7158
Low Current Scanner Card
10-channel
The Model 7158 provides quality low-current
switching at an affordable price . The offset
current error generated is specified <1pA, with
typical performance at <30fA . When used with a
voltage source and electrometer or picoammeter,
this card can easily automate insulation resistance
tests, reverse leakage tests on semiconductor
junctions, or gate leakage tests on FETs .
CONTACT LIFE:
>10
closures at maximum signal levels;
>10
closures at low signal levels .
CONTACT RESISTANCE:
<1W .
CONTACT POTENTIAL:
<200µV .
OFFSET CUR RENT:
<1pA (<30fA typical) .
3dB BANDWIDTH:
1MHz typical .
ACTUATION TIME:
<1ms, exclusive of mainframe .
CHANNEL ISOLATION:
>10
W .
INPUT ISOLATION: Differential:
>10
W,
<50pF .
Common Mode:
>10
W,
<150pF .
COMMON MODE VOLTAGE:
<30V maximum .
Use with 7001 and 7002 switch mainframes
sub-picoamp offset current
Maintains current path for
unselected channel
bNC connectors
ordering Information
7158
low Current
scanner Card
The Model 7158 is designed to maintain the
current path even when the channel is deselected .
Input connectors are BNC for shielding of the
sensitive measurements and for compatibility with
low noise coaxial cables such as Keithley accessory
cables Models 4801 and 4803 . Two outputs are
provided to allow for chaining several scanner
cards to one measurement instrument, and an
isolation relay in the output HI minimizes interac-
tion between cards .
CHANNELS PER CARD:
10 .
CONTACT CONFIGURATION:
Single pole, simultaneous break
and make for signal HI input . Signal LO is common for all 10
channels and output . When a channel is off, signal HI is con-
nected to signal LO .
CONNECTOR TyPE:
BNC .
RELAy DRIVE CURRENT:
100mA per card typical (regardless
of channel closures selected) .
MAXIMUM SIGNAL LEVEL:
30V, 100mA peak (resistive load) .
aCCEssorIEs aVaIlablE
4801
4802-10
4803
Low Noise Male to Male BNC Input Cable
Low Noise BNC to Unterminated Cable, 10 ft .
Low Noise BNC Cable Kit for 7158
sErVICEs aVaIlablE
7158-3Y-EW
1-year factory warranty extended to 3 years
from date of shipment
Ch. 1
Ch. 2 – 9
Ch. 10
Output
accessories supplied
low Noise Male to
4801
Male bNC Input Cable
Output
Nanovolt Scanner Card
8-channel, 2-pole
The Model 7168 is an 8-channel, 2-pole card
with <30nV of thermal offset . It will switch any
one of eight signals to one output in less than
3ms . Channel offset leakage current is <50pA
at 23°C . When the 7168 is used with the Model
2182A, the noise and drift performance of the
2182A is not degraded .
CHANNELS PER CARD:
8 .
CONFIGURATION:
Two poles per channel, input HI and LO .
CONNECTOR TyPE:
Screw terminal to bare copper printed
circuit pad .
MAX. SIGNAL LEVEL:
10V, 50mA peak (resistive load only) .
CONTACT RESISTANCE:
<12W .
CONTACT POTENTIAL (HI to LO) BETWEEN CHANNELS:
<30nV when properly zeroed with supplied leads (see manual
for recommended procedure) . Typically <60nV without
zeroing .
CONTACT TyPE:
Solid state JFET switch .
ACTUATION TIME:
<3ms, exclusive of mainframe .
INPUT LEAKAGE:
<50pA per channel at 23°C .
INPUT ISOLATION:
>10
W,
<40pF between any input terminals
or between any input terminal and earth .
30V peak .
MAXIMUM VOLTAGE BETWEEN ANy TWO TERMINALS:
10V .
WARM-UP:
2 hours in mainframe for thermal stability .
HI
Ch. 1
LO
HI
Ch. 2
LO
Ch. 3 – 7
HI
Ch. 8
LO
HI
Output
LO
<30nV contact potential
bare copper terminal
connections
OPERATING ENVIRONMENT:
0°–40°C; up to 35°C at 70% R .H .
STORAGE ENVIRONMENT:
–25° to 60°C .
aCCEssorIEs suPPlIED
2107-4
7168-316
Low Thermal Input Cable for 2182A (1 supplied)
Low Thermal Input Cables for 7166 (8 supplied)
7168
8-channel Nanovolt
scanner Card
sErVICEs aVaIlablE
7168-3Y-EW
1-year factory warranty extended to 3 years from
date of shipment
1.888.KEITHLEY
(U.S. only)
www.keithley.com
A Greater Measure of Confidence
SWITCHING AND CONTROL