2182a
Nanovoltmeter
The two-channel Model 2182A Nanovoltmeter
is optimized for making stable, low noise
voltage measurements and for characterizing
low resistance materials and devices reliably
and repeatably . It provides higher measurement
speed and significantly better noise performance
than alternative low voltage measurement
solutions .
The Model 2182A represents the next step
forward in Keithley nanovoltmeter technology,
replacing the original Model 2182 and offering
enhanced capabilities including pulse capability,
lower measurement noise, faster current rever-
sals, and a simplified delta mode for making
resistance measurements in combination with a
reversing current source, such as the Model 6220
or 6221 .
Make low noise measurements at
high speeds, typically just 15nV
p-p noise at 1s response time,
40–50nV p-p noise at 60ms
Delta mode coordinates
measurements with a reversing
current source at up to 24Hz
with 30nV p-p noise (typical) for
one reading. averages multiple
readings for greater noise
reduction
synchronization to line provides
110db NMrr and minimizes
the effect of aC common-mode
currents
Dual channels support measuring
voltage, temperature, or the ratio
of an unknown resistance to a
reference resistor
built-in thermocouple
linearization and cold junction
compensation
flexible, Effective speed/Noise Trade-offs
The Model 2182A makes it easy to choose the best speed/filter combination for a particular applica-
tion’s response time and noise level requirements . The ability to select from a wide range of response
times allows optimizing speed/noise trade-offs . Low noise levels are assured over a wide range of
useful response times, e .g ., 15nV p-p noise at 1s and 40-50nV p-p noise at 60ms are typical .
Figure 1
illustrates the Model 2182A’s noise performance .
nV
Keithley 2182A
nV/µΩ Meter
-100
Number of Readings
figure 1. Compare the Model 2182a’s DC noise performance with a nanovolt/micro-ohm-
meter’s. all the data shown was taken at 10 readings per second with a low thermal short
applied to the input.
1.888.KEITHLEY
(U.S. only)
www.keithley.com
A Greater Measure of Confidence
LOW LEVEL MEASURE & SOURCE
-50
Low noise measurements for research, metrology, and other low voltage testing applications
2182A
Nanovoltmeter
Reliable Results
Power line noise can compromise measurement accuracy significantly at the nanovolt level. The
Model 2182A reduces this interference by synchronizing its measurement cycle to line, which
minimizes variations due to readings that begin at different phases of the line cycle. The result is
exceptionally high immunity to line interference with little or no shielding and filtering required.
Optimized for Use with Model 6220/6221 Current Sources
Device test and characterization for today’s very small and power-efficient electronics requires sourc-
ing low current levels, which demands the use of a precision, low current source. Lower stimulus
currents produce lower—and harder to measure—voltages across the devices. Linking the Model
2182A Nanovoltmeter with a Model 6220 or 6221 Current Source makes it possible to address both of
these challenges in one easy-to-use configuration.
When connected, the Model 2182A and Model 6220 or 6221 can be operated like a single instrument.
Their simple connections eliminate the isolation and noise current problems that plague other solu-
tions. The Model 2182A/622X combination allows making delta mode and differential conductance
measurements faster and with less noise than the original Model 2182 design allowed. The Model
2182A will also work together with the Model 6221 to make pulse-mode measurements.
The 2182A/622X combination is ideal for a variety of applications, including resistance measure-
ments, pulsed I-V measurements, and differential conductance measurements, providing significant
advantages over earlier solutions like lock-in amplifiers or AC resistance bridges. The 2182A/622X
combination is also well suited for many nanotechnology applications because it can measure
resistance without dissipating much power into the device under test (DUT), which would otherwise
invalidate results or even destroy the DUT.
An Easy-to-Use Delta Mode
Keithley originally created the delta mode method for measuring voltage and resistance for the
Model 2182 and a triggerable external current source, such as the Model 2400 SourceMeter
SMU
instrument. Basically, the delta mode automatically triggers the current source to alternate the signal
polarity, and then triggers a nanovoltmeter reading at each polarity. This current reversal technique
Ordering Information
2182A
Nanovoltmeter
Accessories Supplied
2107-4 Low Thermal Input Cable
with spade lugs, 1.2m (4 ft).
