GlobTek, Inc.
www.globtek.com
186 Veterans Drive,
Northvale, NJ 07647
Tel. (201) 784-1000
Fax (201) 784-0111
Drawing Title: GTM96600-5409-T3, ITE / Medical Power Supply/Class 2/Household Power Supply, Efficiency Level: VI, Case
Color/Style: GlobTek Black, Desktop/External, Regulated Switchmode AC-DC Power Supply AC Adaptor, Input Rating:
100-240V~, 50-60Hz, IEC 60320/C14 , Output Rating: 54W, 9V@6A, Output Configuration: 1200 mm, 16/2 Cond, Round
Black, T-plug system, Ferrite: None, GlobTek Black, GT Cord # C6192, Pin Out: Polarity Configuration T00 All plugs are
Medical Grade, Locking, Polarized once installed, and single use only. Product demonstration link:
https://en.globtek.com/news/t00-changeable-output-plug-system-for-power-supplies MOLDED CONNECTORS AVAILABLE,
Blades/Input Cord: Not Applicable, Approvals: LPS India IP41 Option CB 230V CoC Tier 2 ETL EAC South African cETLus
cETLus cETLus S-Mark IRAM Ukraine CB Class I ETL NOM CCC RCM CB S-Mark ETL cUL Taiwan BSMI Korea PSE PSE
PSE WEEE VCCI CE Level VI China RoHS RoHS FCC
Model No. GTM96600-5409-T3
Part No./Ordering No. TR9CD6000T00CIMR6B
Rev D
REVISION HISTORY
REV
SUB
QA
QA
rchiu
rchiu
DATE
08/31/2018
03/17/2017
02/22/2017
11/22/2016
APPROVED
QA
QA
QA
QA
Update RoHS 2 to RoHS, Ukraine UKRSepro earth, CCC, Add ETL.EAC
Change ferrite size to 17.5*28.5*12.5 mm, 2 pass (1 turn)
Updating output cord from T00/C6179 to T00/C6192
Initial release
In Addition to GlobTek Inc.'s renewed ISO9001:2008 - Quality Management System Certification, GlobTek Inc. is now certified to:
ISO13485:2003 - Medical Devices Quality Management System Certification
ISO14001:2004 - Environmental Management System Certification
ISO Certificates are available online at
http://www.globtek.com/iso-certificates/
Customer Approval of Specification:
Please approve, sign and send back to GlobTek so we can complete order processing.
A delay in receipt of this form will delay delivery schedule.
Company Name:
Customer P/N:
Quote Number:
Date:
Authorized Representative Name:
Authorized Representative Signature:
The acceptance of this specification and use of the product described in this document indicates the acceptance and binding of the customer to
Globtek terms and conditions, which supersede all other agreements, terms, and conditions (
http://en.globtek.com/terms-and-conditions/
Footnote:
GlobTek Inc. will not be liable for the safety and performance of these power supplies if unauthorized access and repair occurs. End user should consult applicable UL, CSA or EN
standards for proper installation instruction.
Limitation of Use:
GlobTek product are not authorized for use as mission critical components in life support hazardous environment, nuclear or aircraft applications without prior written approval from
the CEO of GlobTek Inc.
Contents of this document are subject to change without prior notice.
GTM96600-5409-T3, 54W, 9V@6A, Desktop/External, Regulated Switchmode AC-DC Power Supply AC Adaptor
PROPRIETARY INFORMATION
PROPRIETARY OF GLOBTEK, INC. ANY REPRODUCTION, DISCLOSURE OR USE OF THIS DRAWING, IN WHOLE OR IN PART, IS HEREBY PROHIBITED EXCEPT AS SPECIFIED IN
WRITING BY GLOBTEK, INC.
guest /11/23/2018 6:49:51 PM CLONED FROM TR9CE5000YL3CIMR6B
Page 1 of 15
GlobTek, Inc.
www.globtek.com
186 Veterans Drive,
Northvale, NJ 07647
Tel. (201) 784-1000
Fax (201) 784-0111
Rev D
POWER SUPPLY INFORMATION
TYPE:
Desktop/External
TECHNOLOGY:
CASE COLOR:
GlobTek Black
NAMEPLATE RATED
100-240V~, 50-60Hz, 1.5A
INPUT:
INPUT CONFIG:
IEC 60320/C14
WATTS:
54.0
VOLTS OUT:
CURRENT OUT (Amps):
BLADE/CORD INCLUDED: N/A:
BLADE/CORD INSTALLED: N/A:
GROUNDING
(E1) Earthed output
CONFIGURATION
EFFICIENCY LEVEL:
VI
OUTPUT
1200 mm, 16/2 Cond, Round Black, T-plug system, Ferrite: None, GlobTek Black, GT Cord # C6192
CONFIGURATION
Polarity Configuration T00 All plugs are Medical Grade, Locking, Polarized once installed, and single use only. Product
CONNECTOR PIN OUT:
demonstration link: https://en.globtek.com/news/t00-changeable-output-plug-system-for-power-supplies MOLDED
LABEL SPECS:
Standard GT,L-1186
PACK SPEC:
Packing for DT plug model with T00 plug and manual
Notes/Modifications/Customizations (notes in this section supersede the values shown in the standard spec):
DIMENSIONS ARE IN MM UNLESS SPECIFIED OTHERWISE.
