Ultra-Low Pressure
Digital Sensor
SM9333/SM9336 Series
Differential Pressure Sensors
Pressure ranges: 125 and 250 Pa Differential
Accuracy after Autozero: < 1%FS
16 bit I
C Digital interface
Pressure calibrated and temperature compensated output
Compensated temperature range: -20 to 85
o
The SM9333/SM9336 are digital, ultra-low pressure sensors offering
state-of-the-art MEMS pressure transducer technology and CMOS
mixed signal processing technology to produce a digital, fully
conditioned, multi-order pressure and temperature compensated
sensor in JEDEC standard SOIC-16 package with a dual vertical
porting option. It is available in a differential pressure configuration.
The total accuracy after board mount and system level autozero is
less than 1%FS. The excellent warmup behavior and long term
stability further assures its expected performance over the life of the
part.
Combining the pressure sensor with a signal-conditioning ASIC in a
single package simplifies the use of advanced silicon micro-machined
pressure sensors. The pressure sensor can be mounted directly on a
standard printed circuit board and a high level, calibrated pressure
signal can be acquired from the digital interface. This eliminates the
need for additional circuitry, such as a compensation network or
microcontroller containing a custom correction algorithm.
The SM9333/SM9336 are shipped in sticks or tape & reel.
Medical
Sleep Apnea
CPAP
Ventilators
Gas Flow Instrumentation
Air Flow Monitors
Life Sciences
Industrial
HVAC
Airflow Measurement
Pressure Transmitters
Pneumatic Gauges
Pressure Switches
Safety Cabinets
Silicon Microstructures, Inc. 2001-2018. All rights reserved
DOC # 40DS9302.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 1
www.si-micro.com
1. Absolute Maximum Ratings
All parameters are specified at Vdd = 3.3 V supply voltage at 25
o
C, unless otherwise noted.
No.
Characteristic
Supply Voltage
Digital IO Voltage
Max. Digital IO Current (DC)
Storage Temperature
(a,b)
Proof Pressure
(a, c)
Burst Pressure
(a, d)
DD
IO,DIG
IO,DIG
STG
Proof
Burst
Minimum
3.0
-0.3
-10
-40
7 (1.0)
20 (2.9)
Maximum
5.5
VDD+0.3
+10
+125
Units
kPa (PSI)
kPa (PSI)
Notes:
a. Tested on a sample basis.
b. Clean, dry air compatible with wetted materials. Wetted materials include silicon, glass, RTV (silicone), gold, aluminum, copper, nickel,
palladium, epoxy, stainless steel and plastic (mold compound).
c. Proof pressure is defined as the maximum pressure to which the device can be taken and still perform within specifications after
returning to the operating pressure range.
d. Burst pressure is the pressure at which the device suffers catastrophic failure resulting in pressure loss through the device.
2. ESD
No.
2.1
ESD HBM Protection at all Pins
ESD(HBM)
Minimum
-2
Maximum
Units
kV
3. External Components
No.
Supply bypass capacitor
I2C Data and clock pull up resistors
VDD
Min.
Typ.
Units
nF
kOhm
Not tested in production
4. Calibrated Pressure Ranges
No.
Device Type
MIN
(Pa)
SM9333 – 125 Differential
SM9336 – 250 Differential
-125
-250
MAX
(Pa)
+125
+250
DOC # 40DS9302.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 2
5. Recommended Operating Conditions
www.si-micro.com
The recommended operating conditions must not be exceeded in order to ensure proper functionality of the device. All parameters
specified in the following sections refer to these recommended operating conditions unless stated otherwise.
No.
Supply Voltage
Low level input voltage at SDA, SCL
High level input voltage at SDA, SCL
Compensated Temperature
(b)
VDD
IN,I2C,lo
IN,I2C,hi
COMP
Min.
3.0
-0.3
0.8 * V
VDD
-20
-20
Typ.
3.6
0.9
VDD
+0.3
+85
85
Units
Notes:
6. OPERATING CHARACTERISTICS TABLE
All parameters are specified at Vdd = 3.3 V DC supply voltage at 25
o
No.
Characteristic
Supply Current
(e)
Digital Pressure Output
MIN
MAX
Digital Full Scale Span
Resolution (Digital Output)
Update Rate
Digital AZ SM9333 – 125 Differential
Output
Accuracy
(f, g, h)
Digital Output SM9333 – 125 Differential
Accuracy
(f)
Bandwidth
Position Sensitivity
Power-up
time
from supply VDD > 3.0 V
to output settled to 90%
of final value
VDD
OUT
MIN
OUT
MAX
FS
Minimum
Typical
-26,215
+26,214
52,429
Maximum
3.8
Units
Counts
Counts
Counts
Bits
Hz
ACC
AZ
ACC
BW
UP
-1.25
-1
-2.75
-1.5
0.2
0.1
± 0.5
± 0.3
1.25
%FS
%FS
%FS
%FS
Hz
%FS/g
ms
pressure step response;
Step response
output rising from 10% to
time
90% of final value
Step response
settling time
output settling to full
accuracy
RESP
ms
SETTLE
ms
Notes:
e. Supply current given for continuous operation. Device can be set to sleep mode with significantly lower power consumption. Refer to
section 11 for details.
f. The accuracy specification applies over all operating conditions in dry clean air. This specification includes the combination of linearity,
repeatability, and hysteresis errors over pressure, temperature, and 3.3 V supply voltage.
g. Based on system level autozero at 25
o
C after board mount or solder reflow.
h. Based on sample testing during initial product qualification, not tested in production.
DOC # 40DS9302.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 3
www.si-micro.com
7 I
C Interface
No.
SDA output low voltage
Low-to-High transition threshold
High-to-Low transition threshold
C clock frequency
Bus free time between a START and STOP condition
Clock low time
Clock high time
START condition hold time
Data setup time
Data hold time
Setup time for repeated START condition
Setup time for STOP condition
Rise time of SDA and SCL signals
Fall time of SDA and SCL signals
Condition
SDA
= 3 mA
pins SA0, SCL
pins SA0, SCL
SDA,OL
SDA,LH
SDA,HL
f
SCL
BUSF
LO
HI
SH
SU
RSH
PSU
Min.
Typ.
Max. Units
0.4
VDD
VDD
kHz
ns
ns
ns
ns
ns
ns
ns
ns
ns
ns
0.6
0.4
0.7
1300
8. Qualification Standards
REACH Compliant
RoHS Compliant
PFOS/PFOA Compliant
For qualification specifications, please contact Sales at sales@si-micro.com
DOC # 40DS9302.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 4
www.si-micro.com
9. Package Reference
SOIC-16 (C) Vertical Package Dimensions
Notes:
All dimensions in units of [mm]
Moisture Sensitivity Level (MSL): Level 3
Clean, dry air compatible with wetted materials. Wetted materials include : Wetted materials include silicon, glass, RTV (silicone), gold,
aluminum, copper, epoxy and mold compound.
Tolerance on all dimensions ±0.13 mm unless otherwise specified.
[B] is tube connected to bottom side of sensor die.
[T] is tube connected to top side of sensor die. Topside pressure is positive pressure. An increase in topside pressure will result in an
increase in sensor output
Bottom plate is stainless steel
Robust JEDEC SOIC-16 package for automated assembly
Manufactured according to ISO9001 , ISO14001 and ISO/TS 16949 standards
Lot Number Identification
DOC # 40DS9302.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 5