User manual, service
manual, contact cleaner,
line cord, alligator clips.
ACCESSORIES AVAILAbLE
2107-30
2182-KIT
2187-4
2188
4288-1
4288-2
7007-1
7007-2
7009-5
8501-1
8501-2
8503
KPCI-488LPA
KUSB-488B
Low Thermal Input Cable with spade lugs,
9.1m (30 ft)
Low Thermal Connector with strain relief
Low Thermal Test Lead Kit
Low Thermal Calibration Shorting Plug
Single Fixed Rack Mount Kit
Dual Fixed Rack Mount Kit
Shielded GPIB Cable, 1m (3.2 ft)
Shielded GPIB Cable, 2m (6.5 ft)
Shielded RS-232 Cable, 1.5m (5 ft)
Trigger Link Cable, 1m (3.2 ft)
Trigger Link Cable, 2m (6.5 ft)
Trigger Link Cable to 2 male BNC connectors
IEEE-488 Interface/Controller for the PCI Bus
IEEE-488 USB-to-GPIB Interface Adapter
SERVICES AVAILAbLE
1-year factory warranty extended to 3 years
from date of shipment
C/2182A-3Y-ISO 3 (ISO-17025 accredited) calibrations within 3
years of purchase*
* Not available in all countries
2182A-3Y-EW
5nV
AppLICATIONS
Research
• Determining the transition
temperature of superconductive
materials
• I-V characterization of a material
at a specific temperature
• Calorimetry
• Differential thermometry
• Superconductivity
• Nanomaterials
Metrology
• Intercomparisons of standard cells
• Null meter for resistance bridge
measurements
4µV
DC
Measurement
Delta Mode
Measurement
Figure 2. Results from a Model 2182A/6220 using the delta mode to measure a 10mW resistor
with a 20µA test current. The free Model 6220/6221 instrument control example start-up
software used here can be downloaded from www.keithley.com.
1.888.KEITHLEY
(U.S. only)
www.keithley.com
2182a
Nanovoltmeter
cancels out any constant thermoelectric offsets, so the results reflect the
true value of the voltage being measured . The improved delta mode for
the Model 2182A and the Model 622X current sources uses the same basic
technique, but the way in which it’s implemented has been simplified dra-
matically . The new technique can cancel thermoelectric offsets that drift
over time (not just static offsets), produces results in half the time of the
original technique, and allows the current source to control and configure
the Model 2182A . Two key presses are all that’s required to set up the
measurement . The improved cancellation and higher reading rates reduce
measurement noise to as little as 1nV .
Differential Conductance Measurements
Characterizing non-linear tunneling devices and low temperature devices
often requires measuring differential conductance (the derivative of a
device’s I-V curve) . When used with a Model 622X current source, the
Model 2182A is the industry’s fastest, most complete solution for differen-
tial conductance measurements, providing 10X the speed and significantly
lower noise than other instrumentation options . There’s no need to
average the results of multiple sweeps, because data can be obtained in a
single measurement pass, reducing test time and minimizing the potential
for measurement error .
Pulsed Testing with the Model 6221
When measuring small devices, introducing
even tiny amounts of heat to the DUT can raise
its temperature, skewing test results or even
destroying the device . When used with the
Model 2182A, the Model 6221’s pulse capability
minimizes the amount of power dissipated into
a DUT . The Model 2182A/6221 combination
synchronizes the pulse and measurement . A
measurement can begin as soon as 16µs after
the Model 6221 applies the pulse . The entire
pulse, including a complete nanovolt measure-
ment, can be as short as 50µs .
