TRUMPower
D E S C R IP T IO N
25-28 WATT MEDICAL POWER SUPPLIES
TM25 SERIES
( LV D )
The TM25 series of compact, open PCB constructed,
AC-DC switching power supplies are capable of
delivering 25 to 28 watts of continuous output
power. They operate at 85 to 264VAC input voltage
without the need of voltage selection. They are
ideally suited for use in medical equipments, safety
systems and monitoring equipments.
F E AT U R E S
Low safety ground leakage current
100% burn-in
8 5 t o 2 6 4 VAC w i d e i n p u t r a n g e
Input surge current protection
O v e r v o l ta g e p r o t e c t i o n
Overcurrent protection
Open PCB construction
C o m p l i a n t w i t h R o H S r e q u i r e m e n ts
S a f e t y Sta n d a r d A p p r o va l s
UL 2601-1
CSA C22.2 No. 601.1
I N P U T S P E C I F I C AT I O N S
I n p u t v o l ta g e
Input frequency
Input current
8 5 t o 2 6 4 VA C
47 to 63 Hz
0 . 7 A
r m s
f o r 11 5 VA C
0 . 4 A
r m s
f o r 2 3 0 VA C
E a r t h l e a k a g e c u r r e n t:
9 0 µ A m a x . @ 11 5 VA C , 6 0 H z
1 5 0 µ A m a x . @ 2 3 0 VA C , 5 0 H z
TÜV EN60601-1
GENERAL SPECIFICATIONS
Switching frequency:
42KHz ±5KHz
Efficiency:
70% minimum on single output
model with Vo
12V, 68% minimum
on the others
10 msec minimum at 110VAC
±0.5% maximum at full load
10 amps @ 115VAC or 25 amps
@ 230VAC at 25ºC cold start
4000VAC from input to output
1500VAC from input to ground
500VAC from output to ground
500,000 hours minimum at full load
at 25ºC ambient, calculated per
MIL- HDBK- 217F
Hold-up time:
OUTPUT SPECIFICATIONS
Output voltage/current:
Total output power:
Ripple and noise:
Overvoltage protection:
See Rating Chart
See Rating Chart
1% peak to peak max.
Provided on output #1 only;
set at 112-132% of its
nominal output voltage
All outputs protected to short
circuit conditions
Line regulation:
Inrush current:
Withstand voltage:
MTBF:
Overcurrent protection:
Temperature coefficient:
All outputs ±0.04%/ºC
maximum
Transient response:
Maximum excursion of 4%
or better on all models,
recovering to 1% of final
value within 500µs after a
25% step load change
EMC Performance (EN60601-1-2:2001)
E N 5 5 0 11 :
Class B conducted, Class B radiated
FCC:
VCCI:
EN61000-3-2:
EN61000-3-3:
EN61000-4-2:
EN61000-4-3:
EN61000-4-4:
EN61000-4-5:
EN61000-4-6:
EN61000-4-8:
E N 6 1 0 0 0 - 4 - 11 :
Harmonic distortion, Class A and D
Line flicker
ESD, ±8KV air and ±6KV contact
Radiated immunity, 3V/m for 80~2500 MHz
Fast transient/burst, ±2KV
Surge, ±1KV diff, ±2KV com.
Conducted immunity, 3Vrms
Magnetic field immunity, 3A/m
Voltage dips, 30% reduction for 500ms,
60% reduction for 100ms and >95%
reduction for 10ms
E N V I R O N M E N TA L S P E C I F I C AT I O N S
Operating temperature:
0ºC to +70ºC
Storage temperature
Relative humidity
Derating:
-40ºC to +85ºC
5% to 95% non-condensing
Derate from 100% at +50ºC
linearly to 50% at +70ºC
Tumbler Technologies + TRUMPower
3350 Scott Blvd., Bldg. 13, Santa Clara, California 95054, USA
Phone: 408-988-6616
Fax: 408-988-6622
email: sales@trumpower.com
Website: www.TRUMPower.com
MEDICAL GRADE
OUTPUT VOLTAGE/CURRENT RATING CHART
Product No.
TM25-S5.1
TM25-S12
TM25-S15
TM25-S24
TM25-S28
TM25-D23
TM25-D25
TM25-T31
TM25-T32
TM25-T39
NOTES:
1.
2.
3.
Vnom.
5.1V
12V
15V
24V
28V
+5.1V
+5.1V
+5.1V
+5.1V
+5.1V
Output #1
Imin. Imax.
0A
0A
0A
0A
0A
0.4A
0.4A
0.4A
0.4A
0.4A
5.5A
2.3A
1.9A
1.2A
1.0A
2.5A
2.5A
2.5A
2.5A
2.5A
Tol.
2%
1%
1%
1%
1%
3%
3%
3%
3%
3%
+12V
+24V
+12V
+15V
+24V
Vnom.
Output #2
Imin. Imax. Tol.
(N/A)
(N/A)
(N/A)
(N/A)
(N/A)
0.2A
1.5A
0.1A
0.2A
0.1A
0.1A
0.8A
1.5A
1.0A
0.8A
5%
5%
5%
5%
5%
-12V
-15V
-12V
Vnom.
TM25 SERIES
Output #3
Imin.
Imax. Tol.
(N/A)
(N/A)
(N/A)
(N/A)
(N/A)
(N/A)
(N/A)
0.05A
0.05A
0.05A
0.2A
0.2A
0.2A
4%
4%
4%
Maximum
Output Power
28W
28W
28W
28W
28W
25W
25W
25W
25W
25W
All multiple output models may be operated at no-load without damage. At no-load, output voltage tolerance
increases to 10%.
Ripple and noise: Peak-to-peak with 20MHz bandwidth and 10µF in parallel with a 0.1µF capacitor at rated line
voltage and load ranges.
Safety agency approvals are for the above listed models in PCB format. Add suffix “B” for L-bracket format, or
suffix “C” for enclosed format, e.g. TM25-T31C.
MECHANICAL SPECIFICATIONS
Single Output Models
NOTES:
1.
2.
3.
4.
5.
Dimensions shown in inch [mm]
Tolerance 0.03 [0.76] maximum
Input connector mates with Molex housing 09-50-3031 and Molex
2878 series crimp terminal.
Output connector mates with Molex housing 09-50-3041 and
Molex 2878 series crimp terminal.
Weight: 163 grams
Multiple Output Models
NOTES:
1.
2.
3.
4.
5.
Output connector mates with Molex housing 09-50-3061 and
Weight: 175 grams
[24.38]
Phone: 408-988-6616