MINT1500
500W Single Output
Medical & ITE Series
3.3” x 7” x 1.5” Package, Ideal for 1U
Up to 500W w/air, 350W convection cooled
Universal Input 90 to 270Vac
Active current share
Class I Input
Standby and Fan Output Voltages
Inhibit, Power Fail, Output OK Signals
Approved to IEC60601-1, 3
rd
Edition
& EN60950 2
nd
Edition
2 x MOPP Isolation
93% Efficiency @ 230V Input, 92% @ 115V
Input
Optional Cover
3 mounting orientations
3 Year Warranty
Low inrush Current
A superior performance 500 Watt AC to DC power supply designed for Medical or Industrial applications. Feature rich and
highly efficient, MINT1500 product family with active current share for redundant applications can easily fit in 1U or 2U
chassis and provides 350 Watts without cover for convection, or 500 Watts with moving air. Input & output monitoring
alarms plus 12V/0.8A fan output and 5V standby voltage are standard features of the MINT1500 family. All models are CE
rd
nd
marked to low voltage directive and approved to IEC60601-1 3 edition, EN60950 2 edition.
Model Selection
Model
Number
MINT1500A2414E01
MINT1500A4814E01
MINT1500A5614E01
Notes:
Volts
24V
48V
56V
Output Current
w/200LFM air Convection
20.8A
10.4A
8.9A
14.6A
7.3A
6.3A
Fan
Output
12Vdc/0.8A
12Vdc/0.8A
12Vdc/0.8A
5V Standby
Output
5V@200mA
5V@200mA
5V@200mA
Ripple &
Noise¹
1%
1%
1%
Total
Regulation
±2%
±2%
±2%
OVP
Threshold
27.6 ± 1.0V
55.5 ± 2.0V
64.3 ± 2.0V
1) Measured with noise probe directly across output terminals with 0.1µF ceramic and 10µF low ESR capacitors.
For main output load of less than 5%, total noise & ripple will increase to 2%.
www.slpower.com
MINT1500 REV 6.0, 12-18-14
MINT1500
500W Single Output
General Specifications
90-270Vac, 47-63Hz agency approved from
90Vac to 264Vac
120–300Vdc
(External fuse required for DC input)
Power supply is protected against brown out
condition
115Vac: 5A, 230Vac: 2.5A
AC Input
Turn On Time
<500mS @115Vac
Input Current
Hold-up Time
Power Factor
Correction
Overload
Protection
Main Output
Short Circuit
Protection
Overvoltage
Protection
Overtemperature
Protection
Switching
Isolation
Operating
Temperature
Storage
Temperature
Altitude
Relative
Humidity
Shock
20 mS minimum at full load & 100 Vac input
Inrush Current
270Vac, cold start: will not exceed 15A
Meets EN61000-3-2, Class A, and D
Meets Class C for 500 Watt output
Input Fuses
Earth Leakage
Current
Efficiency
F1, F2: T10A, 250Vac, provided on all models
120% to 140% of current rating, Hiccup Mode
<275µA@264Vac, 60Hz, NC; <450µA SFC
Self-recovering
92% typical at 115Vac, 93% typical at 230Vac
500W continuous with 200 lfm airflow, up to
350W convection cooled @ 100Vac, 50°C
ambient.
500 uS typical for return to within 0.5% of
nominal, 50% load step from 5% to 100%.
∆i/∆t:
<0.2A/µS.
Maximum Voltage Deviation = 3%
See models chart
Hiccup Mode, Self-recovering see models chart
for trip ranges.
Provided, self-recovering
Automatic power shutdown at TC 135°
C/115°
PFC: Variable, 50-500kHz
Main Converter: Variable 40-180 kHz, 65Khz
typical
Input-Output: 4000Vac, 2 x MOPP
Input-Ground: 1800Vac, 1 x MOPP
Output-Ground: 700Vdc
-10° to +70°
(Derate output power linearly by 2.5% /˚C
above 50° to 50% at 70°
C)
Output Power
Transient
Response
Ripple and Noise
Main Output
Output Voltage
Adjustability
Minimum Load
See models chart
+/-5% from nominal
-40° to +85°
Operating: Up to 3000m (derate 5˚C for natural
Convection Cooling Applications)
Non-operating: -150 to 12,000m
5% to 95%, non-condensing
Operating: Half-sine, 40gpk, 8ms, 3 axes, 6
shocks total
Not required
Main Output: ±2%.
