ScopeMeter
120 and 190 Series
ScopeMeter 215C,
225C and 190C Series:
Speed, performance
and analysis power
QU
AL
ANAGEMENT
YM
IT
incl. 225C and 215C
Technical Data
1,000 V CAT II and 600 V
CAT III safety certified
Up to 1,000 V independently
floating isolated inputs
5000 count DMM and paperless
recorder built-in
Four hours rechargeable
NiMH battery pack
IF
ED
TO
MEET
IS
For demanding applications, the
ScopeMeter 215C, 225C and 190C
Series high-performance oscillo-
scopes offer specifications usually
found on top-end bench instru-
ments. With up to 200 MHz band-
width, 2.5 GS/s real time sampling
and a deep memory of 27,500
points per input, they’re ideal
for engineers who need the full
capabilities of a high-performance
scope in a handheld, battery pow-
ered instrument.
Dual input - 200, 100 or
60 MHz bandwidth
Up to 2.5 GS/s real time
sampling per input
Bus Health Test capability for
industrial buses (225C and 215C)
High waveform resolution of
3000 datapoints per channel
Frequency Spectrum using FFT
analysis
Connect-and-View
automatic
triggering, a full range of manual
trigger modes plus external trig-
gering
Digital Persistence for analyzing
complex, dynamic signals like on
an analog oscilloscope
Fast display update rate for
seeing dynamic behavior
instantaneously
Automatic capture and replay of
100 screens
27,500 points and more per
input record length using
ScopeRecord
mode
EM
ST
90
CE
01
RT
ScopeMeter 120 Series:
Three-in-one simplicity
The compact ScopeMeter 120 Series
is the rugged solution for industrial
troubleshooting and installation
applications. It’s a truly integrated
test tool, with oscilloscope, multi-
meter and “paperless” recorder in
one affordable, easy-to-use instru-
ment. Find fast answers to problems
in machinery, instrumentation,
control and power systems.
A dual input 40 MHz or 20 MHz
digital oscilloscope
Two 5,000 counts true-rms
digital multimeters
Cursor measurements
(Fluke 124, 125)
A dual input TrendPlot recorder
Connect-and-View trigger
simplicity for hands-off operation
Shielded test leads for oscillo-
scope, resistance, continuity
and capacitance measurements
Full bandwidth, VPS40 10:1
40 MHz probe included standard
with Fluke 124, 125
Up to seven hours battery
operation
600 V CAT III safety certified
Optically isolated RS-232
interface
Rugged, compact case
Bus Health test for industrial bus
systems (Fluke 125)
Power measurements and har-
monics measurement (Fluke 125)
Technical specifications
215C, 225C and 190C Series
Oscilloscope mode
Vertical deflection
Bandwidth
Rise time
Fluke 225C
Fluke 199C
200 MHz
1.7 ns
Fluke 215C
Fluke 196C
100 MHz
3.5 ns
Fluke 192C
60 MHz
5.8 ns
Horizontal
Pulse width triggering:
Pulse width qualified by time.
Allows for triggering <t, >t, =t,
≠t,
where t is select-
able in minimal steps of 0.01 div. or 50 nsec.
Time delay:
One full screen of pre-trigger view or up
to 100 screens (= 1200 divisions) of post-trigger delay
Dual slope triggering:
Triggers on both rising and
falling edges alike
N-cycle triggering:
Triggers on N-th occurrence of a
trigger event; N to be set in the range 2 to 99
Bandwidth limiter:
User selectable: 10 kHz, 20 MHz
or off
Number of inputs:
Two plus external trigger.
All inputs references isolated from each other and
ground.
Input coupling:
AC or dc, with ground level indicator
Input sensitivity:
2 mV/div to 100 V/div
Normal/invert:
On both input channels; switched
separately
Variable attenuator:
Variable gain on input channel A
Input voltage:
1000 V CAT II, 600 V CAT III rated—see
“general specifications” for further details
Vertical resolution:
8 bit
Accuracy:
± (1.5 % of reading + 0.04 x range/div)
Input impedance:
1 MW ± 1 % // 15 pF ± 2 pF
Fluke 225C
Fluke 199C
2.5 GS/s
5 ns/div
to 5 s/div
Fluke 215C
Fluke 196C
1 GS/s
Fluke 192C
500 MS/s
10 ns/div
to 5 s/div
Automatic capture of 100 screens
The instrument ALWAYS memorizes the last 100
screens (no user setup required). When an anomaly
occurs on screen, the REPLAY button can be pressed
to review the full screen sequence over and over.
