A high performance AC to DC external power supply family designed for medical applications. The ME60A Medical Series low
power external AC-DC power supplies are approved to safety EN/IEC/UL60601-1, 3rd edition and EN/IEC/UL60601-1-11:2010
for Home Healthcare (non-hospital use) applications with and isolation levels which satisfy the 2 MOPP requirements and
designed to UL/EN/IEC60601-1-2, 4th edition for EMC
. The ME60A Series models will operate at universal input range of 90 to
264Vac over the wide temperature range of -20°C to +70°C, delivering full rated output power up to +40°C and applicable
output power derating at 70°C. These models are available in desktop versions, include an IP22 rating per IEC60529 for the
enclosure, and the output cable can be terminated at a variety of output connectors.
Consult Factory for Table 9 compliance information.
Model Selection
Model
Number
ME60A0551F01
ME60A0903F01
ME60A1203F01
ME60A1503F01
ME60A1803F01
ME60A2403F01
ME60A4803F01
ME60A0551N01
ME60A0903N01
ME60A1203N01
ME60A1503N01
ME60A1803N01
ME60A2403N01
ME60A4803N01
ME60A0551Q01
ME60A0903Q01
ME60A1203Q01
ME60A1503Q01
ME60A1803Q01
ME60A2403Q01
ME60A4803Q01
Notes:
Volts
5.0V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
5.0V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
5.0V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
Output
Current
7.00A
6.00A
5.00A
4.00A
3.30A
2.70A
1.35A
7.00A
6.00A
5.00A
4.00A
3.30A
2.70A
1.35A
7.00A
6.00A
5.00A
4.00A
3.30A
2.70A
1.35A
Output
Power
35W
56W
60W
60W
60W
60W
60W
35W
56W
60W
60W
60W
60W
60W
35W
56W
60W
60W
60W
60W
60W
Ripple &
Noise¹
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
Line
Regulation
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
Load
Regulation
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
Output
Connector
6 pin Molex Type
2.5 x 5.5 x 9.5mm
Straight Barrel
Type, center
positive
6 pin Molex Type
2.5 x 5.5 x 9.5mm
Straight Barrel
Type, center
positive
6 pin Molex Type
2.5 x 5.5 x 9.5mm
Straight Barrel
Type, center
positive
Output
Cable
1150mm, #18AWG
9V:1150mm
18AWG
All others:
1500mm, #18AWG
1150mm, #18AWG
9V:1150mm
18AWG
All others:
1500mm, #18AWG
1150mm, #18AWG
9V:1150mm
18AWG
All others:
1500mm, #18AWG
Input
Configuration
Class I Desktop,
IEC60320 C14
Receptacle
Class II Desktop,
IEC60320 C8
Receptacle
Class II Desktop,
IEC60320 C18
Receptacle
1. Measured at the output connector, with noise probe directly across output and load terminated with 0.1μF ceramic and 10μF low ESR capacitors. For 5V
models, values listed are typical, 100mV pk-pk maximum.
2. Molex p/n 39-01-2060 or equivalent. See outline drawing for pinout information.
3. For Input Class I models: For AC GND connected to output common (-), insert a “B” in the part number where the “A” is located (TE60B1203F01).
4. All specifications are typical at nominal input, full load, at 25°C ambient unless noted.
www.slpower.com
ME60 Rev 2.9
9-NOV-17
ME60
60W Single Output External Power
Medical Series
General Specifications
AC Input
Specification
100-240Vac, ±10%, 47-63Hz, 1∅
Turn On Time
Specification
Less than 1 sec @115Vac, full load
Input Current
100Vac: 1.5A, 240Vac: 0.7A
Hold-up Time
Overtemperature
Protection
Overload Protection
Short Circuit
Protection
Overvoltage
Protection
Isolation
20mS min., at full Load, 100Vac input
Inrush Current
264Vac, cold start: will not exceed 40A
Will shutdown upon an overtemperature
condition, auto-recovery.
Input Fuses
Earth Leakage
Current
(Input to Ground)
Patient Leakage
Current
(Output to Earth)
Efficiency
F1, F2: 2A, 250Vac fuses (line & neutral lines)
provided on all models
<500μA@264Vac, 60Hz, NC
<1mA@264Vac, 60Hz, SFC
<100μA@264Vac, 60Hz, NC
<500µA@264Vac, 60Hz, SFC
130 to 180% of rating, Hiccup Mode
Hiccup Mode, auto recovery.
130 to 150% of output voltage (max. 60V on
48V model), hiccup mode
Input-Output: 2 MOPP
Input-Ground: 1 MOPP
Output-Ground: 1 MOPP
>88%, typical
Output Power
No Load Input
Power
Ripple and Noise
60W continuous – See models chart for
specific voltage model ratings.
