For user manuals and dimensional drawings, visit the product page resources tab on ni.com.
Last Revised: 2012-08-31 15:35:38.0
16-Channel Thermocouple Module
NI 9213
Built-in CJC (cold-junction compensation)
High-speed mode for up to 1,200 S/s (aggregate)
250 Vrms channel-to-earth ground safety isolation
24-bit ADC for up to 0.02 °C measurement sensitivity
Autozero channel for offset error compensation
J, K, T, E, N, B, R, and S types supported
Included backshell kit (NI 9940) for strain relief
The NI 9213 is a high-density thermocouple module for NI C Series carriers designed for higher-channel-count systems. With this module, you can add thermocouples to
mixed-signal test systems without taking up too many slots.
The NI 9213 is similar to the NI 9211 four-channel thermocouple module except it features four times the channel count and almost 100 times the sample rate.
You can use up to eight NI 9213 modules in an NI CompactDAQ chassis or CompactRIO chassis for 128 thermocouple measurements in a single chassis, or deploy a single
module in any of the USB, Ethernet, or Wi-Fi carriers for C Series modules.
Each shipping kit contains:
-NI 9213 module with spring-terminal connectivity
-NI 9940 backshell for cabling and strain relief
-Spring-terminal tool for signal wire insertion
Back to Top
Requirements and Compatibility
OS Information
Real-Time OS
Windows
Driver Information
NI-DAQmx
NI-RIO
Software Compatibility
LabVIEW
LabWindows/CVI
Measurement Studio
SignalExpress
Visual Basic
Visual C#
Visual C++
Visual Studio
Visual Studio .NET
Back to Top
Comparison Tables
1/8
www.ni.com
Thermocouple Module
NI 9213
NI 9219
NI 9211
Channels
Sample Rate
1,200 S/s
50 S/s/ch
14 S/s
Resolution
24-bit
24-bit
24-bit
Feature
Lowest cost/channel
Channel-to-channel isolation
Low-channel count
Back to Top
Application and Technology
Channel Density for High-Channel-Count Systems
Temperature measurements can vary greatly over the different components in a machine, the surface of a device, or the volume of a given space. By taking more temperature
measurements, you can better quantify the thermal gradients and achieve more accurate test results. The NI 9213 was designed with density in mind to enable a large number of
thermocouple channels in a compact space. With up to eight NI 9213 modules installed in an NI CompactRIO or NI CompactDAQ chassis, you can measure up to 128
thermocouples in a system that fits inside a 9 by 9 by 26 cm box. For even more portable temperature measurement systems, you can install the NI 9213 into one of the single
module carriers for NI C Series modules to make 16-channel thermocouple measurements over USB, Ethernet, or Wi-Fi (802.11b/g).
CJC and Autozero for a More Accurate Measurement
The NI 9213 has built-in cold-junction compensation (CJC) to eliminate error caused by the physical connection of the sensor to the instrumentation. The contact of two dissimilar
metals creating a voltage potential is the principle on which thermocouples are designed, and, as such, the contact of the dissimilar thermocouple metal with the spring terminal
must be removed from the calculation. This is done through a process known as cold-junction compensation, with the cold junction being the mating of the sensor to the
instrument. All NI instrumentation designed for thermocouple measurement contains CJC circuitry including the NI 9211, NI 9219, and NI 9213. In addition to CJC, the NI 9213
features an extra, internal-only channel known as the autozero channel. By measuring the autozero channel at the beginning of each channel scan, you can eliminate further
offset errors to provide a more accurate temperature measurement.