Instrument can be set up for triggering on glitches or
intermittent anomalies and will operate in “baby-sit”
mode capturing 100 events.
Replay:
Manual or continuous replay. Displays the
captured 100 screens as a “live” animation, or
under manual control. Each screen has date and
time-stamp.
Replay storage:
Up to 2 sets of 100 screens each
can be saved for later recall and analysis
Maximum
real-time
sample rate
Number of
digitizers
Time base
range
FFT – Frequency spectrum analysis
Shows frequency content of oscilloscope waveform
using Fast Fourier Transform
Window:
Automatic, Hamming, Hanning or None
Automatic window:
Digitally re-samples acquired
waveform to get optimum frequency resolution in FFT
resultant
Vertical scale:
Linear/logarithmic, in volts
Frequency axis:
Logarithmic; frequency range
automatically set as function of timebase range of
oscilloscope
Maximum record length:
3000 points per input in
Scope mode; 27,500 points per input in ScopeRecord
roll mode (5 ms/div to 2 min/div)
Accuracy:
± (0.01 % of reading + 1 pixel)
Glitch capture:
50 nsec (5 μsec/div to 1 min/div)
Display and acquisition
Display:
144 mm full-color LCD, with backlight
Display modes:
Input A, input B, dual, average,
Replay
Visible screen width:
12 divisions
Waveform mathematics:
A + B, A - B, A * B, all
with user selectable scaling of resultant; A versus B
(X - Y mode); Frequency spectrum using FFT analysis
Acquisition modes:
Normal, auto, single shot,
ScopeRecord
roll, glitch capture, waveform
compare, waveform compare with automatic
“Pass / Fail testing”, Bus Health test mode (225C and
215C only), Eyepattern Display of single ended or
differential bus signal (Fluke 225C and 215C only).
Waveform compare and pass/fail testing
Waveform compare:
Provides storage and display
of a reference waveform for visual comparison with
newly acquired waveforms. Reference is derived from
an acquired waveform and can be modified in the
ScopeMeter or externally using FlukeView
Software.
Pass/Fail Testing:
In waveform compare mode, the
Color Scopemeter can be set up to store only matching
(“Pass”) or only non-matching (“Fail”) acquired wave-
forms in the replay memory bank for further analysis.
Automatic scope measurements
V dc, V ac rms, V ac+dc, Vpeak max, Vpeak min, Vpeak
to peak, A ac, A dc, A ac+dc, frequency (Hz), risetime,
falltime, power factor, Watts, VA, VA reactive, phase,
pulse width (pos/neg), dutycycle (pos/neg), temperature
°C, temperature °F, dBV, dBm into 50
and 600
Vpwm ac, Vpwm ac+dc for measurement on pulse
width modulated motordrives and frequency inverters
Trigger and delay
Source:
Input A, input B, external trigger input.
All input references isolated from each other and
from ground
Modes:
Automatic Connect-and-View
free run, single
shot, edge, delay, video, video line, selectable
pulsewidth, dual slope, N-cycle
Connect-and-View
Advanced automatic triggering
that recognizes signal patterns, automatically sets up
and continuously adjusts triggering, time base and
amplitude. Automatically displays stable waveforms
of complex and dynamic signals like motor drive and
control signals. Can be switched off if desired.
Video triggering:
NTSC, PAL, PAL+, SECAM. Includes
field 1, field 2 and line select
Cursor measurements
Source:
Input A, input B or the mathematical result
trace (excluding A vs B curve)
Dual horizontal lines:
Voltage at cursor 1 and 2,
voltage between cursors
Dual vertical lines:
Time between cursors, 1/T
between cursors (in Hz), voltage between markers,
risetime with markers, falltime with markers; Vrms
between cursors, Watts between cursors
Single vertical line:
Min-Max and average voltage at
cursor position; frequency and rms-value of individual
frequency component in FFT result
2 Fluke Corporation
ScopeMeter 215C, 225C and 190C Series, 120 Series
Zoom
Recorded timespan:
Up to 22 days with a resolution
of 1 minute
Up to 16x horizontal zoom
Recording mode:
Continuous roll for the duration of
the full recordable timespan
Bus health test mode
Measurement speed:
5 measurements per second
Horizontal scale:
Time from start, time of day
Zoom:
Up to 64x zoom
Bus Health automatically analyzes the electrical signals
Memory:
Up to 2 TrendPlot recordings can be saved
on the industrial bus system to measure individual
parameters and to give waveform information.
