C Series Analog Input Modules
NI 9201, NI 921x, NI 9221, NI 923x
• Signal conditioning for high voltage
(±60 V), thermocouples, RTDs,
accelerometers, microphones,
strain gages, current inputs
• Advanced features such as smart TEDS
sensor capability, antialiasing filters,
open-thermocouple detection
• ±80 mV, ±10 V, or ±60 V analog
input ranges
• 12-, 16-, or 24-bit
(delta-sigma) resolution
• Up to 800 kS/s multiplexed or up to
100 kS/s simultaneous-sampling
analog-to-digital converter (ADC)
• Up to 32 channels per module
• Up to 2,300 V
rms
isolation (withstand),
up to 250 V
rms
isolation (continuous)
• NIST-traceable calibration certificate
for guaranteed accuracy
Model CompactRIO NI CompactDAQ
NI 9201
NI 9203
NI 9205
NI 9206
NI 9211
NI 9215
NI 9217
NI 9221
NI 9233
NI 9237
Signal Type
Current
CAT I Isolated
Thermocouple
RTD
IEPE
Bridge
Channels
32 SE/16 DI
16 DI
Resolution
(bits)
24
Max Sampling
Rate (S/s)
500 k
200 k
250 k
250 k
100 k/ch
800 k
50 k/ch
50 k/ch
Signal
Input Ranges
±10 V
±20 mA, 0 to 20 mA
±10, ±5, ±1, ±0.2 V
±10, ±5, ±1, ±0.2 V
±80 mV
±10 V
0 to 400
±60 V
±5 V
±250 mV
Simultaneous Antialiasing
Isolation Connector Options
Sampling
Filters
Screw Terminal, D-Sub
Screw Terminal
Spring Terminal, D-Sub
Spring Terminal
Screw Terminal
Screw Terminal, BNC
Screw Terminal
BNC
RJ50
Table 1. C Series Analog Input Modules Selection Guide
High-accuracy C Series analog input modules for National Instruments
CompactRIO and NI CompactDAQ provide high-performance measurements
for a wide variety of industrial, in-vehicle, and laboratory sensors and
signal types. Each module includes built-in signal conditioning and an
integrated connector with screw terminal or cable options for flexible
and low-cost signal wiring. All modules feature NI CompactRIO Extreme
Industrial Certifications and Ratings.
high-speed decision making for all C Series analog modules. For
instance, with CompactRIO, you can implement custom triggering for
any analog sensor type on a per-channel basis using the flexibility and
performance of the FPGA and the numerous arithmetic and comparision
function blocks built into NI LabVIEW FPGA.
Key Features
• High-accuracy, high-performance analog measurements for any
CompactRIO embedded system, R Series expansion chassis, or
NI CompactDAQ chassis
• Screw terminals, BNC, D-Sub, spring terminals, strain relief, high-
voltage, cable, solder cup backshell, and other connectivity options
• Available channel-to-earth ground double-isolation barrier for safety,
noise immunity, and high common-mode voltage range
• CompactRIO Extreme Industrial Certifications and Ratings
• Built-in signal conditioning for direct connection to sensors and
industrial devices
Visit
ni.com/compactrio
or
ni.com/compactdaq
for up-to-date
information on module availability, example programs, application notes,
and other developer tools.
System Compatibility
You can use NI C Series modules in multiple system types depending
on available software. See Table 1 for CompactRIO and NI CompactDAQ
module compatibility because not all modules work with both systems.
Many of the advanced features described apply only to reconfigurable
I/O systems and not to NI CompactDAQ.
Advanced Features
When used with CompactRIO, NI C Series analog input modules
connect directly to reconfigurable I/O (RIO) FPGA hardware to create
high-performance embedded systems. The reconfigurable FPGA
hardware within CompactRIO provides a variety of options for custom
timing, triggering, synchronization, filtering, signal processing, and
North American
Hazardous Locations
C Series Module Accessories
Connectivity Accessories
CompactRIO and NI CompactDAQ systems are designed to provide
flexible options for low-cost field wiring and cabling. Most C Series
modules have a unique connector block option that offers secure and
safe connections to your C Series system. Table 2 contains all of the
connector blocks available for C Series I/O modules.