Model 2182A
2182A
NANOVOLTMETER
RS-232
Trigger Link
Model 622X
6220
DC AND AC CURRENT SOURCE
GPIB or
Ethernet
DUT
figure 3. It’s simple to connect the Model 2182a to the Model
6220 or 6221 to make a variety of measurements. The instrument
control example start-up software available for the Model 622X
current sources includes a step-by-step guide to setting up the
instrumentation and making proper connections.
Competition
2182a
2182a in delta mode
100µs
0.5µA
figure 4. The Model 2182a produces the lowest transient currents of any nanovoltmeter available.
In the delta, differential conductance, and pulse
modes, The Model 2182A produces virtually no transient currents, so it’s
ideal for characterizing devices that can be easily disrupted by current spikes
(see
Figure 4) .
Metrology applications
The Model 2182A combines the accuracy of a digital multimeter with
low noise at high speeds for high-precision metrology applications . Its
low noise, high signal observation time, fast measurement rates, and
2ppm accuracy provide the most cost-effective meter available today for
applications such as intercomparison of voltage standards and direct
measurements of resistance standards .
Nanotechnology applications
The Model 2182A combined with the Model 622X current source or Series
2400 SourceMeter
SMU instrument is a highly accurate and repeatable
solution for measuring resistances on carbon nanotube based materials
and silicon nanowires .
research applications
The Model 2182A’s 1nV sensitivity, thermoelectric EMF cancellation, direct
display of “true” voltage, ability to perform calculations, and high measure-
ment speed makes it ideal for determining the characteristics of materials
such as metals, low resistance filled plastics, and high and low temperature
superconductors .
1.888.KEITHLEY
(U.S. only)
www.keithley.com
2182a
Nanovoltmeter
220
215
210
205
190
185
180
(nV)
Temperature
(°C)
–5
–10
25 33 42 50 58 67 75 83 92 100 108 117 125
Minutes
Three Ways to Measure Nanovolts
DC nanovoltmeters.
DC nanovoltmeters
and sensitive DMMs both provide low noise
DC voltage measurements by using long
integration times and highly filtered readings
to minimize the bandwidth near DC .
Unfortunately, this approach has limitations,
particularly the fact that thermal voltages
develop in the sample and connections vary,
so long integration times don’t improve
measurement precision . With a noise
specification of just 6nV p-p, the Model 2182A
is the lowest noise digital nanovoltmeter
available .
AC technique.
The limitations of the long
integration and filtered readings technique
have led many people to use an AC technique
for measuring low resistances and voltages .
In this method, an AC excitation is applied
to the sample and the voltage is detected
synchronously at the same frequency and
an optimum phase . While this technique
removes the varying DC component, in many
experiments at high frequencies, users can
experience problems related to phase shifts
caused by spurious capacitance or the L/R
time constant . At low frequencies, as the
AC frequency is reduced to minimize phase
shifts, amplifier noise increases .
The current reversal method.
The Model
2182A is optimized for the current reversal
method, which combines the advantages of
both earlier approaches . In this technique,
the DC test current is reversed, then the
difference in voltage due to the difference
in current is determined . Typically, this
measurement is performed at a few hertz (a
frequency just high enough for the current
to be reversed before the thermal voltages
can change) . The Model 2182A’s low noise
performance at measurement times of a
few hundred milliseconds to a few seconds
means that the reversal period can be set
quite small in comparison with the thermal
time constant of the sample and the con-
nections, effectively reducing the impact of
thermal voltages .
figure 5. The Model 2182a’s delta mode provides extremely stable results, even in the pres-
ence of large ambient temperature changes. In this challenging example, the 200nV signal
results from a 20µa current sourced by a Model 6221 through a 10mW test resistor.
optional accessory: Model 2187-4 low Thermal
Test lead Kit
The standard cabling provided with the Model
2182A Nanovoltmeter and Model 622X Current
Sources provides everything normally needed
to connect the instruments to each other and to
the DUT . The Model 2187-4 Low Thermal Test
Lead Kit is required when the cabling provided
may not be sufficient for specific applications,
such as when the DUT has special connection
requirements . The kit includes an input cable
with banana terminations, banana extensions,
sprung-hook clips, alligator clips, needle probes,
and spade lugs to accommodate virtually any
DUT . The Model 2187-4 is also helpful when the
DUT has roughly 1GW impedance or higher .
