Pressure Sensors
Low-cost, Stainless Steel, Isolated Sensors
Rugged, isolated stainless
steel package
Reliable semiconductor
technology
Calibrated and
temperature compensated
Voltage or current supply
options
Absolute and sealed gage
pressures
Low-cost cell package
Absolute pressure
0 psi to 500 psi, 1000 psi,
2000 psi, 3000 psi, and
5000 psi
TYPICAL APPLICATIONS
Industrial controls
Hydraulic controls
Tank pressure
Pressure transmitters
Process control systems
13 mm Series
Honeywell’s 13 mm Series stainless steel pressure sensors are designed
for high pressure applications that involve measurement of hostile media in
harsh environments. This series uses Honeywell’s proven piezoresistive
semiconductor sensor chip in an oil-isolated housing with or without an
integral ceramic for temperature compensation and calibration. This design
has proven to be highly reliable, stable and accurate.
These sensors feature a weld-ring collar and special back support ring for
high cycle life capability as they are designed for further package integration
in OEM (Original Equipment Manufacturer) applications. Products are
available with pressure ranges from 500 psi through 5000 psi and can be
used with voltage or current supplies.
EQUIVALENT CIRCUIT
13 mm compensated series
13 mm uncompensated series
WARNING
PERSONAL INJURY
DO NOT USE these
products as safety or
emergency stop devices or
in any other application
where failure of the product
could result in personal
injury.
Failure to comply with
these instructions could
result in death or serious
injury.
WARNING
MISUSE OF DOCUMENTATION
The information presented in this product sheet is for reference only. Do
not use this document as a product installation guide.
Complete installation, operation, and maintenance information is provided
in the instructions supplied with each product.
Failure to comply with these instructions could result in death or
serious injury.
Sensing and Control
Pressure Sensors
PRESSURE SENSOR CHARACTERISTICS
(all devices)
Environmental specifications
Compensated temperature range
Operating temperature range
Storage temperature range
Vibration
Shock
Life
Insulation resistance
Maximum ratings
Voltage version “K”
Voltage version “L”
0 °C to 82 °C [32 °F to 180 °F]
-40 °C to 125 °C [-40 °F to 257 °F]
10 G at 20 Hz to 2000 Hz
100 G for 11 ms
1 million cycles min.
100 mΩ at 50 Vdc
Supply voltage V
= +15 Vdc
Supply current I
= +2.0 mA
13 mm Series
PERFORMANCE CHARACTERISTICS – COMPENSATED SERIES
Characteristic
Zero pressure offset
(3)
Pressure non-linearity
(3)
Pressure hysteresis
Repeatability
(4)
Temp. effect on span
(4)
Temp. effect on offset
Thermal hysteresis (0 °C to 82 °C [32 °F to 180 °F])
(5)
Long-term stability of offset and span
(6)
Response time
(9)
Common mode voltage (voltage version “K”)
Input resistance (current version “L”)
Input resistance (voltage version “K”)
Output resistance
Minimum
-2
0.50
2.0
8.0
3.0
Typical
±0.1
±0.015
±0.010
±0.5
±0.5
±0.1
±0.1
0.1
1.25
4.5
Maximum
+2
±0.25
±0.030
±0.030
±1.0
±1.0
±0.3
±0.3
2.0
8.0
Units
mV
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
ms
Vdc
kΩ
kΩ
kΩ
PRESSURE RANGE SPECIFICATIONS – COMPENSATED SERIES
Catalog Listing
13C 0500P (A,S) (1,4,5,6) (K,L)
13C 1000P (A,S) (1,4,5,6) (K,L)
13C 2000P (A,S) (1,4,5,6) (K,L)
13C 3000P (A,S) (1,4,5,6) (K,L)
13C 5000P (A,S) (1,4,5,6) (K,L)
Pressure range
0 psi to 500 psi
0 psi to 1000 psi
0 psi to 2000 psi
0 psi to 3000 psi
0 psi to 5000 psi
Full scale span
98 mV to 102 mV
98 mV to 102 mV
98 mV to 102 mV
98 mV to 102 mV
148 mV to 152 mV
(2)
Proof pressure
1500 psi
3000 psi
6000 psi
9000 psi
10,000 psi
(7)
Burst pressure
2500 psi
5000 psi
10,000 psi
10,000 psi
10,000 psi
(8)
NOTES
1. Reference conditions (unless otherwise noted): T
= 25 °C [77 °F]; Supply V
= 10 Vdc ±0.01 Vdc or I
= 1.5 mA
±0.0015 mA.
