19-1481; Rev 1; 7/02
IT
TION K
VALUA
BLE
AVAILA
+3V, 18-Bit, Low-Power, Multichannel,
Oversampling (Sigma-Delta) ADC
General Description
o
18-Bit Resolution, Sigma-Delta ADC
o
16-Bit Accuracy with No Missing Codes to 480sps
o
Matched On-Board Current Sources (200µA) for
Sensor Excitation
o
Low Quiescent Current
250µA (operating mode)
2µA (power-down mode)
o
3 Fully Differential or 5 Pseudo-Differential Signal
Input Channels
o
2 Additional, Fully Differential Calibration
Channels/Auxiliary Input Channels
o
Programmable Gain and Offset
o
Fully Differential Reference Inputs
o
Converts Continuously or On Command
o
Automatic Channel Scanning and Continuous
Data Output Mode
o
Operates with Analog and Digital Supplies
from +2.7V to +3.6V
o
SPI™/QSPI™-Compatible 3-Wire Serial Interface
o
28-Pin SSOP Package
MAX1403
The MAX1403 18-bit, low-power, multichannel, serial-
output analog-to-digital converter (ADC) features
matched 200µA current sources for sensor excitation.
This ADC uses a sigma-delta modulator with a digital
decimation filter to achieve 16-bit accuracy. The digital
filter’s user-selectable decimation factor allows the con-
version resolution to be reduced in exchange for a
higher output data rate. True 16-bit performance is
achieved at an output data rate of up to 480sps. In
addition, the modulator sampling frequency may be
optimized for either lowest power dissipation or highest
throughput rate. The MAX1403 operates from +3V.
This device offers three fully differential input channels
that may be independently programmed with a gain
between +1V/V and +128V/V. Furthermore, it can com-
pensate an input-referred DC offset up to 117% of the
selected full-scale range. These three differential chan-
nels may also be configured to operate as five pseudo-
differential input channels. Two additional, fully
differential system-calibration channels are provided for
gain and offset error correction.
The MAX1403 can be configured to sequentially scan all
signal inputs and provide the results via the serial inter-
face with minimum communications overhead. When
used with a 2.4576MHz or 1.024MHz master clock, the
digital decimation filter can be programmed to produce
zeros in its frequency response at the line frequency and
associated harmonics, ensuring excellent line rejection
without the need for further postfiltering.
The MAX1403 is available in a 28-pin SSOP package.
Pin Configuration
TOP VIEW
CLKIN 1
28 SCLK
27 DIN
26 DOUT
25 INT
24 V
DD
Portable Industrial Instruments
Portable Weigh Scales
Loop-Powered Systems
Pressure Transducers
CLKOUT 2
CS 3
RESET 4
DS1 5
DS0 6
OUT2 7
OUT1 8
AGND 9
MAX1403
23 DGND
22 CALOFF+
21 CALOFF-
20 REFIN+
19 REFIN-
18 CALGAIN+
17 CALGAIN-
16 AIN6
15 AIN5
Ordering Information
PART
MAX1403CAI
MAX1403EAI
TEMP RANGE
0°C to +70°C
-40°C to +85°C
PIN-PACKAGE
28 SSOP
28 SSOP
V+ 10
AIN1 11
AIN2 12
AIN3 13
AIN4 14
SPI and QSPI are trademarks of Motorola, Inc.
SSOP
________________________________________________________________
Maxim Integrated Products
For pricing, delivery, and ordering information, please contact Maxim/Dallas Direct! at
1-888-629-4642, or visit Maxim’s website at www.maxim-ic.com.
MAX1403
ABSOLUTE MAXIMUM RATINGS
V+ to AGND, DGND .................................................-0.3V to +6V
DD
to AGND, DGND ...............................................-0.3V to +6V
AGND to DGND.....................................................-0.3V to +0.3V
Analog Inputs to AGND................................-0.3V to (V+ + 0.3V)
Analog Outputs to AGND .............................-0.3V to (V+ + 0.3V)
Reference Inputs to AGND...........................-0.3V to (V+ + 0.3V)
CLKIN and CLKOUT to DGND...................-0.3V to (V
DD
+ 0.3V)
All Other Digital Inputs to DGND..............................-0.3V to +6V
All Digital Outputs to DGND .......................-0.3V to (V
DD
+ 0.3V)
Maximum Current Input into Any Pin ..................................50mA
Continuous Power Dissipation (T
= +70°C)
28-Pin SSOP (derate 9.52mW/°C above +70°C) ........524mW
Operating Temperature Ranges
MAX1403CAI .....................................................0°C to +70°C
MAX1403EAI...................................................-40°C to +85°C
Storage Temperature Range .............................-60°C to +150°C
Lead Temperature (soldering, 10s) .................................+300°C
Stresses beyond those listed under “Absolute Maximum Ratings” may cause permanent damage to the device. These are stress ratings only, and functional
operation of the device at these or any other conditions beyond those indicated in the operational sections of the specifications is not implied. Exposure to
absolute maximum rating conditions for extended periods may affect device reliability.
