These products are not customer serviceable. Repairs may only be carried out by TDK-Lambda.or their authorized agents. These products are not authorized for
use as critical components in nuclear control systems, life support systems or equipment for use in hazardous environments without the express written approval of
the Managing Director of TDK-Lambda Corporation.
Safety Class of Protection
These products are designed for the following parameters : Material Group IIIa, Pollution Degree 2, Over voltage Category II, Class I (earthed), Indoor use as part
of an overall equipment such that the product is accessible to service engineers only.
Input markings and symbols
Caution refer to supplementary documents
EMC performance
Immunity
Comments
Air discharge 8kV
Contact discharge 6kV
Electromagnetic field
IEC61000-4-3
Level 3
10V/m
Fast / burst transient
IEC61000-4-4
Level 3
Tested to 2kV
Surge immunity
IEC61000-4-5
Level 3,2
Common mode to 2.0kV
Differential mode 1.0kV
Conducted RF immunity
IEC61000-4-6
Level 3
10V
Power frequency magnetic field
IEC61000-4-8
Level 4
30A/m
Voltage dips, variations, interruptions
IEC61000-4-11
Class 3
*1 , *2
*1. At the condition of 60% dips and 100ms duration, the output voltage will recover after the power turn on again.
*2. At the condition of 100% dip and 5000ms duration, the output voltage will recover after the power turn on again.
Emissions
Test
Radiated electric field
Conducted emissions
Conducted harmonics
Flicker
Standard
EN55022
EN55022
IEC61000-3-2
IEC61000-3-3
Comments
Class B (as per CISPR 22)
Compliant
Compliant
Test
Electrostatic discharge
Standard
IEC61000-4-2
Passed
Level 3
General installation instructions
1)These products are Class
Ⅰ
and must therefore be reliably earthed and professionally installed in accordance with the prevailing electrical
wiring regulations and the safety standards covered herein.
2)These products are IPX0, and therefore chemicals/solvents, cleaning agents and other liquids must not be used.
3)The first protective earth connection in the final installation must be marked with the protective earth symbol.
-2-
TDK-Lambda
MWS65
Instruction Manual
60601-
Special Instructions for IEC/EN/UL/ 60601-1
1)
These products are designed for continuous operation within an overall enclosure, and must be mounted such that access to the mains terminals is restricted. See
Clause 16, IEC/EN/UL60601-1.
2) These products are NOT suitable for use in the presence of flammable anaesthetic mixtures with air or with oxygen or with nitrous oxide.
3) These products are classed as ordinary equipment according to IEC/EN/UL60601-1 and are NOT protected against the ingress of water.
4) Connect only apparatus complying with IEC/EN/UL60601-1 to the signal ports.
5) Reference should be made to local regulations concerning the disposal of these products at the of their useful life.
6) These products have not been assessed to IEC/EN60601-1-2 (EMC) but EMC test data is available from TDK-Lambda Corporation
-3-
TDK-Lambda
MWS65
Instruction Manual
Model name identification method
MWS 65 – 5
Rated Output Voltage
Output Power type
Series Name
Terminal Explanation
①
②
③
④
⑤
⑥
⑦
⑧
L : AC Input terminal Live line (Fuse in line.)
N : AC Input terminal Neutral line (Fuse in line.)
Mounting hole (hole diameter :
φ3.5mm)
The hole is connected to Protective Earth.
φ3.5mm)
These holes must be connected to electrically conductive spacer. The mounting surface of the spacer should be within Maxφ6mm.
φ3.5mm)
These holes are not connected to Protective Earth.
+ : + Output Terminal
Output Terminal
V.ADJ : Output voltage adjust trimmer. The output voltage rises when a trimmer is turned clockwise.
-4-
TDK-Lambda
MWS65
Instruction Manual
3. Terminal Connection Method
Pay attention to the input wiring. If it is connected to wrong terminal, the power supply will be damaged.
Input must be off when making connections.
Connect terminal of mounting hole to protective earth of the equipment.
Output current of each terminal pin must be less than 7A.
The output load line and input line shall be separated to improve noise sensitivity.
Do not apply stress to PCB, when connecting or removing connector.
Use input/output connector (housing) specified by the table below.
Use recommended crimping tool. Connector is not included with this product. (Refer to the following)
据外媒报道,萨里大学(University of Surrey)的研究人员开发出一种无需依赖GPS即可在人口密集的城市地区精确定位设备位置的人工智能系统。该系统可将定位误差从734米缩小到22米以内,这对于自动驾驶汽车和救援车辆等技术的发展意义重大。 图片来源: 萨里大学 在发表于《IEEE Robotics and Automation Letters》的论文中,研究人员介绍了PEn...[详细]