Accessory
NI 9932
NI 9933
NI 9934
NI 9935
NI 9936
NI 9939
10-position strain relief and high-voltage screw-terminal connector kit
37-pin D-Sub connector kit with strain relief and D-Sub shell
25-pin D-Sub connector kit with strain relief and D-Sub shell
15-pin D-Sub connector kit with strain relief and D-Sub shell
10-position screw-terminal plugs (quantity 10)
16-position connector kit with strain relief
The NI 9933 includes a screw-terminal connector with strain relief
as well as a D-Sub solder cup backshell for creating custom cable
assemblies for any module with a 37-pin D-Sub connector.
To meet shock and vibration requirements, you must affix ferrules to the ends of the wires on all
screw-terminal connectors.
Figure 2. NI 9933 37-Pin D-Sub Connector Kit with Strain Relief and D-Sub Shell
Table 2. Connector Blocks for C Series I/O Modules
Table 3 lists the recommended connector block accessories for each
C Series analog input module.
NI 9201
NI 9201 with D-Sub
NI 9211
NI 9215
NI 9217
NI 9221
NI 9221 with D-Sub
Requires a 25-pin D-Sub connector such as the NI 9934 accessory kit.
The NI 9934 includes a screw-terminal connector with strain relief
assemblies for any module with a 25-pin D-Sub connector.
Recommended Module Accessory
NI 9932, NI 9936
NI 9934
NI 9932, NI 9936
NI 9932, NI 9936
NI 9939
NI 9932, NI 9936
NI 9934
Table 3. Recommended Connector Block Accessories
Figure 3. NI 9934 25-Pin D-Sub Connector Kit with Strain Relief and D-Sub Shell
The NI 9932 kit provides strain relief and operator protection from
high-voltage signals for any 10-position screw-terminal module.
The NI 9935 includes a screw-terminal connector with strain relief as
assemblies for any module with a 15-pin D-Sub connector.
Figure 1. NI 9932 10-Position Strain Relief and High-Voltage Screw-Terminal
Connector Kit
Figure 4. NI 9935 15-Pin D-Sub Connector Kit with Strain Relief and D-Sub Shell
BUY ONLINE
at
ni.com
or
CALL 800 813 3693 (U.S.)
The NI 9936 consists of 10-position screw-terminal plugs for any
Figure 5. NI 9936 10-Position Screw-Terminal Plugs
Visit
ni.com/compactrio
or
ni.com/compactdaq
for up-to-date
information on availability of accessories.
BUY ONLINE
at
ni.com
or
CompactRIO Ordering Information
Developer Kits
NI cRIO-9961 Embedded System Developer
Hardware Evaluation Kit ..................................................779107-01
NI Reconfigurable I/O
Software Development Kit................................................779149-03
NI 9211 ..................................................................................779001-01
NI 9215 ..................................................................................779011-01
NI 9215 with BNC..................................................................779138-01
NI 9217 ..................................................................................779592-01
NI 9221 ..................................................................................779014-01
NI 9221 with D-Sub ..............................................................779373-01
NI 9233 ..................................................................................779015-01
NI 9237 ..................................................................................779521-01
Analog Output Module
NI 9263 ..................................................................................779012-01
NI 9265 ..................................................................................779334-01
Digital Input and Counter/Timer Modules
NI 9401 ..................................................................................779351-01
NI 9411 ..................................................................................779005-01
NI 9421 ..................................................................................779002-01
NI 9421 with D-Sub ..............................................................779136-01
NI 9422 ..................................................................................779522-01
NI 9423 ..................................................................................779009-01
NI 9425 ..................................................................................779139-01
NI 9435 ..................................................................................779010-01
Digital Output and Relay Modules
NI 9472 ..................................................................................779004-01
NI 9472 with D-Sub ..............................................................779137-01
NI 9474 ..................................................................................779003-01
NI 9476 ..................................................................................779140-01
NI 9477 ..................................................................................779517-01
NI 9481 ..................................................................................779006-01
Accessories
NI 9912 ..................................................................................779019-01
NI 9915 ..................................................................................779018-01
NI 9917 ..................................................................................779454-01
NI 9918 ..................................................................................779455-01
NI 9932 ..................................................................................779017-01
NI 9933 ..................................................................................779103-01
NI 9934 ..................................................................................779104-01
NI 9935 ..................................................................................779016-01
NI 9936 ..................................................................................779105-01
NI 9937 ..................................................................................779106-01
NI 9939 ..................................................................................192692-01
Ethernet Cables
CAT 5
5 m ....................................................................................182219-05
CAT 5 Crossover
5 m ....................................................................................187375-05
OEM Discounts Available
NI offers aggressive OEM pricing discounts for high-volume OEM
applications. Contact NI for a quote.