Automatically compares the measurement results to
Cursor measurements—all recorder modes
preset values and present ‘good’, ‘weak’ or ’false’
Source:
Input A, B or DMM input
indicator with each parameter.
Min-Max or Average
Bus types and reference standards used:
voltage. Time between cursors
AS-i (EN50295, 166 kb/s);
Min-Max or Average
CAN-bus (ISO-11898, up to 1 Mb/s);
voltage. Absolute date and time or time from start
Modbus (EIA-232 up to 115 kb/s and EIA-485
up to 10 Mb/s);
Meter mode
Foundation Fieldbus H1 (61158 type 1, 31.25 kb/s);
Profibus DP (EIA-485 up to 10 Mb/s) and PA (61158
Via 4 mm banana inputs. Fully isolated from scope
inputs and scope ground. The specified accuracy is
Ethernet [10Base2 (coaxial) and 10BaseT (UTP)],
valid over the temperature range 18 °C to 28 °C
10 Mb/s;
(65 °F to 82 °F). Add 10 % of specified accuracy for
Ethernet 100BaseT (100 Mb/s);
each degree C below 18 °C or above 28 °C.
RS-232 (EIA-232, up to 115 kb/s);
Maximum resolution:
5,000 counts
RS-485 (EIA-485, up to 10 Mb/s).
Voltmeter ranges:
500 mV, 5 V, 50 V, 500 V, 1,000 V
Measured parameters (where applicable):
Accuracy:
Bias voltage level, signal amplitude, pulse width or
V dc ± (0.5 % + 5 counts)
baud rate, risetime, fall time, jitter, signal distortion,
V ac true rms
noise HF, noise LF, in-band noise.
15 Hz to 60 Hz: ± (1 % + 10 counts)
60 Hz to 1 kHz: ± (2.5 % + 15 counts)
Recorder mode
V ac+dc true rms
dc to 60 Hz: ± (1 % + 10 counts)
ScopeRecord—roll mode
Dual input waveform storage mode
Ohms:
Source and display:
Ranges: 500
W,
5 kW, 50 kW, 500 kW, 5 MW, 30 MW
Memory depth:
27,500 points per input. Each point
Accuracy: ± (0.6 % + 5 counts)
consists of Min-Max pair
Min-Max values:
Min-Max values are measured at
Other meter functions
high sample rates ensuring capture and display of
Continuity:
Beeper on < 50
(± 30
W)
glitches
Diode test:
Up to 2.8 V
Time base range
5 ms/div to 1 min/div
2 min/div
Amps:
A dc, A ac, A ac+dc using an optional current
clamp or shunt. Scaling factors: 0.1 mV/A ... 100 V/A
Recorded timespan
6 sec to 24 hr
48 hr
Temperature (°C, °F):
With optional accessories. Scale
Glitch capture
50 ns
250 ns
factors 1 mV/°C or 1 mV/°F
Sample rate
20 MS/s
4 MS/s
Input impedance:
1 MW ± 1 %/10 pF ± 2 pF
Resolution
200 μsec to 2 sec
4.8 sec
Advanced meter functions:
Auto/manual ranging,
relative measurements (zero reference), TrendPlot
Recording modes:
Single sweep, continuous roll,
recording
Start-on-Trigger (through external), Stop-on-Trigger
(through external)
General specifications
Stop-on-Trigger (through external):
ScopeRecord
mode can be stopped by an individual trigger event, or
Input voltage ratings
by an interruption of a repetitive trigger signal
Horizontal scale:
Maximum probe voltage:
CAT II 1,000 V, CAT III
Zoom:
Up to 100x
600 V (Maximum voltage between 10:1 probe tip
Memory:
Up to 2 dual input ScopeRecord waveforms
(VPS210) and reference lead)
Floating voltage:
CAT II 1,000V, CAT III 600 V
(Maximum voltage between earth ground and any
TrendPlot
recording
terminal (signal input or shielding))
Single or dual input electronic paperless chart
Independently isolated inputs:
CAT II 1,000 V,
recorder. Plots, displays and stores meter and scope
CAT III 600 V (Maximum voltage between any terminal
measurements.
of one input or probe (VPS210) and any other terminal
Source and display:
Input A, input B or DMM input
of another input or probe (VPS210))
Memory depth:
18,000 points recording. Per record
Maximum voltage on BNC input directly
point a minimum, a maximum and an average value,
(input A or B):
CAT III 300 V
plus a date and timestamp are recorded.
