MINT1065D
65W Single Output Series
Medical, 3
rd
Edition Certified Universal Input Power Supply
65W Convection Cooled AC-DC Power Supply
Universal Input 90-264 VAC
Class I or Class II input versions
Less 0.5W No Load Power Dissipation
Rugged EMC Compliant Design
2” x 4” x 1.2” Package
ANSI AAMI ES/CSA/EN/IEC 60601-1, 3rd Edition
3 Year Warranty
Certified
Meets Efficiency Level IV Efficiency
Compliant (LVD, RoHS)
Optional Chassis/Cover
All Specifications are typical at nominal input, full load at 25
C unless otherwise stated.
AC Input
100-240VAC +/- 10%, 47-63 Hz single phase.
Input Current
115VAC: 1.5A, 230VAC: 0.7A.
Inrush Current
The inrush at 264VAC, at a cold start, will not exceed
50A.
Input Fuses
F1, F2: 2.5A, 250VAC fuses provided on all models.
Efficiency
Up to 90% typical.
Earth Leakage Current
Earth Leakage Current <300μA, NC; <500μA SFC @ 264Vac, 60Hz
Touch Current is <100µA NC, <500µA SFC @ 264Vac, 60Hz,
measured from output to ground
Output Power
65W continuous.
Ripple and Noise
1% pk-pk max., 20MHz BW, differential mode. Measured with noise
probe directly across output terminals.
Output Voltage
See chart.
Voltage Adjustability
Voltage pre-set by fixed resistance.
Minimum Load
Not required.
Total Regulation
+/- 2%. See chart.
Transient Response
500μs max., 50% load step, typical.
Turn On Time
Less than 1 second under all rated load conditions.
Hold-up Time
<17 ms at 65W, nominal input (120VAC).
Overvoltage Protection
Built-in
Overload Protection
Hiccup Mode
Short Circuit Protection
Hiccup Mode
Isolation
Input-Output: 4000Vac/2 MOOP
Input-Ground: 1500Vac/1MOOP
Medical Safety Standards
Switching Frequency
Variable 50-75kHz typical.
0 to 70C (derate from full rated power above 40C).
Cooling
Convection
Output Power Derating
See derating chart
Storage Temperature
-40 to +85C.
Relative Humidity
5% to 95%, non-condensing
______________________________________________________________________________________________________________________________
1 of 3
SL Power Electronics Corp
6050 King Drive Ventura, CA 93003
Phone 805.486.4565
Fax 858.712.2040
Email: info@slpower.com
www.slpower.com
MINT1065D
Medical 3
rd
Output Parameters
Model
Number
MINT1065D1275K01
MINT1065D1375K01
MINT1065D1575K01
MINT1065D1875K01
MINT1065D2475K01
MINT1065D4875K01
Volts (V)
12V
13.2V
15V
18V
24V
48V
Output Current
Open Frame
w/cover
5.25A
4.80A
4.33A
3.50A
2.70A
1.35A
4.00A
3.63A
3.20A
2.66A
2.00A
1.00A
Ripple
& Noise
120mV
140mV
150mV
180mV
240mV
480mV
Total
Regulation
±2%
±2%
±2%
±2%
±2%
±2%
OVP
Threshold
15.0 ± 1.5V
16.0 ± 1.5V
18.0 ± 1.5V
21.0 ± 2.0V
27.0 ± 2.0V
55.0 ± 4.0V
3 Year Warranty
(continued)
Vibration
Operating:
0.003 g
/Hz, 1.5 g
rms
overall, 3 axes, 10 min. / axis
Non-Operating:
0.026 g
/Hz, 5.0 g
rms
overall, 3 axes, 1 hr. / axis
Pollution Degree
Dimensions
W 2.0” x L 4.0” x H 1.17”. See outline drawing.
Weight
150 grams
Shock
Operating:
Half-sine, 20 g
pk
, 10 ms, 3 axes, 6 shocks total
Non-Operating: Half-sine, 40 g
pk
Overvoltage Category
OVCII (2500V for Mains), OVCI (1500V for Secondary)
Operating Altitude
Up to 4000m.
Non-operating Altitude
-500 to 40,000 ft.
EMI/EMC Compliance
Conducted Emissions
Radiated Emissions
Static Discharge Immunity
Radiated RF Immunity
EFT/Burst Immunity
Line Surge Immunity
Conducted RF Immunity
Power Frequency Magnetic Field Immunity
Voltage Dip Immunity
EN55011 Class B, FCC Part 15, Class B.
EN55011 Class A, FCC Part 15, Class A.
EN61000-4-2, 6kV Contact Discharge, 8kV air discharge
EN61000-4-3, 3V/m.
EN61000-4-4, 2kV/5kHz..
EN61000-4-5, 1kV differential, 2kV common-mode
EN61000-4-6, 3Vrms
EN61000-4-8, 3A/m
EN61000-4-11 240VAC, 0%/0.5 cycle, 40%/5cycles, 70%/25
cycles
Model Number Key
MINT 1 065 X 12 75 K 01
Model:
Input Connector:
Output Connector:
Output Voltage:
Configuration:
Output Pow er:
# of Outputs:
Product Fam ily:
"01" = Standard Model, 02 and higher indicates a m odified m odel.
"K" = 3 pin Header (class I input); "C" = 2 pin Header (class II input).
"75" = Output Connector, 4 pin header. Other options available, consult factory.
"12" = 12V, "24" = 24V, etc.
"D" = EN60601-1 3rd Edition Certified, "C" = Optional chassis & cover (class I only).
"065" = 65W
"1" = Single output
"M" = Medical, "I" = Internal, "NT" = New Technology
2 of 3
Phone 805.486.4565
Fax 858.712.2040
www.slpower.com
MINT1065D
Medical 3
rd
Mechanical Drawings and Connector Information
3 Year Warranty
MINT1065 Series Connectors
J1
INPUT:
AMP P/N 641937-1
PIN 1) AC LINE
PIN 2) AC NEUTRAL
OUTPUT:
AMP P/N 640445-4
PIN 3) -Vout
PIN 4) -Vout
J2
PIN 1) +Vout
PIN 2) +Vout
MATING CONNECTORS:
P1
Molex 09-50-3031
P2
Molex 09-50-3041
Pins: Molex 08-50-0105
GND CONNECTION (Class I version)
0.250" FASTON TAB
Output Power Derating Curve
Data Sheet © 2011 SL Power Electronics Corp. The information and specifications contained herein are believed to be correct at the time of publication.
Rev. 2-28-13
However, SL Power accepts no responsibility for consequences arising from reproduction errors or inaccuracies. Specifications are subject to change without notice.
3 of 3
Phone 805.486.4565
Fax 858.712.2040
www.slpower.com