1~4 Output Medical Type
200W Medical
series
Universal AC input / Full range
Low leakage current <180 A
Protections: Short circuit / Overload / Over voltage / Over temperature
Free air convection for 140W and forced air convection for 200W
UL60601-1 medical safety approved
With power good and fail signal output
Built-in remote ON-OFF control
Built-in remote sense function
Fixed switching frequency at 100KHz
3 years warranty
EN60601-1
IEC60601-1
Specification
AC INPUT VOLTAGE
90~264 VAC, 47~440Hz / 127~370VDC.
POWER FACTOR (Typ.)
PF>0.95/230VAC
PF>0.98/115VAC at full load
WORKING TEMP.
Whole series can operate from -20~70 . Please refer
to the derating curves.
WORKING HUMIDITY
20~90% RH non-condensing.
STORAGE TEMP., HUMIDITY
-40~+85 , 10~90% RH
TEMP. COEFFICIENT
0.04%/ on all outputs at full load between 0~50
of ambient temperature.
AC INPUT CURRENT
(Typ.)
Maximum input current 3.5A at 115VAC, 60Hz or
1.6A at 230VAC, 60Hz with 100% output load.
INRUSH CURRENT
(Typ.)
Inrush current is less than 25A at 115VAC or
less than 40A at 230VAC under cold start
conditions. Limiting provided by internal
thermistors.
SETUP, RISE TIME
1000ms, 20ms / 230VAC at full load
3000ms, 20ms / 115VAC at full load
HOLD-UP TIME
(Typ.)
16ms / 230VAC at full load
16ms / 115VAC at full load
LEAKAGE CURRENT
Leakage current is less than 180 A at 264VAC for
earth leakage current
Leakage current is less than 100 A at 264VAC for
patient leakage current
DC OUTPUT ADJ. RANGE
DC output voltage (or CH1 of multiple output models)
can be adjusted between -5%~+10% rated output
voltage by potential meter.
OVERLOAD PROTECTION
Fully protected against short circuit and output overload.
The hiccup type protection will be activated at 120~160%
rated load and recovers automatically after fault condition
is removed.
OVER VOLTAGE PROTECTION
Provided on output channel 1 only at 115%~135%
rated output voltage.
Output will be shut down when this protection is activated.
OVER TEMPERATURE PROTECTION
When the temperature of TSW1 which detect on heat sink of
power transistor reaches 95
This protection is activated.
Then output will be shut down and recovers automatically
after temperature goes down.
POWER GOOD / FAIL SIGNAL
TTL logic high for power good and TTL low for
power fail. When the output voltage reaches 90% of
rated value, a +5V TTL signal will be sent out with a
10~500ms delay; At least 1ms before the output
voltage goes below 90% of the rated value, the TTL
signal will be turned off.
* MPS-200-3.3 does not have this function.
REMOTE CONTROL
RC+/RC-:0 ~ 0.8V=power on; 4 ~ 10V=power off sink current<4~10mA
VIBRATION
2G of acceleration, vibrating frequency adjust from 10Hz ~500Hz within a
10-minute cycle. 6 testing cycles (60 minutes) each along X, Y, Z axes.
SAFETY STANDARDS
Medical
UL60601-1, TUV EN60601-1
, IEC60601-1 approved
Commercial : Also design refer to UL60950-1, TUV EN60950-1
WITHSTAND VOLTAGE
4000VAC between input and output
1500VAC between input and F.G.
1500VAC between output and F.G.
ISOLATION RESISTANCE
>100M Ohms for I/P-O/P, I/P-FG, O/P-FG by using 500VDC test voltage.
EMI COMPLIANCE
EMI Specifications
Conducted & Radiation
Harmonic distortion
Voltage flicker
EMS COMPLIANCE
EMS Specification
ESD air
ESD contact
RF field susceptibility
EFT(Electrical Fast Transient)/Burst
Lightning/Surge
Conducted RF susceptibility
Magnetic field immunity
Voltage dip, interruption
Digital phone carrier immunity
MTBF
262,100 hours min. at full load and 25
calculated per MIL-HDBK-217F.
DIMENSION (L*W*H)
177.8x107.2x35.5mm or 7"x4.22"x1.4"
PACKING
0.66Kg; 24pcs/16.8Kg/0.99CUFT
File Name:200W MEDICAL-SPEC 2010-10-18
Compliance Level
EN55011, Class B
EN55022, Class B
EN61000-3-2
EN61000-3-3
Compliance Level
EN61000-4-2, Level 3, 8KV
EN61000-4-2, Level 2, 4KV
EN61000-4-3, Level 2, 3V/m
Level 3, 10V/m
EN61000-4-4, Level 2, 1KV/5KHz
Level 3, 2KV/5KHz
EN61000-4-5, Level 4, 2KV/Line-Line
4KV/Line-Earth
EN61000-4-6, Level 2, 3Vrms/m
Level 3, 10Vrms/m
EN61000-4-8, Level 2, 3A/m
Level 3, 10A/m
EN61000-4-11,Compliance
ENV50204, Level 2, 3V/m, 900MHz
Level 3, 10A/m, 900MHz
of ambient temperature,