The MINT3110 Series are medical grade, high efficiency, triple output power supplies in a small 2” x 4” size. The
MINT3110 family is ideal for medical applications in surgical and laboratory instruments requiring low leakage, multiple
rd
outputs and small footprint. Surpassing the strict requirements of IEC 60601-1 3 edition, the MINT3110 models feature
outputs of 5/±12V, 5/±15V, 5/±24V, Power Fail/DC OK signal, very low earth leakage current of 200µA, Class B
Conducted EMI, 2 MOPP (means of patient protection) isolation, and are available in Class I and Class II configurations.
Model Selection
Model
Number
V1
MINT3110A0508K01
V2
V3
V1
MINT3110A1708K01
V2
V3
V1
MINT3110A1908K01
Notes:
Volts*
5V
12V
-12V
5V
15V
-15V
5V
24V
-24V
Output Current**
w/200LFM air
Convection
14.0A
6.0A
1.0A
14.0A
4.5A
1.0A
12.0A
4.0A
1.0A
10.0A
4.5A
1.0A
10.0A
3.5A
1.0A
8.0A
3.0A
1.0A
Ripple &
Noise***
1.0% pk-pk
1.0% pk-pk
2.0% pk-pk
1.0% pk-pk
1.0% pk-pk
2.0% pk-pk
1.0% pk-pk
1.0% pk-pk
2.0% pk-pk
Total
Regulation
±2%
±3%
±10%
±2%
±3%
±10%
±2%
±3%
±10%
OVP
Threshold
7.5V max.
115%-135%
115%-135%
7.5V max.
115%-135%
115%-135%
7.5V max.
115%-135%
115%-135%
V2
V3
* 5V output is adjustable with +/-10% range
** Total convection power is 80 Watts.
*** Measured with noise probe directly across output terminals, and load terminated with 0.1µF ceramic and 10µF low ESR capacitors. Ripple & Noise of V2 at no
load is 2% maximum.
General Specifications
AC Input
100-240Vac, ±10%, 47-63Hz, 1∅
120–370Vdc
Turn On Time
Less than 2 sec. @115Vac (inversely
proportional to input voltage and thermistor
temperature)
16mS typical at 110W, 120Vac input
Input Current
115Vac: 1.5A, 230Vac: 0.75A
Hold-up Time
MINT3110
110W Triple Output
Medical Series
(continued)
Inrush Current
264Vac, cold start: will not exceed 45A
Signals
Overload
Protection
Short Circuit
Protection
Overvoltage
Protection
Switching
Isolation
Operating
Temperature
Temperature
Derating
Storage
Temperature
Altitude
AC Power Fail, DC OK
Input Fuses
Earth Leakage
Current
Efficiency
F1, F2: 2.5A, 250Vac fuses provided on all
models
150%-300% above rating for V2 & V3, 110%-
200% for V1. Hiccup Mode
Provided - no damage will occur if the output is
shorted.
<290µA@264Vac, 60Hz, NC
87% typical at 230Vac
110W continuous with 200 lfm airflow, 80W
convection cooled – See chart for specific
voltage model ratings.
500µS typ. for return to within 0.5% of
nominal, 50% load step.
∆i/∆t
<0.2A/µS.
Max Volt Deviation = 3%
See models chart
See models chart for trip range.
PFC: Variable 30-400kHz.
Main Converter: Variable 35-180kHz, 65-70kHz
at full load.
Input-Output: 4000Vac, 2 x MOPP
Input-Ground: 1800Vac, 1 x MOPP
Output-Ground: 500Vac
-10° to +70°
Output Power
Transient
Response
Ripple and Noise
Output Voltage
Adjustability
Minimum Load
See models chart
Derate output power linearly above 50° to 50%
at 70°
+/-10% from nominal on 5V output
-40° to +85°
Not required
Operating: -500 to 10,000 ft.
Non-operating: -500 to 40,000 ft.
Total Regulation
See models chart
Operating: 0.003g²/Hz, 1.5grms overall, 3
axes, 10 min/axis
Non-Operating: 0.026g²/Hz, 5.0grms overall,
3 axes, 1 hr/axis
W: 2.0” x L: 4.0” x H: 1.3”
Relative Humidity
5% to 95%, non-condensing
Operating: Half-sine, 20gpk, 10ms, 3 axes, 6
shocks total
Non-Operating: Half-sine, 40 gpk, 10 ms, 3
axes, 6 shocks total
EN/CSA/UL/IEC 60601-1, 3
rd
Edition
Vibration
Shock
Dimensions
Safety Standards
Weight
200g
MTBF
245,000 hours, 25°C Ambient, 110Vac input
Auxiliary Signals
AC Power Fail:
During normal operation, stays HIGH. Signal
goes LOW with at least 6mS warning before
loss of DC output from AC failure.
(5V output, optional) Will compensate for 0.5V
drop min. Will operate without remote sense
connected. Reverse connection protected.
DC OK:
Open collector logic signal goes and stays HIGH
100mS to 500mS after main output reaches
regulation.