ELECTRICAL CHARACTERISTICS
(V+ = +2.7V to +3.6V, V
DD
= +2.7V to +3.6V, V
REFIN+
= +1.25V, REFIN- = AGND, f
CLKIN
= 2.4576MHz, T
= T
MIN
to T
MAX
, unless
otherwise noted. Typical values are at T
= +25°C.)
PARAMETER
STATIC PERFORMANCE
Noise-Free Resolution
Output Noise
Integral Nonlinearity
(Note 1)
Nominal Gain (Note 2)
Unipolar Offset Error
Unipolar Offset Drift
Bipolar Zero Error
Bipolar Zero Drift
Positive Full-Scale Error
(Note 3)
Full-Scale Drift (Note 4)
Gain Error (Note 5)
Gain-Error Drift (Note 6)
Bipolar Negative Full-Scale Error
Bipolar Negative Full-Scale Drift
For gains of 1, 2, 4
For gains of 8, 16, 32, 64, 128
For gains of 1, 2, 4, 8, 16, 32, 64
For gain of 128
For gain of 128
For gain of 128
For gain of 128
-2.5
-3.5
0.8
0.3
-2
-3
3.5
-2.5
-3.5
0.8
0.3
Relative to nominal of 1% FSR
-2.0
0.8
0.3
3.5
-1
2.0
INL
No missing codes guaranteed by design;
for filter settings with FS1 = 0
Depends on filter setting and selected gain
Bipolar mode; FS1 = 0; MF1, MF0 = 0
FS1 = 0; MF1, MF0 = 1, 2, 3
(Tables 16a, 16b)
-0.0015
±0.001
0.98
%FSR
µV/°C
%FSR
µV/°C
%FSR
µV/°C
%FSR
ppm/°C
%FSR
µV/°C
0.0015
%FSR
Bits
SYMBOL
CONDITIONS
MIN
TYP
MAX
UNITS
_______________________________________________________________________________________
MAX1403
ELECTRICAL CHARACTERISTICS (continued)
DD
= +2.7V to +3.6V, V
REFIN+
CLKIN
= 2.4576MHz, T
= T
MIN
to T
MAX
, unless
= +25°C.)
PARAMETER
OFFSET DAC
Offset DAC Range (Note 7)
Offset DAC Resolution
Offset DAC Full-Scale Error
Offset DAC Zero-Scale Error
Additional Noise from Offset
DAC (Note 8)
DAC code = 0000
Unipolar mode
Bipolar mode
Unipolar mode
Bipolar mode
Input referred
Gain = 1, 2, 4, 8, 16, 32, 64
Gain = 128
-2.5
-3.5
-58.35
16.7
8.35
+2.5
+3.5
116.7
58.35
%FSR
%FSR
%FSR
%FSR
µV
RMS
SYMBOL
CONDITIONS
MIN
TYP
MAX
UNITS
ANALOG INPUTS/REFERENCE INPUTS
(Specifications for AIN and REFIN, unless otherwise noted.)