In this case, measuring with the Model 2182A
figure 6. Model 2187-4 Test lead Kit
directly across the DUT will lead to loading
errors . The Model 2187-4 Low Thermal Test Lead Kit provides a banana cable and banana jack
extender to allow the Model 2182A to connect easily to the Model 622X’s low impedance guard
output, so the Model 2182A can measure the DUT voltage indirectly . This same configuration also
removes the Model 2182A’s input capacitance from the DUT, so it improves device response time,
which may be critical for pulsed measurements .
figure 7. Model 2182a rear panel
1.888.KEITHLEY
(U.S. only)
www.keithley.com
2182a
Volts specifications
(20% over range)
Nanovoltmeter
Input
resistance
>10 GW
>10 GW
>10 GW
>10 GW
10 MW ±1%
>10 GW
>10 GW
>10 GW
accuracy: ±(ppm of reading + ppm of range)
(ppm = parts per million) (e.g., 10ppm = 0.001%)
24 Hour
90 Day
1 Year
2 Year
Cal
±1°C
Cal
±5°C
Cal
±5°C
Cal
±5°C
20 + 4
40 + 4
50 + 4
60 + 4
10 + 3
25 + 3
30 + 4
40 + 5
7+2
18 + 2
25 + 2
32 + 3
2 + 1
25 + 2
32 + 3
10 + 3
25 + 3
35 + 4
52 + 5
10 + 6
7+2
2 + 1
18 + 2
18 + 2
30 + 7
25 + 2
25 + 2
40 + 7
32 + 3
32 + 3
Temperature
Coefficient
0°–18°C & 28°–50°C
(1 + 0 .5)/°C
(1 + 0 .2)/°C
(1 + 0 .1)/°C
(1 + 0 .1)/°C
(1 + 0 .5)/°C
(1 + 1 )/°C
(1 + 0 .5)/°C
(1 + 0 .5)/°C
CONDITIONS:
1PLC with 10 reading digital filter or 5PLC with 2 reading digital filter .
Channel 1
range
10 .000000 mV
2, 3, 4
100 .00000 mV
1 .0000000 V
10 .000000 V
100 .00000 V
Channel 2
6, 10
100 .00000 mV
1 .0000000 V
10 .000000 V
resolution
1 nV
10 nV
100 nV
1 µV
10 µV
10 nV
100 nV
1 µV
* 0 .0028ppm for the 100mV range, 0 .0016ppm for ranges 1V and above .
** 8nV/
Hz
for ranges above 10mV . meas_time (seconds) = pulsewidth – pulse_meas_delay in 33µs incr .
DC Noise Performance
(DC noise expressed in volts peak-to-peak)
Response time = time required for reading to be settled within noise levels from a stepped input, 60Hz operation .