Maximum voltage on meter input:
CAT II 1,000 V,
Ranges:
5 s/div to 30 min/div in normal view mode;
CAT III 600 V
5 min/div to 48 hr/div in view all mode, giving
overview of total record
Fluke Corporation 3
Memory save and recall
Scope memories:
15 memory locations that each
can contain two waveforms plus corresponding setup.
With each storage action, a user specified name
(20 ASCII characters long) can be assigned to the
stored data, for easier reference.
Recorder memories:
2 memory locations that each
can contain 100 captured dual input scope screens,
or a dual input ScopeRecord (27,500 Min-Max pairs
per input), or a dual input Trendplot (18,000 min-max
pairs + average values)
CAT II 1000 V/ CAT III 600 V
(Maximum voltage between 10:1 probe tip [VPS200]
and reference lead)
Floating voltage:
terminal [signal input or shielding])
(Maximum voltage between any terminal of one input
or probe [VPS200] and any other terminal of another
input or probe [VPS200])
(input A or B):
CAT III 300 V
CAT II 1,000 V/ CAT III 600 V
Real-time clock
Time and date stamp for ScopeRecord, 100 captured
screens and TrendPlots
Mechanical data
Size:
256 mm x 169 mm x 64 mm
(10.1 in x 6.6 in x 2.5 in)
Weight:
2 kg (4.4 lbs)
Environmental
Operating temperature:
0 °C to +50 °C
Storage temperature:
-20 °C to +60 °C
Humidity:
10 °C to 30 °C: 95 % RH non condensing
30 °C to 40 °C: 75 % RH non condensing
40 °C to 50 °C: 45 % RH non condensing
Maximum operating altitude:
3,000 m (10,000 feet)
Maximum storage altitude:
12 km (40,000 feet)
Electro-Magnetic-Compatibility (EMC):
EN 61326-1
for emission and immunity
Case
Design:
Rugged, shock proof with integrated
protective holster
Drip and dust proof:
IP51 according to IEC529
Shock and vibration:
Shock 30g, Vibration
(sinusoidal) 3g according to MIL-PRF-28800F Class 2
Display:
Bright full-color LCD with backlight
Brightness:
80 Cd/m
typ. using power adapter
Display size:
115.2 mm x 86.4 mm (4.54 in x 3.4 in);
144 mm (5.67 in) diagonal
Resolution:
320 x 240 pixels
Contrast and brightness:
User adjustable,
temperature compensated
Brightness:
80 cd/m
Optically-isolated PC/printer interface
To printer:
Supports HP Laserjet,
DeskJet,
Epson FX/LQ, Seiko DPU-414 and Postscript printers
via optional PAC 91
To PC:
Transfer instrument settings, screen images
and waveform data, compatible with FlukeView
software for Windows
via optional OC4USB or
PM9080.
Power
Line power:
Country-specific line voltage
adapter/battery charger included
Battery power:
Rechargeable NiMH (installed)
Battery operating time:
4 hours
Battery charging time:
4 hours
Battery power saving functions:
Auto power down
with adjustable power down time. On-screen battery
status indicator.
Warranty
Three-years, parts and labor on mainframe instrument
One-year on accessories
Accessories
Standard accessories
Rechargeable battery
pack (installed)
Line voltage adapter/
battery charger
Voltage probes and
accessories
Multimeter test leads
User manual
Fluke 215C, 225C and 190C Series
BP190
BC190
10:1 voltage probe (VPS200, 1 red + 1 grey) including hook clip,
ground lead with mini alligator clip, ground spring for probe tip
TL75 Hard Point test lead set (1 red, 1 black)
Multi-lingual CD-ROM
Safety
Compliance:
EN61010-1-2001, Pollution degree 2;
UL61010B, with approval;
CAN/CSA C22.2, No. 61010-1-04, with approval;
ANSI/ISA-82.02.01
215C and 225C additional accessories
Three breakout adapt-
ers for bus systems
using RJ45, DB9, and
M12 connectors.
BHT190
4 Fluke Corporation
Technical specifications 120 Series
Oscilloscope mode
Vertical deflection
Bandwidth and risetime
Bandwidth (risetime)
with VPS40 probes
Fluke 125, 124
40 MHz
40 MHz
12.5 MHz
8.75 ns
Fluke 123
20 MHz
20 MHz
12.5 MHz
17.5 ns
Cursor measurements (125, 124)
Sources:
Input A, Input B
Modes:
Single or dual vertical cursor, dual horizontal
cursor, risetime or falltime
Measurements:
Single vertical line: Average, min value, max value,
time from start of recording in roll mode
∆V
at markers, time between
cursors, 1/T between cursors (in Hz)
Dual horizontal lines: High, low or
∆V
– readout,
risetime and falltime: transition time, 0 %-level,
100 %-level, with markers at 10 % and 90 %
Accuracy:
As oscilloscope
input A and B directly
with STL120 Shielded
Test Leads
Instrument risetime
(input directly)
Number of inputs:
two
Input coupling:
AC, dc with ground level indicator
Input sensitivity:
5 mV to 500 V/div (with the
included VPS40 (Fluke 125, 124) and STL120 shielded
test leads measure up to 600 Vrms, CAT III)
Input voltage:
600 V CAT III. See “general specifica-
tions” for more detailed information.