At DC
Common-Mode Rejection
CMR
For filter notch 50Hz, ±0.02
f
NOTCH
MF1 = 0, MF0 = 0, f
CLKIN
= 2.4576MHz (Note 9)
For filter notch 60Hz, ±0.02
f
NOTCH
MF1 = 0, MF0 = 0, f
CLKIN
Normal Mode 50Hz Rejection
(Note 9)
Normal Mode 60Hz Rejection
(Note 9)
Common-Mode Voltage Range
(Note 10)
Absolute Input Voltage Range
Absolute and Common-Mode
AIN Voltage Range
DC Input Leakage Current
(Note 11)
AIN Input Current (Note 11)
NMR
NMR
f
NOTCH
MF1 = 0, MF0 = 0, f
CLKIN
= 2.4576MHz
f
NOTCH
MF1 = 0, MF0 = 0, f
CLKIN
= 2.4576MHz
REFIN and AIN for BUFF = 0
BUFF = 1
REFIN and AIN for
BUFF = 0
BUFF = 1
Gain = 1
AIN Input Capacitance
(Notes 12)
BUFF = 0
Gain = 2
Gain = 4
Gain = 8, 16, 32, 64, 128
BUFF = 1, all gains
AIN Differential Voltage Range
(Note 13)
Unipolar input range (U/B bit = 1)
Bipolar input range (U/B bit = 0)
34
38
45
0 to V
REF
/ gain
±V
REF
/ gain
pF
= +25°C
= T
MIN
to T
MAX
90
AGND
AGND
- 30mV
AGND
+ 200mV
V+
V+
+ 30mV
V+
- 1.5
dB
dB
pA
nA
nA
dB
MAX1403
DD
= +2.7V to +3.6V, V
REFIN+
CLKIN
= 2.4576MHz, T
= T
MIN
to T
MAX
, unless
= +25°C.)
PARAMETER
AIN and REFIN Input Sampling
REFIN+ - REFIN- Voltage
(Note 14)
LOGIC INPUTS
Input Current
Input Low Voltage
Input High Voltage
Input Hysteresis
LOGIC OUTPUTS
Output Low Voltage (Note 15)
Output High Voltage (Note 15)
Floating-State Leakage Current
Floating-State Output
Current
Initial Tolerance
Drift
TRANSDUCER EXCITATION CURRENTS
Current
Initial Tolerance
Drift
Match
Drift Match
Compliance Voltage Range
POWER REQUIREMENTS
V+ Voltage
DD
Power-Supply Rejection V+
(Note 17)
V+
DD
PSR
For specified performance
2.7
2.7
(Note 18)
3.6
3.6
dB
AGND
OUT1 to OUT2
V+ - 1.0
±1
EXC
uA
ppm/°C
ppm/°C
OL
OH
DOUT and
INT,
SINK
= 100µA
CLKOUT, I
SINK
= 10µA
DOUT and
INT,
SOURCE
= 100µA
CLKOUT, I
SOURCE
= 10µA
DD
- 0.3
DD
- 0.3
-10
0.4
µA
pF
IN
IL
IH
HYS
All inputs except CLKIN
CLKIN only
CLKIN only
+10
0.4
0.4
µA
mV
SYMBOL
f
±5% for specified performance; functional
with lower V
REF
CONDITIONS
MIN
TYP
(Table 15)
1.25
MAX
UNITS
Hz
TRANSDUCER BURN-OUT
(Note 16)
BO
0.1
±10
±0.05
µA
%/°C
DD
= +2.7V to +3.6V, V
REFIN+
CLKIN
= 2.4576MHz, T
= T
MIN
to T
MAX
, unless
= +25°C.)
PARAMETER
SYMBOL
CONDITIONS
MIN
TYP
MAX
UNITS
MAX1403
ANALOG POWER-SUPPLY CURRENT
(Measured with digital inputs at either DGND or V
DD
, external CLKIN, burn-out and
transducer excitation currents disabled, X2CLK = 0, CLK = 0 for 1.024MHz, CLK = 1 for 2.4576MHz.)
V+ Standby Current (Note 19)
PD bit = 1, external clock stopped
Normal mode,
MF1 = 0,
MF0 = 0
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
Buffers off
Buffers on
370
250
610
245
610
0.42
1.2
0.42
1.2
1.8
4.8
1.8
4.8
1.8
4.8
2.2
420
µA
µA
2X mode,
MF1 = 0,
MF0 = 1
V+ Current
V+
4X mode,
MF1 = 1,
MF0 = 0
8X mode,
MF1 = 1,
MF0 = 1
DIGITAL POWER-SUPPLY CURRENT
DD
DD
Normal mode,
MF1 = 0, MF0 = 0
2X mode,
MF1 = 0, MF0 = 1
Digital Supply Current
DD
4X mode,
MF1 = 1, MF0 = 0
8X mode,
MF1 = 1, MF0 = 1
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
1.024MHz
2.4576MHz
0.17
0.11
0.22
0.15
0.32
0.50
0.40
0.35
µA
µA