C Series Compatibility
The NI C Series hardware family features more than 50 measurement modules and several chassis and carriers for deployment. With this variety of modules, you can mix and
match measurements such as temperature, acceleration, flow, pressure, strain, acoustic, voltage, current, digital, and more to create a custom system. Install the modules in one
of several carriers to create a single module USB, Ethernet, or Wi-Fi system, or combine them in chassis such as NI CompactDAQ and CompactRIO to create a
mixed-measurement system with synchronized measurements. You can install up to eight modules in a simple, complete NI CompactDAQ USB data acquisition system to
synchronize all of the analog output, analog input, and digital I/O from the modules. For a system without a PC, CompactRIO holds up to eight modules and features a built-in
processor, RAM, and storage for an embedded data logger or control unit. For higher-speed control, CompactRIO chassis incorporate a field-programmable gate array (FPGA)
that you can program with NI LabVIEW software to achieve silicone-speed processing on I/O data from C Series modules.
Back to Top
Ordering Information
For a complete list of accessories, visit the product page on ni.com.
Products
NI 9213 Thermocouple Input Module
NI 9213
Requires: 1 Connectivity Accessories ;
Part Number Recommended Accessories
Connectivity Accessories:
screwTerminal - NI 9940 Strain relief, operator protection (qty 1)
779567-01
Back to Top
Software Recommendations
LabVIEW Professional Development System for Windows
Advanced software tools for large project development
Automatic code generation using DAQ Assistant and Instrument I/O Assistant
Tight integration with a wide range of hardware
Advanced measurement analysis and digital signal processing
Open connectivity with DLLs, ActiveX, and .NET objects
Capability to build DLLs, executables, and MSI installers
Back to Top
Support and Services
System Assurance Programs
NI system assurance programs are designed to make it even easier for you to own an NI system. These programs include configuration and deployment services for your NI PXI,
CompactRIO, or Compact FieldPoint system. The NI Basic System Assurance Program provides a simple integration test and ensures that your system is delivered completely
2/8
www.ni.com
assembled in one box. When you configure your system with the NI Standard System Assurance Program, you can select from available NI system driver sets and application
development environments to create customized, reorderable software configurations. Your system arrives fully assembled and tested in one box with your software preinstalled.
When you order your system with the standard program, you also receive system-specific documentation including a bill of materials, an integration test report, a recommended
maintenance plan, and frequently asked question documents. Finally, the standard program reduces the total cost of owning an NI system by providing three years of warranty
coverage and calibration service. Use the online product advisors at ni.com/advisor to find a system assurance program to meet your needs.
Calibration
NI measurement hardware is calibrated to ensure measurement accuracy and verify that the device meets its published specifications. To ensure the ongoing accuracy of your
measurement hardware, NI offers basic or detailed recalibration service that provides ongoing ISO 9001 audit compliance and confidence in your measurements. To learn more
about NI calibration services or to locate a qualified service center near you, contact your local sales office or visit ni.com/calibration.
Technical Support
Get answers to your technical questions using the following National Instruments resources.
Support
- Visit ni.com/support to access the NI KnowledgeBase, example programs, and tutorials or to contact our applications engineers who are located in NI sales
offices around the world and speak the local language.
Discussion Forums
- Visit forums.ni.com for a diverse set of discussion boards on topics you care about.
Online Community
- Visit community.ni.com to find, contribute, or collaborate on customer-contributed technical content with users like you.
Repair
While you may never need your hardware repaired, NI understands that unexpected events may lead to necessary repairs. NI offers repair services performed by highly trained
technicians who quickly return your device with the guarantee that it will perform to factory specifications. For more information, visit ni.com/repair.
Training and Certifications
The NI training and certification program delivers the fastest, most certain route to increased proficiency and productivity using NI software and hardware. Training builds the skills
to more efficiently develop robust, maintainable applications, while certification validates your knowledge and ability.
Classroom training in cities worldwide
- the most comprehensive hands-on training taught by engineers.
On-site training at your facility
- an excellent option to train multiple employees at the same time.
Online instructor-led training
- lower-cost, remote training if classroom or on-site courses are not possible.
Course kits
- lowest-cost, self-paced training that you can use as reference guides.
Training memberships
and training credits - to buy now and schedule training later.
Visit ni.com/training for more information.