<0.210W (meets DoE Efficiency Level VI
Requirements)
Safety Standards
Operating
Temperature
Storage Temperature
-20°C to +70°C. See curve for derating.
See models chart on pg 1.
-40°C to +85°C
Output Voltage
Transient
Response
Regulation
500µs response time for return to within 0.5%
of final value for any 50% load step over the
range of 5% to 100% of rated load,
∆i/∆t<
0.2A/µs. Max. voltage deviation is +/-3.5%.
Operating: 0.003g/Hz, 1.5grms overall, 3
axes, 10 min/axis, 1-500Hz.
Non-Oper.: random waveform, 3 minutes per
axis, 3 axes and Sine waveform, Vib.
frequency/acceleration: 10-500Hz/1g, sweep
rate of 1 octave / minutes, Vibration time of 10
sweeps / axes, 3 axes
W: 2.67” x L: 4.25” x H: 1.29”
W: 67.9mm x L: 108mm x H: 32.7mm
Altitude
Operating: to 5000m.
Non-operating: -500 to 40,000 ft.
Relative Humidity
5% to 95%, non-condensing
Drop Test
1.4m from table top to wooden platform, 4
faces.
Operating: Half-sine, 20gpk, 10mS, 3 axes,
6 shocks total
Non-Operating: Half-sine waveform, impact
acceleration of 100G, Pulse duration of 6
mS, Number of shocks: 3 for each of the
three axis
>250,000 hours, full load, 110 & 220Vac
input, 25°C amb., per Telcordia 332 Issue 6.
>7 year life based on calculations at
115Vac/60Hz & 230Vac/50Hz, ambient 25°C
at 24 hrs per day, 365 days/year, 6 power up
cycles per day.
(80% load on 5V, 12V model)
Vibration
Shock
Dimensions
MTBF
Weight
400g
E-Cap Life
www.slpower.com
ME60 Rev 2.9
9-NOV-17
ME60
Medical Series
EMI/EMC Compliance
Conducted Emissions:
Radiated Emissions:
Common Mode Noise:
Electro-Static Discharge (ESD)
Immunity on Power ports:
Radiated RF EM Fields Susceptibility
Electrical Fast Transients (EFT) /Bursts:
Surges, Line to Line (Diff Mode) and Line
to GND (CMN Mode)
Conducted Disturbances induced by RF
Fields
Rated Power frequency magnetic fields
Specification
EN55011/CISPR11 Class B, FCC Part 15.107, Class B: 6db margin typ, at 115 and 230Vac
EN55011/CISPR11 Class B, FCC Part 15.109, Class B: 3db margin typ, at 115 and 230Vac
EN55024/IEC61000-4-5, Level 4, +/-2kV DM, +/-4kV CM, Criteria A
Surpasses IEC60601-1-2, 4
th
Edition requirements.
EN55022/IEC61000-4-6, 3.6V/m – Level 4, 0.15 to 80Mhz; and 12V/m) in ISM and amateur
radio bands between 0.15Mhz and 80Mhz, 80% AM at 1KHz
IEC60601-1-2, 4
th
Edition, Table 5
EN55024/IEC1000-4-8, Level 4: 30 A/m, 50/60 Hz
IEC60601-1-2, 4th Edition, Table 4
EN55024/IECEN61000-4-11: --100% dip for 10 mS, at 0, 45, 90, 135, 180, 225,
270 and 315 degrees, 100% dip for 20mS, 0 deg., Criteria A
--100% dip for 5000mS (250/300 cycles), Criteria B
--60% dip for 100mS, Criteria B
--30% dip for 500mS, Criteria A
IEC60601-1-2, 4th Edition, Table 5
EN55011/EN61000-3-2, Class A
EN61000-3-3
Voltage Interruptions, Dips, Sags &
Surges
Harmonic Current Emissions
Flicker Test
All specifications are typical at nominal input, full load, at 25°C ambient unless noted. Consult factory for information regarding testing for or usage under special
environments.
Mechanical Drawing
See
Note 3:
V through 48V Models: 2.5 x 5.5 x 9.5mm
Barrel Connector, center positive²
Notes:
1) All dimensions in mm.
2) Other options are available.
3) Cable length on 12V through 48V models is 1500mm, nominal.
4) The unit should not be covered or enclosed to protect against excessive case temperature rise.
1 Introduction
In the mid-1960s, American scientist Maas conducted extensive experimental research on the charging process of open-cell batteries and proposed an acceptable charging curve for ...[详细]