12V Fan Output: ±10%
(with >0.1A load on main o/p)
5Vsb Output: ±2%
Operating: 0.003g²/Hz, 1.5grms overall, 3 axes,
1 hr/axis
Operating: 0.026g²/Hz, 5.0grms overall, 3 axes,
10 min/axis
W: 3.3” x L: 7.0” x H: 1.5”
W: 84mm x L: 178mm x H: 38mm
With Fan Option: Top Mount: 3.3” x 7.0” x 2.09”
With IEC: 3.3” x 8.15” x 1.58”
Total Regulation
Vibration
Dimensions
Safety Standards
EN/CSA/UL/IEC 60601-1, 3
rd
Edition
Weight
680g. 750g w/cover
MTBF
Over 500,000 hours, 25°C Ambient, 115Vac
input
Single wire current share for up to 3 units.
Please consult Application notes for proper
wiring
Cover Option
Replace Letter “A” to “C” for this option
Example: MINT1500C2414E01
Current Share
www.slpower.com
MINT1500
500W Single Output
Auxiliary Signals
Power Good:
Signal goes HIGH 100-250mS after main
output is in regulation, and goes LOW with
7ms warning time before loss of main output
due to loss of AC input (Output is measured
above 90% nominal voltage)
Logic High or Open = On
Low/ground
= Off
Low or Open = ON
Logic High = OFF
DC OK:
During normal operation, this signal is logic
HIGH. Signal will go LOW for output less than
90% of nominal.
Inhibit:
PS Off:
Current Sharing
Active single wire, for up to 3 units in parallel
EMI/EMC Compliance
Conducted Emissions
Radiated Emissions
Static Discharge Immunity
Radiated RF Immunity
EFT/Burst Immunity
Line Surge Immunity
Conducted RF Immunity
Power Frequency Magnetic Field Immunity
Voltage Dip Immunity
Line Harmonic Emissions
Flicker Test
EN55011/22 Class B, FCC Part 15, Subpart B, Class B
EN55011/22 Class A; FCC Part 15, Subpart A, Class A
EN61000-4-2, Criteria A, 6kV Contact Discharge, 8kV air discharge
EN61000-4-3, 3V/m;10V/M Criteria A, B
EN61000-4-4, 2kV/5kHz
EN61000-4-5, 1kV differential, 2kV common-mode
EN61000-4-6, 3Vrms
EN61000-4-8, 3A/m
EN61000-4-11, 0% Vin, 10mS; 40% Vin, 100mS (60% load); 70% Vin,
500mS (80% load); 0% 5000mS; Criteria A, A, A, B.
EN61000-3-3, Complies
Isolation Specifications
Insulation Safety Rating
Conditions/Description
Input/Ground
Input/Output
Output/Ground
Input/Ground
Input/Output
Output/Ground
Min
Nom
1 MOPP
2 MOPP
Operational
Units
Electric Strength Test Voltage
1800
4000
700
Vac
Vac
Vdc
www.slpower.com
MINT1500
500W Single Output
Mechanical Drawings
www.slpower.com
MINT1500
500W Single Output
Connector Information
Input Connector
J1 (E Version))
PIN 1) GROUND
PIN 3) AC NEUTRAL
PIN 5) AC LINE
Mating Connector:
AMP 770849-5 or 647402-5
Pins= 3-770522-1 or 3-647409-1
DC Output
Connector
(2) Buss Bars (M5
x 0.5 Screws):
1) +Vout
2) RTN
Mating Connector:
#M3 Spade or
Ring Lugs
Fan Output Connector
J301
PIN 1) +12Vfan
PIN 2) RTN
Signal ConnectorJ2
(see Detail B on prior page)
Mating Connector:
AMP 1375820-2, 3-640441-2, or
Molex 22-01-3027
Pins = AMP 1375819-1 or Molex 08-
50-0114
PIN 1) Remote Sense +
PIN 2) Remote Sense –
PIN 3) +5Vsb RTN
PIN 4) RTN
PIN 5) +5Vsb
Mating Connector:
Molex 90142-0010
Pins = 90119-2109 or 2120
PIN 6) Power_Good
PIN 7) Current Share
PIN 8) PS_Off
PIN 9) Enable
PIN 10) DC_OK
Fan Options
Consult Factory
Characteristic Curves
Output vs. Temperature
350W convection cooled and 500W continuous with 200 LFM airflow. Derate output power to 50% at 70°C.
Output Power (Watts)
Output Power vs. Ambient
Temperature
Ambient Operating Temperature (
C)
Convection
200LFM Cooling
www.slpower.com