Extended Warranty
NI offers options for extending the standard product warranty to meet the life-cycle requirements of your project. In addition, because NI understands that your requirements may
change, the extended warranty is flexible in length and easily renewed. For more information, visit ni.com/warranty.
OEM
NI offers design-in consulting and product integration assistance if you need NI products for OEM applications. For information about special pricing and services for OEM
customers, visit ni.com/oem.
Alliance
Our Professional Services Team is comprised of NI applications engineers, NI Consulting Services, and a worldwide National Instruments Alliance Partner program of more than
700 independent consultants and integrators. Services range from start-up assistance to turnkey system integration. Visit ni.com/alliance.
Back to Top
Detailed Specifications
The following specifications are typical for the range – 40 to 70 °C unless otherwise noted.
Works over temperature ranges defined by NIST (J, K, T, E, N, B, R, S
thermocouple types)
3/8
www.ni.com
Timing modes
Timing Mode
Conversion Time (Per Channel) Sample Rate
(All Channels
1 S/s
75 S/s
High-resolution 55 ms
High-speed
740
μ
s
Common-mode voltage range
Channel-to-COM
COM-to-earth ground
Common-mode rejection ratio
High-resolution mode (at DC and 50–60 Hz)
Channel-to-COM
COM-to-earth ground
High-speed mode (at 0–60 Hz)
Channel-to-COM
COM-to-earth ground
Input bandwidth
High-speed mode
High-resolution noise rejection (at 50 and 60 Hz)
Overvoltage protection
Differential input impedance
Input current
Input noise
High-speed mode
Gain error
High-speed mode
Offset error
High-speed mode
Offset error from source impedance
Cold-junction compensation accuracy
0 to 70 °C
– 40 to 70 °C
MTBF
0.8 °C typ, 1.7 °C max
1.1 °C typ, 2.1 °C max
852,407 hours at 25 °C; Bellcore Issue 2, Method 1, Case 3, Limited Part
Stress Method
Contact NI for Bellcore MTBF specifications at other temperatures or for MIL-HDBK-217F specifications.
4 μV typ, 6 μV max
14 μV typ, 17 μV max
Add 0.05 μV per Ω, when source impedance >50 Ω
0.03% typ at 25 °C, 0.07% typ at – 40 to 70 °C, 0.15% max at – 40 to 70 °C
0.04% typ at 25 °C, 0.08% typ at – 40 to 70 °C, 0.16% max at – 40 to 70 °C
200 nV
rms
7 μV
rms
14.4 Hz
78 Hz
60 dB
±30 V between any two inputs
78 MΩ
50 nA
70 dB
>150 dB
100 dB
>170 dB
±1.2 V min
±250 V
Temperature Measurement Accuracy
Measurement sensitivity
Types J, K, T, E, N
Types B, R, S
High-speed mode
Types J, K, T, E
Type N
<0.25 °C
<0.35 °C
<0.02 °C
<0.15 °C
4/8
www.ni.com
Type B
Types R, S
<1.2 °C
<2.8 °C
Figures 6, 7, 8, 9, and 10 show the errors for each thermocouple type when connected to the NI 9213 with the autozero channel on. The figures display the maximum errors over
a full temperature range and typical errors at room temperature. The figures account for gain errors, offset errors, differential and integral nonlinearity, quantization errors, noise
errors, 50
lead wire resistance, and cold-junction compensation errors. The figures do not account for the accuracy of the thermocouple itself.
随着阅读器与标签价格的降低和全球市场的扩大,射频标识 RFID(以下简称RFID)的应用与日俱增。标签既可由阅读器供电(无源标签),也可以由标签的板上电源供电(半有源标签和有源标签)。由于亚微型无源 CMOS 标签的成本降低,库存和其他应用迅速增加。一些评估表明,随着无源标签的价格持续下降,几乎每一个售出产品的内部都将有一个 RFID 标签。由于无源 RFID 标签的重要性及其独特的工程实现...[详细]