AC DC
130 Watts
JPS Series
High Efficiency
up to 90%
xppower.com
Medical Approval (JPS130 M Versions)
3” x 5” Footprint
100 W Convection cooled
130 W Forced cooled
Single & Multi Outputs
3 Year Warranty
Specification
Input
• 90-264 VAC (130-370 VDC)
• 47-63 Hz
• 2.0/1.0 A max at 115/230 VAC (JPS130)
2.6/1.3 A max at 115/230 VAC (JPS130-M)
Inrush Current
• 30 A at 115 VAC, 60 A at 230 VAC,
cold start at 25 °C
Power Factor
• 0.9 typical
Earth Leakage Current
• 1.5 mA max at 264 VAC/60 Hz (JPS130)
260
µA
max at 275 VAC/63 Hz (JPS130-M)
Input Protection
• JPS130 internal T3.15 A/250 VAC fused line,
JPS130-M fused line & neutral
Input Voltage
Input Frequency
Input Current
General
Efficiency
Isolation
• Up to 90% nominal line, full load
• 3000 VAC Input to Output (JPS130),
4000 VAC Input to Output (JPS130-M
2 x MOPP), 1500 VAC Input to Ground
(JPS130-M 1 x MOPP),
500 VAC Output to Ground
• 120 kHz, ±10% for both active PFC & PWM
• 6.7 W/In
• 200 kHrs to MIL-HDBK-217F at 50 °C, GB
Switching Frequency
Power Density
MTBF
Environmental
• 0 °C to +70 °C, (see derating curves)
Cooling
• JPS130PS03 83 W convection-cooling,
100 W, with 18 CFM airflow.
All others, 100 W with convection
cooling,130 W with 18 CFM airflow
Operating Humidity
• 5-95% RH, non-condensing
Storage Temperature
• -40 °C to +85 °C (JPS130)
-20 °C to +85 °C (JPS130-M)
Operating Altitude
• 3000 m
Output
• See table
• ±10% on single output models,
±5% on multi output models,
auxiliary outputs will track by the same
percentage
Initial Set Accuracy
• At 60% rated load ±1% single outputs,
±3% on V1 for multi outputs
Minimum Load
• Single output: no minimum load required
Multi-outputs: (see note 1)
Start Up Rise Time
• <20 ms
Hold Up Time
• 16 ms typical at 115 VAC & rated load (JPS130)
20 ms typical at 115 VAC & rated load (JPS130-M)
Drift
• <±0.2% after 20 mins warm up
Line Regulation
• ±0.5% Vmin to Vmax
Load Regulation
• ±1%, ±5% for auxiliary outputs,
20%-100% of rated load
Transient Response
• 4% max deviation, recovery to within
1% in 500 µs for a 25% load change
Ripple & Noise
• ±1% max pk-pk (see note 4)
Overvoltage Protection
• 110-140% Vnom on V1 only,
recycle input to reset
Overload Protection
• 110-150%, auto recovery
Short Circuit Protection
• Trip & restart (hiccup mode), auto recovery
Temperature
• 0.05%/°C
Coefficient
Remote Sense
• Single output models only, compensates for
up to 0.5 V drop
Fan Output
• 12 V, 450 mA multi-output models only
Output Voltage
Output Voltage Trim
EMC & Safety
Emissions
• EN55022, Class B conducted
FCC 20780, level B conducted
EN60601-1-2, (JPS130-M)
• EN61000-3-2 Class A
• EN61000-3-3
• EN61000-4-2, level 3 contact, level 3 air,
Perf Criteria A
• EN61000-4-3, 10 V/m Perf Criteria A
• EN61000-4-4, level 3 Perf Criteria A
• EN61000-4-5, Installation Class 3
Perf Criteria A
• EN61000-4-6, level 2, Perf Criteria A
• EN61000-4-11, 30% 10 ms, 60% 100 ms,
95% 5000 ms Perf Criteria A, B, B
• JPS130:
EN60950-1, UL60950-1,
CSA C22.2 No. 60950-1,
CE Mark LVD.
JPS130-M: UL60601-1, EN60601-1,
CSA C22.2 No. 601.1
CE Mark LVD.