Page 2 of 15
GlobTek, Inc.
www.globtek.com
186 Veterans Drive,
Northvale, NJ 07647
Tel. (201) 784-1000
Fax (201) 784-0111
Rev D
ELECTRICAL SPECIFICATIONS
A) ELECTRICAL SPECIFICATIONS:
01. Input Voltage: Specified 90-264 Vac, Nameplate rated: 100-240Vac
90-264 Vac range @ 100% of rated load current
85-264 Vac range @ 85% of rated load current
110-370 VDC range @ 100% of rated load current
02. Input Frequency: Specified 47-63 Hz, Nameplate rated 50-60Hz
03. Output Regulation: +/- 5% measured at the output connector
04. Line Voltage Regulation: +/- 0.5% typical measured at full load
05. Green Power On Indicator LED
06. Output Ripple (Vp-p): 1% or 100 mV whichever is greater, measured at 20 MHz bandwidth with 0.1 uf ceramic capacitor in parallel with
a low impedance 47 uf electrolytic capacitor connected at the end of the output connector
07. Turn-ON/OFF Overshoot: 5% maximum, 1 mS typical recovery time for 40% to 70% step load
08. Turn-ON Delay: 1 second maximum @ full load, nominal line
09. Hold-Up Time: 8 mS typical @ nominal input voltage and full load
10. Inrush Current: 30A/60A maximum cold start @ 115/230Vac input
11. Efficiency: Compliant with Efficiency Level VI and CoC Tier 2 standards
12. No Load Standby Power: <0.075 W @ 230Vac for <= 49W output power rating,
<0.15W @ 230Vac for > 49W output power rating
B) PROTECTION
01. Input Protection: Input line fusing and 300Vac MOV
02. Short Circuit/ Overload: Electronically Protected unit will auto recover upon removal of fault
Output Current Limit: 110% to 160% of rated output current
(upper end of current limit range is determined by normal output power rating,
not derated output power level)
03. Output Over-Voltage: 110% to 130% of nominal output voltage
C) SAFETY
01. Dielectric Withstand Voltage: 4000Vac or 5656Vdc from input to output,
On Class I models, 3000Vac or 4242Vdc from input to earth
02. Earth Leakage Current: Class I models < 300uA, N/A for Class II models
03. Touch Current: Class I models < 20uA, Class II models < 75uA
04.
Output Isolation Options
a) C8 or C18 Inlet, Class II
b) C6 or C14 Inlet, Class II FE, Output Isolated from Earth contact
c) C6 or C14 Inlet, Class I, Output negative directly attached to Earth contact
05. Earth Continuity Test: < 0.1 Ohm between Earth Pin at AC input and PCB termination point (Class I models only)
06. Means of Protection: 2 x MOPP
07. Compliant Standards: See listings at end of this drawing for specifics
D) EMC
EN 60601-1-2, 4
th
edition
Emissions, per EN 55032, EN 61000-6-3, EN 61000-6-4
Conducted Emissions: Class B, FCC Part 15, Class B
Radiated Emissions: Class B, FCC Part 15, Class B
Line Frequency Harmonics EN61000-3-2, Class A
Voltage Fluctuations/Flicker EN61000-3-3
Immunity, per EN 55024, EN 61000-6-1, EN 61000-6-2
Static Discharge Immunity EN61000-4-2, 10kV Contact Discharge, 20kV air discharge
Radiated RF Immunity EN61000-4-3, 10V/m 80-1000MHz, 3V/m 1-2.7GHz, 80% 1KHz AM.
EFT/Burst Immunity EN61000-4-4, 4kV/100kHz.
Line Surge Immunity EN61000-4-5, 2kV differential, 4kV common-mode
Conducted RF Immunity EN61000-4-6, 3Vrms, 80% 1KHz AM
Power Frequency Magnetic Field Immunity EN61000-4-8, 3A/m
Voltage Dip Immunity EN61000-4-11, Criteria
E) OTHER:
01. MTBF: 1,000,000 Hours @ 40°C ambient temperature, Full Load
02. GTM96600 Family: -20°C to 40°C ambient temperature with full load
Page 3 of 15
GlobTek, Inc.
www.globtek.com
186 Veterans Drive,
Northvale, NJ 07647
Tel. (201) 784-1000
Fax (201) 784-0111
Rev D
Regarding Operating Temperature,
a- See below derating table for output power capability at alternate temperature
b- Extended low end temperature range available as custom option
03. Operating Humidity: 0% to 95% relative humidity, non-condensing
04. Storage Temperature: -30°C to 80°C
05. Operating Altitude: 5000 Meters
06. Ingress Protection: IP40 standard, IP52 available as special option
07. ROHS: Compliant with latest regulations, see approvals section below
F) ENCLOSURE
01. Housing: High impact plastic, 94V0 polycarbonate, non-vented
02. Markings: Label or Laser Printed
03. AC Input mechanical options: Desktop C6, C8, C14 or C18 IEC Inlet.
Hybrid option (Desktop or Changeable Blade Wall Plug-in) Class I or Class II input
G) SPECIAL OPTIONS
01. Cost reduction, removal of LED
02. Custom Cordsets, various cordage types, and connector types
03. Custom Markings
04. Short term Output Surge Capability
05. Reduced Leakage Current versions
06. Tightened output voltage tolerance
07. Reduced output ripple level
08. Reduced output power rating marking
09. High Rel PCB laminate with Plated through Holes for IPC610 Class 2 Compliance
10. Low Temperature operation, down to -40°C
11. Special Housing Colors and Cordset Colors
H) OUTPUT CONNECTORS
Several output connector options are available with various output current ratings.
GlobTek can supply 10A rated 2.1mm and 2.5mm style DC Power Jacks, to complement our
10A output rated 2.1mm and 2.5mm DC power plugs used on our output cordsets.
Please visit https://en.globtek.com/news/high-current-coaxial-barrel-plug-jacks-for-high-wattage-power-supplies for a real time product
offering of mating connectors.
Page 4 of 15
GlobTek, Inc.
www.globtek.com
186 Veterans Drive,
Northvale, NJ 07647
Tel. (201) 784-1000
Fax (201) 784-0111
Rev D
Page 5 of 15