Visit
Reconfigurable Embedded Systems
Real-Time Controllers
NI cRIO-9002 ........................................................................779000-01
NI cRIO-9004 ........................................................................779055-01
Reconfigurable Chassis
NI cRIO-9101 ........................................................................779052-01
NI cRIO-9102 ........................................................................779007-01
NI cRIO-9103 ........................................................................779053-01
NI cRIO-9104 ........................................................................779054-01
Power Supplies
NI PS-1 ..................................................................................777567-01
NI PS-2
U.S. 120 VAC/Japan 100 VAC ..........................................777584-01
North American 240 VAC ..................................................777584-05
NI PS-3
U.S. 120 VAC/Japan 100 VAC ..........................................777585-01
North American 240 VAC ..................................................777585-05
NI PS-4 ..................................................................................778586-90
NI PS-5 ..................................................................................778805-90
R Series Expansion Systems
Expansion Chassis
NI cRIO-9151 chassis ............................................................779008-01
SH68-C68-S cable, 2 m ........................................................186381-02
R Series Devices
PCI
NI PCI-7811R digital I/O ........................................................779363-01
NI PCI-7813R digital I/O ........................................................779370-01
NI PCI-7830R multifunction I/O ............................................779361-01
NI PCI-7831R multifunction I/O ............................................778797-01
NI PCI-7833R multifunction I/O ............................................779359-01
PXI
NI PXI-7811R digital I/O ........................................................778800-01
NI PXI-7813R digital I/O ........................................................779362-01
NI PXI-7830R multifunction I/O ............................................779364-01
NI PXI-7831R multifunction I/O ............................................778668-01
NI PXI-7833R multifunction I/O ............................................779360-01
C Series Modules
NI 9201 ..................................................................................779013-01
NI 9201 with D-Sub ..............................................................779372-01
NI 9203 ..................................................................................779516-01
NI 9205 ..................................................................................779519-01
NI 9205 with D-Sub ..............................................................779357-01
NI 9206 ..................................................................................779526-01
ni.com/compactrio
for up-to-date information on module availability.
BUY NOW!
For complete product specifications, pricing, and accessory
information, call 800 813 3693 (U.S.) or go to
ni.com/compactrio.
BUY ONLINE
at
ni.com
or
NI CompactDAQ Ordering Information
System Chassis
NI cDAQ-9172
U.S. (120 VAC) ..................................................................779508-01
Switzerland (220 VAC) ......................................................779508-02
Australia (240 VAC) ..........................................................779508-03
Europe (240 VAC) ..............................................................779508-04
UK (240 VAC) ....................................................................779508-06
Japan (100 V) ....................................................................779508-07
Accessories
NI 9901 ..................................................................................779473-01
NI 9940 ..................................................................................779567-01
NI 9941 ..................................................................................779568-01
C Series Modules
Digital Input Modules
Academic Bundles
Standard Bundle ....................................................................777448-60
Instrumentation Bundle ........................................................777448-61
DSA (Sound and Vibration) Bundle ......................................777448-62
Visit
ni.com/compactrio
or
ni.com/compactdaq
BUY NOW!
ni.com/compactdaq.
BUY ONLINE
at
ni.com
or