8 bit
Accuracy:
± (1 % of reading + 0.05 x range/div)
Input impedance:
1 MW ± 1 % // 225 pF with
STL120 shielded test leads; 1 MW ± 1 % // 20 pF
± 3 pF with BB120; 5 MW ± 1 % // 15.5 pF with
VPS40, 10:1 voltage probe
Bus health tester (Fluke 125 only)
Bus health automatically analyzes the electrical signals
on the network to give waveform data and measure
individual parameters. Automatic comparison of the
measurement results to the standards results in “good”
or “false” indicators to be displayed per parameter.
Interbus S (EIA-485, up to 10 Mb/s)
ControlNet (61158 type 2, 2.5 Mb/s)
Modbus (EIA-232, up to 115 kb/s and
and H2 (61158 type 1, up to 10 Mb/s);
Profibus DP (EIA-485, up to 10 Mb/s) and
PA (61158 type 1, 31.25 kb/s);
Ethernet [10Base2 (coaxial) and 10BaseT (UTP)], 10 Mb/s;
RS-485 (EIA-485, up to 10 Mb/s);
or user defined system
Measured parameters:
Baud rate, risetime, falltime,
high level, low level, distortion, amplitude and jitter,
with comparison to system’s standard values
Horizontal
Max. sample rate (both channels simultaneously):
Fluke 125, 124: 2.5 GS/s for repetitive signals;
25 MS/s for single shot
Fluke 123: 1.25 GS/s for repetitive signals; 25 MS/s
for single shot
Number of digitizers:
two
Time base range:
10 ns/div to 1 min/div
(Fluke 125, 124)
512 Min/Max points
per input
Accuracy:
± (0.1 % of reading + 1 pixel)
Glitch detect:
40 ns
Power measurements (Fluke 125 only)
Measure types:
Watt, VA, VAR, Power Factor (PF)
Power configuration:
Single phase or balanced
3-phase (delta-configuration) mains supply
Voltage measurement:
Channel A, using STL120,
voltage probe or direct input
Current measurement:
Channel B, using i400s
current clamp (included) or other compatible clamp
Current clamp or shunt sensitivity:
0.1/1/10/100/1000 mV/A, 10 mV/mA and 400 mV/A
Display modes:
Input A, input A and B, envelope,
smooth
Acquisition modes:
Normal (including glitch capture),
single shot, roll
Trigger and delay
Source:
Input A, input B, external via optional ITP120
Modes:
Automatic Connect-and-View, Free Run, Edge,
Single Shot, Video, Video Line
Connect-and-View:
that recognizes signal patterns and automatically sets
up and continuously adjusts triggering, time base and
amplitude. Automatically displays stable pictures of
control signals
Video triggering:
line select
Time delay:
Up to 10 divisions pre-trigger view
Harmonics mode (Fluke 125 only)
Converts waveform information into a harmonics dis-
play (using FFT processing), which shows the relative
amplitudes of the 1st up to the 33rd harmonic
Harmonics frequency range:
DC to 33rd harmonic
(fundamental
60 Hz); DC to 24th (fundamental
400 Hz)
Analyzed waveform:
Voltage waveform (Ch.A),
current waveform (Ch.B) or power (Ch.A x Ch.B),
automatically generated
Display:
Bargraph showing 1st up to 33rd harmonic,
amplitude displayed in % relative to fundamental
Timebase setting:
5 ms/div.
Measurements:
Relative amplitude of individual
harmonic; THD in %r or %f
Measurements
Vdc, Vac, Vac+dc, Vpeak max, Vpeak min, Vpeak
to peak, frequency (Hz), positive pulse width, negative
pulse width, positive duty cycle, negative duty
cycle, Amp ac, Amp dc, Amp ac+dc, Phase,
Temperature °C, Temperature °F, dBV, dBm into 50
and 600
W.
(Amps, °C or °F with optional probes)