Medium Pressure
Digital Sensor
SM4221, SM4321, SM4421 Series
Gauge & Differential Pressure Sensors
Pressure ranges from 2.5 t0 14.9 PSI gauge, differential and
asymmetric outputs
14-bit digital, pressure calibrated and temperature compensated
output
C Digital Interface
Compensated temperature range: -20 to 85
o
Insensitive to mounting orientation
Robust JEDEC SOIC-16 package for automated assembly
Manufactured according to ISO9001 and ISO/TS 16949 standards
The SM4X21 Series is a digital, medium pressure MEMS sensor
family offering state-of-the-art pressure transducer technology and
CMOS mixed signal processing technology to produce a digital, fully
conditioned, multi-order pressure and temperature compensated
sensor in JEDEC standard SOIC-16 package with a dual vertical
porting option. It is available in both compound gage or differential
pressure configurations. With the dual porting, a vacuum-gage
measurement is possible to minimize altitude errors due to changes
in ambient pressure.
Combining the pressure sensor with a signal-conditioning ASIC in a
single package simplifies the use of advanced silicon micro-machined
pressure sensors. The pressure sensor can be mounted directly on a
standard printed circuit board and a high level, calibrated pressure
signal can be acquired from the digital interface. This eliminates the
need for additional circuitry, such as a compensation network or
microcontroller containing a custom correction algorithm.
Customer-specified pressure ranges and supply voltages are
available.
The SM4X21 Series is shipped in sticks or tape & reel.
Medical
Sleep Apnea
CPAP
Ventilators
Gas Flow Instrumentation
Air Flow Monitors
Negative Pressure Wound Therapy
Industrial
Airflow Measurement
Pneumatic Gauges
Pressure Switches
Safety Cabinets
Life Sciences
Consumer
Sports Equipment
Appliances
Silicon Microstructures, Inc. 2001-2016. All rights reserved
DOC # 40DS4960.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 1
www.si-micro.com
Absolute Maximum Ratings
All parameters are specified at Vdd = 3.3 V supply voltage at 25
o
C, unless otherwise noted.
No.
Characteristic
Supply Voltage
Supply Current
Update Period
Compensated Temperature
(a)
Storage Temperature
(a)
Media Compatibility
(a,b)
Proof Pressure
(a, c)
2.5 – 7.5 PSI
7.5 – 15 PSI
Burst Pressure
(a, d)
2.5 – 7.5 PSI
7.5 – 15 PSI
DD
VDD
Minimum
2.7
Typical
Maximum
6.00
4.00
Units
mS
COMP
OP
STG
-20
-40
-40
+85
+105
+125
Proof
45
75
PSI
PSI
PSI
PSI
Burst
Notes:
a. Tested on a sample basis.
b. Clean, dry gas compatible with wetted materials. Wetted materials include silicon, epoxy, RTV, gold and aluminum.
c. Proof pressure is defined as the maximum pressure to which the device can be taken and still perform within specifications after
returning to the operating pressure range
d. Burst pressure is the pressure at which the device suffers catastrophic failure resulting in pressure loss through the device.
DOC # 40DS4960.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 2
www.si-micro.com
OPERATING CHARACTERISTICS TABLE
All parameters are specified at Vdd = 3.3 V DC supply voltage at 25
o
No.
Characteristic
Supply Voltage*
Pressure Output
MIN
Pressure Output
MAX
Full Scale Span
Resolution
Accuracy
(e, f, g)
DD
OUT
MIN
OUT
MAX
FSP
Minimum
3.0
4.75
Typical
5.0
1,638
14,745
13,107
Maximum
3.6
5.25
Units
Counts
Counts
Counts
Bits
ACC
-1
+1
%FS
No.
Device Type
SM4221 – Gauge
SM4321– Differential
SM4421 -- Asymmetric
Customizable Calibrated Pressure Ranges
MIN
(PSI)
MAX
(PSI)
-14.9 to -2.0
-14.9 to 0
+2.5 to +14.9
+2.5 to +14.9
0 to +14.9
Comment
Absolute value of P
MIN
must match absolute value of P
MAX
Delta between P
MAX
and P
MIN
must be at least 2.0 PSI
Notes:
e. The accuracy specification applies over all operating conditions. This specification includes the combination of linearity, repeatability, and
hysteresis errors over pressure, temperature, and voltage.
f. Maximum 10-year zero pressure offset shift < ±1%FS based on 1000 hours of HTOL , TC and THB testing. Pressure offset shift < ±2%FS
between 2.0 and 2.5psi over 10-year span.
g. For less demanding applications, devices with relaxed accuracy specifications are available
*Custom calibration pressures and voltages are available to meet specific customer demands.
DOC # 40DS4960.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 3
www.si-micro.com
SM4X21 Series – I
C Communications
1. SCL Clock frequency:
100kHz to 400kHz
2. Slave Address
The factory setting for the I
C slave address is 28HEX. The
part will only respond to the set address.
For the Read_DF3 data fetch command (Data Fetch 3
Bytes), the sensor returns three bytes in response to the
master sending the slave address and the READ bit (1): two
bytes of bridge data with the two status bits as the MSBs
and then 1 byte of temperature data (8-bit accuracy). After
receiving the required number of data bytes, the master
sends the NACK and stop condition to terminate the read
operation.
For the Read_DF4 command, the master delays sending the
NACK and continues reading an additional final byte to
acquire the full corrected 11-bit temperature measurement.
In this case, the last 5 bits of the final byte of the packet are
undetermined and should be masked off in the application.
The Read_DF2 command is used if corrected temperature is
not required. The master terminates the READ operation
after the two bytes of bridge data.
3. Read Operations
For read operations, the I
C master command starts with
the 7-bit slave address with the 8th bit = 1 (READ). The
sensor as the slave sends an acknowledge (ACK) indicating
success.
The sensor has four I
C read commands: Read_DF2,
Read_DF3, and Read_DF4. The following figures show the
structure of the measurement packet for three of the four
C read commands, which are further explained below.
3.1 I
C Read_DF (Data Fetch):
For the Data Fetch commands, the number of data bytes
returned by the sensor is determined by when the master
sends the NACK and stop condition.
DOC # 40DS4960.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 4
www.si-micro.com
SCL Clock Frequency
Start Condition Hold Time Relative to SCL Edge
Minimum SCL Clock Low Width1
Minimum SCL Clock High Width1
Data Hold Time on SDA Relative to SCL Edge
Data Setup Time on SDA Relative to SCL Edge
Stop Condition Setup Time on SCL
Bus Free Time Between Stop Condition and Start Condition
FSCL
tHDSTA
tLOW
tHIGH
tSUSTA
tHIDDAT
tSUDAT
tSUSTO
tBUS
Minimum
0.6
0.6
0.1
0.0
0.1
0.1
2.0
Typical Maximum
Units
kHz
µs
µs
µs
µs
µs
µs
µs
µs
DOC # 40DS4960.02
+1-(408) 577-0100
sales@si-micro.com
www.si-micro.com
Page 5