Remote Sense:
Notes:
1. Specifications are for convection rating at factory settings at 115 Vac input, 25°C ambient unless otherwise stated.
2. For DC input an external DC safety rated fuse must be used.
www.slpower.com
MINT3110 REV 1.4, 19-SEP-14
MINT3110
110W Triple Output
Medical Series
Connector Information
Input Connector
J100
PIN 1) AC NEUTRAL
PIN 2) EMPTY
PIN 3) AC LINE
Mating Connector:
Molex 09-50-3031
Pins= 08-52-0072
AMP #640250-3
Pins= 3-640706-1
Ground
J101
0.187” FASTON TAB
DC Output Connector
J200
PIN 1) +V1
PIN 2) +V1
PIN 3) GND
PIN 4) GND
PIN 5) GND
PIN 6) GND
PIN 7) V2
PIN 8) V3
Signal Connector
J300
PIN 1) Power Fail/DC OK
PIN 2) GND
PIN 3) +Remote Sense
PIN 4) –Remote Sense
Mating Connector:
Amp #1375820-4
Pin= Amp #1375819
Mating Connector:
Molex 01-90020001
Mating Connector:
Amp #640250-8
Pins=Amp #3-640706-1
Isolation Specifications
Insulation Safety Rating
Conditions/Description
Input/Ground
Input/Output
Output/Ground
Input/Ground
Input/Output
Output/Ground
Min
Nom
Units
Basic (1 MOPP)
Reinforced (2 MOPP)
Operational
1800
4000
Vac
Vac
Vac
Electric Strength Test Voltage
Leakage Current
Earth Leakage Current
Touch Current
Normal Condition (NC)
Single Fault Condition (SFC)
290µA
420µA
90µA
170µA
Input Specifications
All specifications apply over specified input voltage, output load, and temperature range, unless otherwise noted.
Input Voltage
Input Frequency
Input Current
Input Current
Inrush Current
Min
90
47
Nom
115/230
50/60
63
Units
Vac
Hz
115Vac/max load
230Vac/max load
264Vac, cold start, 25°C
Vi
nom,
Io
nom
MINT3110A0508K01
MINT3110A1708K01
MINT3110A1908K01
45
Efficiency
87%
Output Specifications
Output Current V1
Output Current V2
Output Current V3
Output Current V1
Output Current V2
Output Current V3
Output Current V1
Output Current V2
Output Current V3
MINT3110A0508K01
Min
Nom
10.0
4.5
3.5
3.0
6.0
4.5
4.0
Units
ADC
MINT3110A1708K01
ADC
MINT3110A1908K01
ADC
www.slpower.com
MINT3110
110W Triple Output
Medical Series
(continued)
Static Line Regulation V1
Static Line Regulation V2
Static Line Regulation V3
Static Load Regulation V1
(Droop Characteristic)
Static Load Regulation V2
Static Load Regulation V3
Hold-Up Time
Dynamic Load Regulation V1, V2, V3
Start-Up Time
Ripple & Noise V1
Ripple & Noise V2
Ripple & Noise V3
Vi
min-Vi max,
Vi
nom,
0-100%
Io1
max
Vi
min-Vi max,
Vi
nom,
0-100%
Io2
max
Vi
min-Vi max,
Vi
nom,
0-100%
Io3
max
Vi
min-Vi max,
Vi
nom,
0-100%
Io1
max
Vi min-Vi max, Vi nom, 0-100% Io2 max
Vi min-Vi max, Vi nom, 0-100% Io3 max
Vin
= 120Vac,
Po
= 110W
Load change =50%, di/dt = 0.2A/µS
Vin
= 115Vac
Io
nom
20MHz bandwidth
20MHz bandwidth
20MHz bandwidth
Min
-2
-3
-10
-2
-3
-10
Nom
Units
Vo
nom
Vo
nom
Vo
nom
Vo
nom
% Vo nom
% Vo nom
mS
% Vo nom
% Vo nom
% Vo nom
% Vo nom
1%
1%
2%
Protection
Input Fuse
Input Transient Protection
Short Circuit Protection
Overload Protection
Overvoltage Protection
T2.5A/250V internal fuse in both line & neutral
2KV(CM) and 1KV(DM) surge
Min
Nom
Hiccup Mode
Hiccup Mode
Units
KV (CM)
Not user accessible
Latching Type, recycle AC input to reset
See models chart for trip ranges
Characteristic Curves
Output vs. Temperature
80W convection cooled and 110W continuous with 200 LFM airflow. Derate output power to 50% at 70°C.
射频识别(Radio Frequency Identification,RFID)技术是一种利用无线射频通信实现的非接触式自动识别技术,与目前广泛采用的条形码技术相比,RFID具有容量大、识别距离远、穿透能力强、抗污性强等特点。RFID技术已经发展得比较成熟并获得了大规模商用,但超高频RFID技术相对滞后。本文分析了射频芯片nRF9E5的功能特性,并将其用于RFID系统中,设计了一套有源超高频(...[详细]