2. Full-scale span (FSS) is the algebraic difference between the output voltage at full-scale pressure and the output at zero
pressure. FSS is ratiometric to supply voltage.
3. Pressure non-linearity is based on best-fit straight line from zero to the full-scale pressure. Pressure hysteresis is the
maximum output difference at any point within the operating pressure range for increasing and decreasing pressure.
4. Maximum error band of the offset voltage or span over the compensated temperature range, relative to the 25 °C [77 °F]
reading.
5. Long-term stability over a six-month period.
6. Response time for a 0 psi to FSS pressure step change, 10 % to 90 % rise time.
7. The maximum pressure that can be applied without changing the transducer’s performance or accuracy.
8. The maximum pressure that can be applied to a transducer without rupture of either the sensing element or transducer
case.
9. Common mode voltage as measure from output to ground.
2 Honeywell
Sensing and Control
Pressure Sensors
PERFORMANCE CHARACTERISTICS – UNCOMPENSATED SERIES
Characteristic
(3)
(3)
Pressure hysteresis
Repeatability
(4)
Temp. coefficient of span
°F])
(4)
Temp. coefficient of resistance
°F])
(4)
Temp. coefficient of offset
°F])
Thermal hysteresis
°F])
(5)
(6)
Response time
Input resistance
Output resistance
Minimum
-7.5
360
2700
4.0
4.0
Typical
±0.1
±0.015
±0.010
720
3420
±0.1
±0.1
0.1
4.75
4.75
13 mm Series
Maximum
+7.5
±0.25
±0.030
±0.030
1260
4500
±0.3
±0.3
6.0
6.0
mV/V
%FSS
%FSS
%FSS
ppm/°C
ppm/°C
µV/V/°C
%FSS
%FSS
ms
kΩ
kΩ
PRESSURE RANGE SPECIFICATIONS – UNCOMPENSATED SERIES
Catalog Listing
13U 0500P A 0 K
13U 1000P A 0 K
13U 2000P A 0 K
13U 3000P A 0 K
13U 5000P A 0 K
Pressure range
0 psi to 500 psi
0 psi to 1000 psi
0 psi to 2000 psi
0 psi to 3000 psi
0 psi to 5000 psi
Full scale span
175 mV to 300 mV
175 mV to 300 mV
175 mV to 300 mV
175 mV to 300 mV
290 mV to 500 mV
(2)
Proof pressure
1200 psi
3000 psi
6000 psi
9000 psi
10,000 psi
(7)
Burst pressure
2400 psi
5000 psi
10,000 psi
10,000 psi
10,000 psi
(8)
NOTES
= 1.5 mA
±0.0015 mA.
4. The error band resulting from maximum deviation of a transducer’s output parameter (offset, span, or resistance) as
temperature is varied from 25 °C [77 °F] to any other temperature within the specified range 0 °C to 82 °C [32 °F to
180 °F]. This parameter is not 100 % tested and is guaranteed by process design and tested on a sample basis only.
Temperature coefficient of span is evaluated using a constant current source.
case.
Honeywell
Sensing and Control
Pressure Sensors
PHYSICAL DIMENSIONS
for reference only, in (mm)
Package 1 (see note): Ring with back support
13 mm Series
Non-concentricity effects at the diaphragm weld area may cause runout of up to ±0.006 in between the upper and
lower portions of the sensor body. (It is recommended to use a counter bore to mate with this device to allow for this non-
concentricity).
Package 4: 1/8-27 NPT
4 Honeywell
Sensing and Control
Pressure Sensors
Package 5: 1/4-18 NPT
13 mm Series
Package 6: 7/16-20 UNF
Honeywell
Sensing and Control