiated, radio-frequency, electromagnetic field immunity test
Fast Transient and Burst Immunity
Surge Immunity
Immunity to conducted disturbances, induced by radio-frequency fields
Power Magnetic Field Immunity
Medical electrical equipment Part 1-2: Electromagnetic disturbances
– Requirements and tests
Industrial, scientific and medical equipment – Radio frequency distur-
bance characteristics – Limits and methods of measurement
Surge Immunity
Condition
with external filter
refer to
„EMC Filtering“
Standard / Criterion
EN55032, Class A and B
EN55024:2010 + A1:2015
Air ±8kV, Contact ±4kV
3V/m
DC Power Port: ±1kV
DC Power (Output) Port: ±0.5kV
DC Power (Output) Port: 3V
50Hz, 1A/m
IEC61000-4-2:2009, Criteria A
IEC61000-4-3:2006 + A2:2010, Criteria A
IEC61000-4-4:2012, Criteria A
IEC61000-4-5:2014 + A1:2017, Criteria A
IEC61000-4-6:2013 + C1:2015, Criteria A
IEC61000-4-8:2010, Criteria A
EN60601-1-2:2015
IEC60601-1-2:2014
EN55011:2016+A1:2017, Class B
Air ±15kV, Contact ±8kV
10V/m
DC Power Port: ±2kV
DC Power (Output) Port: ±1kV
DC Power (Output) Port: 3V, 6V
50Hz, 30A/m
IEC61000-4-2:2008, EN61000-4-2:2009
IEC61000-4-3:2006+A1:2007+A2:2010
EN61000-4-3:202006+A2:2010
IEC/EN61000-4-4:2012
IEC/EN61000-4-5:2014+A1:2017
IEC61000-4-6:2013, EN61000-4-6:2014
IEC61000-4-8:2009, EN61000-4-8:2010
www.recom-power.com
REV.: 3/2020
MED-5