715
Volt/mA Calibrator
Instruction Sheet
Introduction
The Fluke 715 Volt/mA Calibrator is a source and measurement
tool for 0 to 24 mA current loop testing and dc voltage from 0 to
20/25 V. The calibrator does not source and measure
simultaneously.
Your calibrator is supplied with a holster, a set of test leads, an
installed 9 V alkaline battery, and this instruction sheet.
Summary of Calibrator Capabilities
Function
dc mV input
0 to 200 mV
dc mV output
dc V input
dc V output
dc mA input
0 to 24 mA
dc mA output
Loop power
output
24 V dc output
N/A
0.001 mA
0 to 25 V
0.001 V
0 to 20 V
0.01 mV
Range
Resolution
PN 650314 July 1997 Rev. 3, 8/05
1997-2005
Fluke Corporation. All rights reserved. Printed in U.S.A.
All product names are trademarks of their respective companies.
If the calibrator is damaged or something is missing, contact the
place of purchase immediately. Contact your Fluke distributor for
information about accessories. To order replacement parts or
spares, see “Replacement Parts.”
To contact Fluke, call one of the following telephone numbers:
USA and Canada: 1-888-99-FLUKE (1-888-993-5853)
Europe: +31 402-675-200
Japan: +81-3-3434-0181
Singapore: +65-
-276-6196
Anywhere in the world: +1-425-356-5500
Address correspondence to:
Fluke Corporation
P.O. Box 9090,
Everett, WA 98206-9090
U.S.A.
Fluke Europe B.V.
P.O. Box 1186,
5602 BD Eindhoven
The Netherlands
Visit us on the World Wide Web at:
www.fluke.com
International Symbols
Meaning
Earth ground
Fuse
Battery
Refer to this instruction sheet for information about
this feature.
Double insulated
Conforms to relevant Canadian Standards
Association directives.
Conforms to European Union directives
Safety Information
Use the calibrator only as specified in this sheet, otherwise the
protection provided by the calibrator may be impaired.
Warning
identifies conditions and actions that pose hazard(s)
to the user; a
Caution
identifies conditions and actions that may
damage the calibrator or the equipment under test.
Warning
To avoid possible electric shock or personal injury:
Never apply more than 30 V between any two
jacks (terminals), or between any jack and earth
ground.
Make sure the battery door is closed and
latched before you operate the calibrator.
Remove test leads from the calibrator before
you open the battery door.
Do not operate calibrator if it is damaged.
Do not operate the calibrator around explosive
gas, vapor, or dust.
When servicing the calibrator, use only
specified replacement parts.
Caution
Use the proper jacks, function, and range for your
measurement or output application.
Turning the Calibrator On
Press the green
pushbutton to turn the calibrator on and off.
For longest battery life:
For mA output, use current simulate mode instead of source
mode when an external 24 to 30 V loop power supply is
available. (See Simulating a Transmitter.)
Turn off the calibrator when not in use.
Auto Shut-Off (Power Saver) Feature
The Calibrator automatically turns off after 30 minutes of
inactivity. To reduce this time or disable this feature:
1.
With the Calibrator OFF, press the
key.
P.S.xx
is
displayed, where
xx
is the turn-off time in minutes.
OFF
means the power saver is disabled.
Press
and/or
to increase or decrease the turn-
To disable, press
until the display shows
OFF.
TM
2.
3.
HART
Resistor Mode
The calibrator has a user selectable 250 ohm HART
TM
resistor to
facilitate use with HART
TM
communication devices. The resistor
can be switched in/out at any time by simultaneously pressing
TM
the
and
keys. Use a HART communicator when
measuring dc mA with loop power or sourcing mA.
Span Check Feature
The calibrator allows the users to store 0% and 100% setpoints
for each output function. Once setpoints are stored, the span
check feature allows the user to quickly toggle back and forth
from 0% to 100% or to step in 25% increments.
Automatic step and ramp modes can be enabled while in the
span check mode by simultaneously pressing the
keys.
First select the desired output mode (V,mV or mA) then preceed
to store the setpoints:
1.
2.
3.
4.
5.
Storing 0% and 100% Setpoints
Using the
controls set the output to the desired
output value for 0%.
Press the
the 0% value.
scroll keys simultaneously to store
Now using the
desired output value for 100%
Again, press the
store the 100% value.
Measuring dc Volts
INPUT
mV or V depending
on range
mV
Press so that
INPUT is on
the display
mV
Press to toggle
mV and
V ranges
it01i.eps
Sourcing dc Volts
OUTPUT
mV or V depending
on range
mV
Press so that
OUTPUT is on
the display
mV
Press to toggle
mV and V ranges
it02i.eps