TECHNICAL DATASHEET
Medical AC-DC Open Frame Power Supply
48V 300W with 5V/0.5A Standby / MDS-300APB48 AA
MDS-300APB48 AA
Highlights & Features
Safety Approvals to IEC 60601-1 3.1rd ed. & IEC 60950-1
Compliant with IEC 60601-1-2 4th Ed. Requirements
Up to 240 Watt Convection in 3" x 5" x 1.4" Package
Up to 800 K Hrs MTBF
Suited for Type BF Medical Products
2 × MOPP isolation
Safety Standards
CB Certified for worldwide use
Model Number:
Unit Weight:
Dimensions (W x L x H):
MDS-300APB48 AA
405 g (0.89 lb.)
76.5 x 127.3 x 36.1 mm
3 x 5 x 1.4 inch
General Description
Delta’s new MDS-300APB48 open frame design offers a high energy density (14.3 Watts/in ) 3x5 inch design for Type BF patient
access medical products. With operating parameters of 90 to 264 Vac universal input voltage, temperatures of -10 to +70 degrees
centigrade, and altitudes of up to 5000 meters (16,400 feet), the design is well suited for a variety of both medical and non-medical
applications. Other features include input surge of 300 Vac, low leakage current, no-load input power < 0.5 Watt, and 800K hour MTBF.
This product is certified for EMC standards EN 55011 for industrial, scientific and medical (ISM) radio-frequency equipment, and EN
55022 for Industrial Technology Equipment (ITE) radio-frequency equipment. In addition, only recognized Japanese capacitors are
used to ensure long product life.
An enhanced feature set, includes a 5 Volt/0.5 amp standby output, remote on/off, remote sense, and a power good signal that are
included as part of the standard product.
The design, which has protection against shock compliant with 2XMOPP and Type BF requirements, has both medical (with risk report
available), and ITE safety approvals, including cURus(US&Canada) /TUV/GB(China); plus, CB certificates with all national deviations.
Full compliance with RoHS Directive 2011/65/EU for environmental protection is included. The unit comes in a standard 3x5 inch
footprint, with height suitable for 1U applications. Robust terminals are capable of multiple connections of the main output to the end
product.
Model Information
( With 5V/0.5A standby available)
Model Number
MDS-300APB48 AA
Input Voltage
90-264Vac
Output Voltage Convection Current Output
48Vdc
0-5A
Forced Air Current Output
0-6.25A (with 10CFM forced air)
Model Numbering
MDS
Delta Medical Power
Supply
Max wattage in the
product series.
→300W
Family Code
A½ Z
Product Type
P: Open Frame
Input Type Code
B: 3pin Class I
48
Output Voltage
48 for 48V
Revision code,
Refer to page.7
All parameters are specified at 25°C ambient unless otherwise noted.
www.DeltaPSU.com (August 2017, Rev. 14)
TECHNICAL DATASHEET
Input Ratings / Characteristics
Nominal Input Voltage
Input Voltage Range
Nominal Input Frequency
Input Frequency Range
Input Current (max)
Input Surge Voltage (max)
Full load Efficiency (typ.)
Standby Power (max)
Inrush Current (max)
Input-PE(protective earth) leakage current (max)
Output-PE(protective earth) leakage current for Type BF application (max)
Power Factor (min)
1) NC: normal condition, SFC: single fault condition
100-240Vac
90-264Vac
50-60Hz
47-63Hz
4A
300Vac for 100ms
92.5% @ 115Vac/60Hz
93.5% @ 230Vac/50Hz, Reference Fig.1
0.5W (only standby working with Inhibit signal high)
@ 115Vac/60Hz, 230Vac/50Hz
40A @ 230Vac, cold start
0.1mA @ NC , 0.3mA @ SFC
0.1mA @ NC , 0.5mA @ SFC
1)
1)
1)
1)
0.9 @ 115V/50Hz, 230V/50Hz, full load
Leakage Current
Input-PE Leakage Current
100Vac/60Hz (Typ) 264Vac/60Hz (Typ)
Normal Condition 17.5uA
Single Fault Condition 54.5uA
Output-PE Leakage Current for Type BF application
Normal Condition 29.5uA
Single Fault Condition 43.7uA
81.5uA
129.4uA
100uA max
500uA max
100uA max
500uA max
43.5uA
153uA
Delta Limit
100uA max
300uA max
IEC60601-1 Limit
5000uA max
10000uA max
Fig.1 Efficiency versus output load
TECHNICAL DATASHEET
Output Ratings / Characteristics
Nominal Output Voltage (Vrated)
Total Regulation
Output Power
Line Regulation (max)
Load Regulation (max)
Ripple & Noise (typ.)