Channel 1
response
range
Time
NPlC, filter
10 mV
100 mV
1V
10 V
25 .0 s
5, 75
6 nV
20 nV
75 nV
750 nV
4 .0 s
5, 10
15 nV
50 nV
150 nV
1 .5 µV
1 .0 s
1, 18
25 nV
175 nV
600 nV
2 .5 µV
667 ms
1, 10 or 5, 2
35 nV
250 nV
650 nV
3 .3 µV
60 ms
1, Off
70 nV
300 nV
700 nV
6 .6 µV
Channel 2
6, 10
25 .0 s
4 .0 s
1 .0 s
85 ms
5, 75
5, 10
1, 10 or 5, 2
1, Off
150 nV
150 nV
175 nV
425 nV
200 nV
200 nV
400 nV
1 µV
750 nV
1 .5 µV
2 .5 µV
9 .5 µV
100 V
75 µV
75 µV
100 µV
150 µV
300 µV
NMrr
110 dB
100 dB
95 dB
90 dB
60 dB
110 dB
100 dB
90 dB
60 dB
CMrr
140 dB
140 dB
140 dB
140 dB
140 dB
140 dB
140 dB
140 dB
140 dB
VolTaGE NoIsE Vs. sourCE rEsIsTaNCE
source
resistance
Noise
6 nV
8 nV
1 kW
15 nV
10 kW
35 nV
100 kW
100 nV
1 MW
350 nV
analog
filter
Off
Off
Off
Off
On
On
Digital
filter
operating Characteristics
13, 14
60Hz (50Hz) operation
function
DCV Channel 1,
Channel 2,
Thermocouple
Digits
7 .5
7 .5
17, 19
6 .5
18, 19
6 .5
18, 19, 20
5 .5
17, 19
4 .5
16, 17, 19
7 .5
7 .5
17, 19
6 .5
18, 20
5 .5
17
4 .5
17
6 .5
readings/s
45
(36)
115 (105)
1 .5 (1 .3)
2 .3 (2 .1)
8 .5 (7 .5)
41
(40)
47 (40 .0)
22
PlCs
0 .01
0 .01
(Displayed in °C, °f, or K. accuracy based on
ITs-90, exclusive of thermocouple errors.)
TYPE
raNGE
–200 to +760°C
–200 to +1372°C
–200 to +1300°C
–200 to +400°C
–200 to +1000°C
0 to +1768°C
0 to +1768°C
+350 to +1820°C
rEsoluTIoN
0 .001 °C
0 .001 °C
0 .001 °C
0 .001 °C
0 .001 °C
0 .1 °C
0 .1 °C
0 .1 °C
TEMPEraTurE (Thermocouples)
aCCuraCY
90 Day/1 Year
23° ±5°C
relative to simulated
reference Junction
±0 .2 °C
±0 .2 °C
±0 .2 °C
±0 .2 °C
±0 .2 °C
±0 .2 °C
±0 .2 °C
±0 .2 °C
Delta with 622X
system speeds
13, 15
RANGE CHANGE TIME:
<40 ms (<50 ms) .
FUNCTION CHANGE TIME:
<45 ms (<55 ms) .
AUTORANGE TIME:
<60 ms (<70 ms) .
ASCII READING TO RS-232 (19.2K Baud):
40/s
(40/s) .
MAX. INTERNAL TRIGGER RATE:
120/s
(120/s) .
MAX. EXTERNAL TRIGGER RATE:
120/s
(120/s) .
1.888.KEITHLEY
(U.S. only)
www.keithley.com
Channel 1/Channel 2 (Ratio),
Delta with 24XX, Scan
Model 2182A specifications
CHANNEL 1/CHANNEL 2 RATIO:
For input signals
≥1%
of the range, Ratio Accuracy =
±{[Channel 1 ppm of Reading + Channel 1 ppm of Range * (Channel 1 Range/Channel 1 Input)] + [Channel 2 ppm of Reading + Channel 2 ppm of Range * (Channel 2 Range/Channel 2 Input)]} .
DELTA (hardware-triggered coordination with Series 24XX, Series 26XXA, or Series 622X current sources for low noise R measurement):
Accuracy = accuracy of selected Channel 1 range plus accuracy of I source range .
DELTA MEASUREMENT NOISE WITH 6220 or 6221:
Typical 3nVrms / Hz (10mV range)
21
. 1Hz achieved with 1PLC, delay = 1ms, RPT filter = 23 (20 if 50Hz) .
PULSE-MODE (WITH 6221):
Line synchronized voltage measurements within current pulses from 50µs to 12ms, pulse repetition rate up to 12Hz .
PULSE MEASUREMENT NOISE (typical rms noise, R
DUT
<10W):
±(0 .009ppm of range*) / meas_time / pulse_avg_count + 3nV** / (2 · meas_time · pulse_avg_count) for 10mV range .