Harmonic Currents
Voltage Flicker
ESD Immunity
Radiated Immunity
EFT/Burst
Surge
Conducted Immunity
Dips & Interruptions
Safety Approvals
AC DC
Models and Ratings
Output
Power
(3)
100 W
JPS130
Ripple & Noise
Pk-Pk
(4)
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
mV
Output
3.3 V
5.0 V
12.0 V
15.0 V
24.0 V
28.0 V
48.0 V
+5.0 V
+12.0 V
(2)
-12.0 V
(2)
+3.3 V
+5.0 V
(2)
+12.0 V
(2)
+5.0 V
+12.0 V
(2)
-12.0 V
(2)
+5.0 V
+15.0 V
(2)
-15.0 V
(2)
+5.0 V
+24.0 V
(2)
+12.0 V
(2)
Output Current
Convection-cooled
18 CFM
25.0
20.0
9.0
7.0
4.5
3.8
2.1
11.5
3.0
8.0
3.0
0.8
10.0
0.8
2.8
0.7
26.0
10.8
8.7
5.4
4.6
2.7
15.0
5.0
15.0
4.1
1.1
15.0
4.1
1.1
4.0
4.0
Standard Model
Number
JPS130PS03
JPS130PS05
JPS130PS12
JPS130PS15
JPS130PS24
JPS130PS48
JPS130PT31
JPS130PT30
Medical Model
Number
JPS130PS05-M
JPS130PS12-M
JPS130PS15-M
JPS130PS24-M
JPS130PS28-M
JPS130PS48-M
130 W
130 W
130 W
130 W
130 W
130 W
JPS130PT31-M
JPS130PT32-M
JPS130PT34-M
1. 20% minimum load required on V1 to maintain stated regulation for
auxiliary rails.
2. Load regulation of auxiliary rails is defined over the range of 20%-100%
of rated load.
3. Maximum power with 18 CFM of forced air.
4. Ripple and noise measured over 15 MHz bandwidth with a 47 µF
electrolytic capacitor and 0.47 µF ceramic capacitor.
5. Covers available, order part number ‘JPS130-M COVER’ or add suffix ‘-
C’ to receive the unit with cover fitted. Overall dimensions 89.5 x 143.0 x
40.0 mm. For industrial units order part number ‘JPS130 COVER’ or add
suffix ‘-C’ to receive the unit with cover fitted. Overall dimensions 90.6 x
142.9 x 39.8 mm. Fixing screws M3, maximum penetration depth 3mm
from cover outer surface.
6. Connector kits and cable harnesses available. Consult sales for options.
Mechanical Details
0.16 (4.0) Dia Mounting Hole (4 Places)
0.22 (5.7) [0.2 (5.0) -M]
3.00 (76.2)
TB1
Voltage Trim
TB3
4.55 (115.6) [4.60 (117.0) -M]
5.00 (127.0)
To Chassis Grounding x 2
1.18 (30.0)
-M version
2.55 (64.8)
[2.60 (66.2) -M]
TB2
TB2
JPS130-M PIN CONNECTIONS
TB2
PS05
PS 12 - 48
PT31
PT32
Return
Return
-12 V
-15 V
Return
Return
+12 V
+15 V
Return
Return
Return
Return
Return
+V
Return
Return
+V
+V
Return
Return
+V
+V
Return
Return
+V
+5 V
+5 V
+V
+5 V
+5 V
+5 V
+5 V
Type *5273-08A *5273-06A
*5273-09A
TB2
PS05
+V
+V
+V
+V
Return
Return
Return
Return
*5273-08A
JPS130 PIN CONNECTIONS
PS03, 12 - 48
PT31
+V
+5 V
+V
+5 V
+V
+5 V
Return
Return
Return
Return
Return
Return
Return
+12 V
-12 V
*5273-06A
*5273-09A
PT34
+12 V
+24 V
Return
Return
Return
Return
+5 V
+5 V
+5 V
0.10
(2.5)
All dimensions are in inches (mm).
Tolerance: 0.02 (0.4) max.
Weight: 0.5 lbs (240 g) approx.
TB3 PIN CONNECTIONS
Pin
Single
Multi
- Sense
0V
+ Sense
+12 V Fan
1. Single output industrial model shown, add extra 0.08” (2.0 mm) to overall
height for multi-output models. Dimensions followed by ‘-M’ apply for
medical units (note output connector orientation).
2. AC input connector (TB1): Molex 5277-02A or equivalent. Mates with
5196-N housings/5194/5225 crimps.
3. DC output connector (TB2): Molex 5273 or equivalent, mates with 5195-N
housing/5194/5225 crimps and 5239-N housing/5167/5168/2478/2578
crimps. Except JPS130PS03 which has a 6-position terminal screw
block: Dinkle part number EHK508V Series accepts 24-12 AWG wire.
Terminal screw tightening torque 4 lbs-in (0.45 Nm) maximum.
4. TB3: Molex 5045-02A or equivalent. Mates with KK 51191
housings/50802 crimps and KK 5051 housings/2759/5159 crimps. Single
output units - TB3 is remote sense. Multi-output units - TB3 is 12 V,
450 mA fan output.
PT30
+3.3 V
+3.3 V
Return
Return
Return
Return
Return
+5 V
+5 V
+12 V
*5273-10A
* Note: MOLEX connector type
Derating Curves
130
Load (W)
JPS130 and ‘-M’ with 18 CFM fan
(except JPS130PS03).
JPS130PS12, PS15, PS24 and PS48
and ‘-M’ with convection cooling.
JPS130PS03 with 18 CFM fan.
JPS130PS05, PT30, PT31, PT32 and
PT34 and ‘-M’ with convection cooling.
JPS130PS03 with convection cooling.
Ambient Temp (˚C)
17-Sept-12