Start-up Time (max)
Hold-up Time (min)
48V
± 3%
300W 10CFM air, 240W convection air
±0.5%
±1%
1% pk-pk Vrated@300W load (Refer to Fig. 2 & Fig. 3)
2000ms @ 115Vac
10ms @ 100Vac and 115Vac, with 300W load
20ms @ 100Vac and 115Vac, with 190W load
± 5% @ with 50-100% load change
1500uF
100ms
Up to 500mV compensation for voltage drop across
external wire connections to load.
Short and reverse connection protected.
Dynamic Response (Overshoot & Undershoot O/P Voltage)
Capacitive load (max)
Rise time (max)
Remote Sense
(a) 115V (measured value=392mV)
(b) 230V(measured value=384mV)
Fig.2 Ripple & Noise example, 20MHz BW
Output Ratings / Characteristics—Standby Output
Nominal Output Voltage of standby output (Vrated)
Total Regulation of standby output
Ripple & Noise of standby output
5V
± 3%
100mV max (Refer to Fig. 3)
PSU
Load
5cm
0.47uF Ceramic
10uF Electrolytic
Test Fixture
Fig. 3 Ripple & Noise measurement circuit
Scope
TECHNICAL DATASHEET
Mechanical
Dimensions(W x L x H)
Unit Weight
76.5x127.3x36.1mm(3x5x1.4inch)
405g (0.89 lb.)
Environment
Surrounding Air Temperature
Operating
Absolute Maximum/Minimum Rating.
-10°C to +70°C. Linear power derate from 100% load at
50°C, to 50% load at 70°C
Note: see power de-rating curves below
-40°C to +85°C
5-95% RH (Non-Condensing)
5,000 meters (16,400 feet or 50kPa)
50G, 11ms, 3 shocks for each direction
5-500Hz, 2Grms, 15 minute for each three axis
Storage
Operating Humidity
Operating Altitude
Non-operating Altitude
Shock Test (Non-Operating)
Vibration (Operating)
Power De-rating curve
Load De-rating Fixture and Test Setup. Fan is DELTA Part Number AFB0512HHD.
Airflow Direction 10CFM
TECHNICAL DATASHEET
Protections (for both 48 Volt and 5Volt Standby outputs, unless otherwise indicated)
Overvoltage (max)
Over load / Over current (max)
Main output 125% of rated voltage, Latch Mode
Standby 125% of rated voltage, Latch Mode
Main output 160% of rated current
Standby 1.06A max
Hiccup Mode(Non-Latching, Auto-Recovery)
Latch Mode for Main output
Hiccup Mode
Over Temperature
Short Circuit
Reliability Data
MTBF(Minimum) at 115Vac, 240W, 35 °C
Operating life(Minimum) at 115Vac, 240W, 25°C
Convection Air Flow
Convection Air Flow
800 Khrs based on Telecordia SR-332
26,280 hrs
Safety Standards / Directives
Medical Safety
IEC60601-1 2 , 3 and 3.1 edition CB report
rd
IEC60601-1 edition 3.1 (2012), EN60601-1 (2006) +
A11 + A1 + A12, CAN/CSA-C22.2 NO. 60601-1:14,
ANSI/AAMI ES60601-1:2005/(R)2012
IEC60950-1 CB report
TUV60950-1
UL60950-1+CAN/CSA60950-1
GB4943.1-2011, GB9254-2008, GB17625.1-2003
MDD Directive 93/42/EEC
RoHS Directive 2011/65/EU Compliant
Input to/Output (2XMOPP) 4000 Vac
1)
Input to/Ground(1XMOPP) 1500 Vac
Output to/Ground(1XMOPP) 1500 Vac (Type BF application rated)
nd
rd
rd
ITE Safety
CE
Environmental
Galvanic Isolation
1)
PSU can support PoE applications with Primary to FG 2500Vac test.